Validation of software used for product realisation

J

JeromeL

#1
Hi all,

I work for a patient-specific medical device manufacturer, and we are using a few software tools that were developed in-house for our product realization.

For instance, we have developped a web-based interface to enable doctors to send us the appropriate patient data for each work order.

Also, we have developped "macros" for our CAD software in order to save valuable production time.

Our quality consultant claims that for each software tool, we need to validate the code line by line, using thorough methods. It honestly looked more complicated than the coding itself.

My feeling, based on everything I have read so far, is that when a process' output is entirely measurable, it is sufficient to demonstrate that it matches the requirements specified as the input, and the process itself doesn't need to be directly validated.

For our web file transfer interface, we could use a significant sample of test cases to demonstrate the transfer occurs the way it should.

For our CAD software (3rd party, well-known) macros, we can easily make virtual measurements on the CAD part, make real measurements on the milled part, and also test it for fit on standard parts to demonstrate the input requirements are met.

Would anybody have suggestions on the matter, or online references, guidelines, etc. I could look up?

Thanks!
 
Elsmar Forum Sponsor

Gert Sorensen

Forum Moderator
Moderator
#3
Hi all,

I work for a patient-specific medical device manufacturer, and we are using a few software tools that were developed in-house for our product realization.

For instance, we have developped a web-based interface to enable doctors to send us the appropriate patient data for each work order.

Also, we have developped "macros" for our CAD software in order to save valuable production time.

Our quality consultant claims that for each software tool, we need to validate the code line by line, using thorough methods. It honestly looked more complicated than the coding itself.

My feeling, based on everything I have read so far, is that when a process' output is entirely measurable, it is sufficient to demonstrate that it matches the requirements specified as the input, and the process itself doesn't need to be directly validated.

For our web file transfer interface, we could use a significant sample of test cases to demonstrate the transfer occurs the way it should.

For our CAD software (3rd party, well-known) macros, we can easily make virtual measurements on the CAD part, make real measurements on the milled part, and also test it for fit on standard parts to demonstrate the input requirements are met.

Would anybody have suggestions on the matter, or online references, guidelines, etc. I could look up?

Thanks!
All software used for production, and documentation is required to be validated. Macros need to be validated more thorough, because they are not a part of the OTS software (which also needs to be validated). Web interface for patient data transfer needs to be validated and it needs to adhere to regulation regarding patient data safety.

The thing about software is that it is much harder to control the actual result and actions of it, than it is for a process. Therefore the focus is on validation, documentation, and built-in checks and balances.

Depending on your product and your markets you may need to look various places for the legislation (you may need a lawyer to help you with that). Besides that, FDA's guideline for software validation is a good starting point.
 
W

wrodnigg

#4
For medical devices there is the standard AAMI TIR36:2007:
Validation of software for regulated processes

Abstract:
Applies to any software used to automate device design, testing, component acceptance, manufacturing, labeling, packaging, distribution, and complaint handling or to automate any other aspect of the quality system as defined by the Quality System Regulation (21 CFR 820). In addition, it applies to software used to create, modify, and maintain electronic records and to manage electronic signatures that are subject to the validation requirements (21 CFR 11). This TIR can also be broadly applied wherever software automates processes regulated by the FDA.
This TIR applies to software used in the production of a device and to software used in implementation of the device manufacturer?s quality system. It does not apply to software used as a component, part, or accessory of a medical device or software that is itself a medical device.
You can find a preview here.
 

yodon

Staff member
Super Moderator
#5
For our web file transfer interface, we could use a significant sample of test cases to demonstrate the transfer occurs the way it should.
Just to add on to the excellent advice so far, the above caught my eye.

This would be, IMO, demonstrating what we call "happy path." It all works quite nicely if everything is perfect. It's often when you venture off the happy path that you get into trouble. Part of validation should include some level of robustness testing, error checking, etc. Since it's a web interface, you also need to consider the likelihood of data transmission errors. Hopefully you have things like checksums built in to avoid that but if not, then you need to consider the event.

The statement about validating the code line by line is quite bogus. You DO need to validate (as the others have indicated) and it does need to be thorough (as your consultant mentioned) but the method(s) are up to you. They should be well-planned in a Validation Plan document which should describe how you will (thoroughly) validate.
 
