Hi,

I was hired to improve the system. There is no proper job done in the current. Especially validation. And many other areas.

Since they are moving to new plant with new facility, they could like to take the chance to get everything in place. Which means everything have to start fresh.

Hope this clear your doubts.

Thanks & Regards,

Lisa
 
Elsmar Forum Sponsor
A

alex.Kennedy

#12
Hi Lisa

You are really asking how to open a hazel nut using a road roller.

It is a new site; it requires to be incorporated into your Validation Master Plan (VMP) and everything requires to be validated.

Facilities - Utilities - processes - equipment - training - documentation.

Your list is not logical. Be very careful - there are no shortcuts that are acceptable to the regulators. cGMP must be implemented throughout your new facility.

If you use any none validated processes or equipment in the production/testing/storage of your product the regulators may declare your product adulterated.

We could certainly advice you - but you are talking about a large amount of work.

What time scale have you?

Alex
 
K

KLDLetts - 2012

#13
I agree with Alex, your plan will need to be all-encompassing. I am currently working on a validation project that moved production (and related equipment) from one plant to another.
We are validating every thing from Facility - Utilities - Analytical Methods - Process IOQ & PQ and Product Validation along with the required documentation.

The amount of change that your process undergoes in the transfer between plants should have also triggered your re-validation criteria.
 

lisawxf80

Involved In Discussions
#14
Thanks Alex.

Our company is aggressively planning everything to be done within few months which I think it must take about years. We have limited resource here. And QA is leading the road.

As such, we could only do the foundamental. As per listed.

Thanks.
 

lisawxf80

Involved In Discussions
#15
I agree with Alex, your plan will need to be all-encompassing. I am currently working on a validation project that moved production (and related equipment) from one plant to another.
We are validating every thing from Facility - Utilities - Analytical Methods - Process IOQ & PQ and Product Validation along with the required documentation.

The amount of change that your process undergoes in the transfer between plants should have also triggered your re-validation criteria.
Hi,

Thanks for the advice. May I know more detail your validation master plan has covered. We have no experience ppl here in the company. I hope to do it in a proper way. Appreciate.

Regards,

Lisa
 
K

KLDLetts - 2012

#16
Lisa,

We are in the exact position you are, we have no experienced staff on site and management wants it done as fast as possible but without using any resources. (Which means it will take a very long time to get done...).
We had a consultant visit earlier this month to help us with preparation, and he had us completely gut our VMP. My task this week is to bring it up to snuff, I will post what I can once that is complete!

He did recommend virtually starting from scratch in the VMP and building it as you progress so that you don't get bogged down in the details. For us this meant taking out our other production lines (that we want to validate in the future, but aren't mission critical to bringing the assigned registered products to our facility), and limiting the scope of the VMP to only the IOQ portion for the lines and utilities. Once those are written up, then I will go through the VMP again to add in the required information for the PQ, and then again for cleaning validation, product validation, analytical methods, etc.
 

v9991

Trusted Information Resource
#17
He did recommend virtually starting from scratch in the VMP and building it as you progress so that you don't get bogged down in the details. For us this meant taking out our other production lines (that we want to validate in the future, but aren't mission critical to bringing the assigned registered products to our facility), and limiting the scope of the VMP to only the IOQ portion for the lines and utilities. Once those are written up, then I will go through the VMP again to add in the required information for the PQ, and then again for cleaning validation, product validation, analytical methods, etc.
1)
if i understand you post, you intend to bring VMP upto IOQ level of details; validate the systems & equipments and then expand the VMP to PQ level of details!!!
Its interesting approach to build upon the MVP in a phased manner; (first IOQ & followed/later add the PQ part) in order to implement this...you might consider additional fail safe requirements for product (which are manufactured during the period of IOQ --> PQ completion); this is usually manageable for lesser-complex products and personnel; But, i haven't seen this approach anytime before. It will be interesting to see how this pans out in context of compliance to regulations, and other areas!!!

I would like to caution you from ignoring the risk assessment;

2)
It might have worked well in your case, because, you seem to have segregated the product specific lines; that seems a practical approach, BUT in case of OPs scenario, they plan to move into new facility (with almost all new systems/equipments in place!!!)
 
K

KLDLetts - 2012

#18
We won't be manufacturing any registered products until all of the validation work up to the product validation is done.

We were also instructed to work on our risk assessment (don't know why I left that out before).