Thread starter Similar threads Forum Replies Date
V Validation of macro - scripts - programs used in statistical software (Minitab-SAS... Qualification and Validation (including 21 CFR Part 11) 5
N Validation of CAD Software used to generate our Engineering Drawings? Qualification and Validation (including 21 CFR Part 11) 11
J Interpreting clause 7.5.2.1 (validation of software used in production & service) ISO 13485:2016 - Medical Device Quality Management Systems 2
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 8
K ERP System Software Validation - ISO13485 2016 4.1.6 Design and Development of Products and Processes 3
D Software validation in Medical Equipment Other Medical Device and Orthopedic Related Topics 20
C Looking for simple Software Validation IQ templates. Qualification and Validation (including 21 CFR Part 11) 4
C Software validation - Off The Shelf Software - Web hosted ISO 13485:2016 - Medical Device Quality Management Systems 6
R Validation of Medical Device Hardware containing Software - How many to Validate ISO 13485:2016 - Medical Device Quality Management Systems 1
F 21 CFR Part 11 - Implicit requirements - Validation plan for a Software as a Service Other US Medical Device Regulations 1
R ISO 13485 Software validation procedure and Quality Objectives Monitoring wanted Document Control Systems, Procedures, Forms and Templates 1
S Validation of COTS Equipment plus Software Qualification and Validation (including 21 CFR Part 11) 12
D Software Validation - Contract manufacturer of Components (PCBA's) Qualification and Validation (including 21 CFR Part 11) 7
Pmarszal Software Validation Training Course - Recommendations Training - Internal, External, Online and Distance Learning 3
T Software Validation Certificate (ISO 13485:2016) ISO 13485:2016 - Medical Device Quality Management Systems 19
R FDA Requirements - Printing Equipment Software Validation Qualification and Validation (including 21 CFR Part 11) 1
G Windows 10 OS build Software Validation US Food and Drug Administration (FDA) 1
S Where to keep Enterprise Resource Planning software (ERP) Validation Records ISO 13485:2016 - Medical Device Quality Management Systems 1
C Software validation (4.1.6 ISO 13485:2016) ISO 13485:2016 - Medical Device Quality Management Systems 20
M Software Validation Guidance Suggestions Various Other Specifications, Standards, and related Requirements 6
S QMS software validation - Documentation ISO 13485:2016 - Medical Device Quality Management Systems 1
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
R CNC Software Validation requirements as per ISO 13485:2016 Other ISO and International Standards and European Regulations 8
R Software validation - off the shelf X-Ray Software Quality Assurance 3
S What is the clause in ISO 13485 for SAP Software Validation? ISO 13485:2016 - Medical Device Quality Management Systems 3
P Software validation of Data Exporter ISO 13485:2016 - Medical Device Quality Management Systems 3
A Process Validation of QMS Software ISO 13485: 2016 Cl. 4.1.6 ISO 13485:2016 - Medical Device Quality Management Systems 26
P Software Validation for Equipment - Question ISO 13485:2016 - Medical Device Quality Management Systems 5
D Medical Device Software Tool Validation - Compilers! IEC 62304 - Medical Device Software Life Cycle Processes 7
N When is Medical Device Software Validation required? ISO 13485:2016 - Medical Device Quality Management Systems 6
O Process Mapping prior to Validation and Software to use to Map Process Maps, Process Mapping and Turtle Diagrams 12
SteveK Software Validation – Clause 4.1.6 of ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 12
B Could I combine IQ, OQ and PQ for Minitab Software Validation ? Software Quality Assurance 3
R Document Management Software : Validation and other requirements Medical Information Technology, Medical Software and Health Informatics 8
E Software Validation - Clinical Trials US Food and Drug Administration (FDA) 3
K What are the minimum requirements for Process Validation (Software)? ISO 13485:2016 - Medical Device Quality Management Systems 5
R ISO - Clause 7.5.1.3 - CMM Software Program Validation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
I Validation Software Experience? (Valgenesis, etc.) Quality Assurance and Compliance Software Tools and Solutions 1
M Injection Molding Machine Software Validation Calibration and Metrology Software and Hardware 2
E FDA Requirements for Implantable Medical Device Software Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
C Change Control Forms Post Software Validation Medical Information Technology, Medical Software and Health Informatics 2
C Thoughts on validation of Legacy Systems for Medical Device Software EU Medical Device Regulations 2
C Cement Mixing Software Validation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Y OTS (Off The Shelf) Software Validation for 510k Traditional 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R ISO 13485 Software Validation Requirements - Help needed ISO 13485:2016 - Medical Device Quality Management Systems 4
M MSA - Validation of Metrology Software such as PCDMIS Calibration and Metrology Software and Hardware 2
T Class II Software Device 510k V&V (Verification and Validation) Criteria and Results 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A ClassIIa Medical Device Software Validation Report Format and Content IEC 62304 - Medical Device Software Life Cycle Processes 5
A Medical Device Software Validation Guidance Software Quality Assurance 3
T Validation of OTS (Off The Shelf) Software in Medical Devices IEC 62304 - Medical Device Software Life Cycle Processes 12