I think my situation is similar to the OPs in that we moved two existing packaging lines from a sister plant to our plant, which necessitated a fair amount of system changes in terms of our utilities, facilities etc. Essentially we are starting from scratch in terms of validation - everything is 'new' to us, as for the OP.
 

lisawxf80

Involved In Discussions
#19
Lisa,

We are in the exact position you are, we have no experienced staff on site and management wants it done as fast as possible but without using any resources. (Which means it will take a very long time to get done...).
We had a consultant visit earlier this month to help us with preparation, and he had us completely gut our VMP. My task this week is to bring it up to snuff, I will post what I can once that is complete!

He did recommend virtually starting from scratch in the VMP and building it as you progress so that you don't get bogged down in the details. For us this meant taking out our other production lines (that we want to validate in the future, but aren't mission critical to bringing the assigned registered products to our facility), and limiting the scope of the VMP to only the IOQ portion for the lines and utilities. Once those are written up, then I will go through the VMP again to add in the required information for the PQ, and then again for cleaning validation, product validation, analytical methods, etc.
Hi,

Thanks, that is wonderful. Appreciate that if can view the VMP once you have.

Thanks a lot.

Regards,

Lisa
 
Thread starter Similar threads Forum Replies Date
F 21 CFR Part 11 - Implicit requirements - Validation plan for a Software as a Service Other US Medical Device Regulations 1
I GMP Stores qualification/validation plan or protocol - New Store requirements ISO 13485:2016 - Medical Device Quality Management Systems 3
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
B Sterilization Validation Plan Other Medical Device Related Standards 3
J Requested Validation plan and reports Manufacturing and Related Processes 4
Mr Skeleton PV (Production Validation) test plan responsibility APQP and PPAP 21
E Sampling plan for orthopedic implant - Process Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
M Does anyone has a good verification and validation plan template? ISO 13485:2016 - Medical Device Quality Management Systems 3
K Computerized system Master Validation Plan, IQ-OQ-PQ Draft or Template Qualification and Validation (including 21 CFR Part 11) 4
M Example Medical Device Validation Plan and Report wanted CE Marking (Conformité Européene) / CB Scheme 4
T Validation Plan for a new Class II Medical Device (Critical Care Medical Ventilator) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 62366-1:2015 Usability Verification / Validation Plan example wanted IEC 62366 - Medical Device Usability Engineering 20
M Justification for Not Having a Validation Plan for a Critical Process ISO 13485:2016 - Medical Device Quality Management Systems 4
C Validation Plan for a Printer Manufacturing and Related Processes 2
F Usability Objectives and Verification & Validation Plan IEC 62366 - Medical Device Usability Engineering 6
M Manufacturing Process Validation Destructive Testing Sampling Plan Advice Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
J Welding Process Validation Plan Qualification and Validation (including 21 CFR Part 11) 19
G Sampling Plan - Sample Size Justification for a Bottling Cleaning Process Validation Qualification and Validation (including 21 CFR Part 11) 2
K Definition MVP (Master Validation Plan) and VMP (Validation Master Plan) - Differences Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 4
S Validation Plan Template for Supporting Software Tools (subversion, bugzilla,etc) IEC 62304 - Medical Device Software Life Cycle Processes 1
E How to write a Validation Risk Management Plan for Equipment ISO 14971 - Medical Device Risk Management 5
T New Product Validation Plan example or template for a Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P Best practice on documents turning into records, test and validation plan ISO 13485:2016 - Medical Device Quality Management Systems 9
A High Level Validation Plan template for a Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Master Validation Plan - What must be included in a "Master Validation Plan"? Qualification and Validation (including 21 CFR Part 11) 5
M Validation Project Plan example Document Control Systems, Procedures, Forms and Templates 3
E Validation master plan - Existing cleanroom facility that will be upgraded ISO 13485:2016 - Medical Device Quality Management Systems 11
D Control Plan need to cover the Product Validation Stage? IATF 16949 - Automotive Quality Systems Standard 13
Ajit Basrur Does anyone have a template for Validation Master Plan to share? Qualification and Validation (including 21 CFR Part 11) 35
N Test plan of Design Validation Plan and Report (DVP & R) Design and Development of Products and Processes 7
T Anybody could share an example of Ford PVP (production validation plan/program) APQP and PPAP 2
Marc DVP&R Design Validation Plan and Report - A Ford, GM and Chrysler thing Design and Development of Products and Processes 2
mustomutlu Process Validation Final Report Other Medical Device and Orthopedic Related Topics 2
P Software verification and validation procedure IEC 62304 - Medical Device Software Life Cycle Processes 4
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 3
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3

Similar threads

Top Bottom