We want to move our Label Database System after moving the operating files from one network drive to the other network drive, what should be validated for this case?
I would be very appreciate on your help in this matter.
 
Elsmar Forum Sponsor
S

shala

#3
Labeling being used for IVD devices. we are using this system for long time and now we want to move this system to other network drive. our labeling software is Microsoft access database program.
we want to validate functionality of labeling system after moving to new drive.
 

Jen Kirley

Quality and Auditing Expert
Staff member
Admin
#4
I would validate using a short test with data that is likely to be needed. If there are multiple sectors or otherwise chunks of data that could become lost, I would outline them and validate all of them if possible.
 
S

shala

#5
Thanks for reply. Do you think printing labels and comparing the similarity with original one and scanning barcode for checking accuracy for all labels type is enough?
do I need to repeat validation of computer or barcode scanner too?
 

Marc

Fully vaccinated are you?
Staff member
Admin
#6
I would think so. Just moving the database to a new drive shouldn't affect the database. The part that will be important is that the program that uses the database for inputs and outputs is reconfigured to point to the new database location.

As Jennifer said, do a short run of labels. Then use your barcode scanner to read them. You'll be validating the database, the barcode scanner and the computer all at the same time. I would also include input, though - Make a new label, print it and then scan it.
 
M

MIREGMGR

#7
Technically, creating a particular label and checking it is verification. To validate a system for which there is no overriding rule or guidance that allows you to do less, you must test in such a way that you've established that the system works correctly with all of the potential variables exercised across their full ranges, and with any interactions among variables investigated.
 
Thread starter Similar threads Forum Replies Date
M Requirement for manufacturing process validation in IATF 16949 IATF 16949 - Automotive Quality Systems Standard 8
S Requirement for Sterilization Validation completion before issuance of ISO cert ISO 13485:2016 - Medical Device Quality Management Systems 0
N ETO Sterilisation Validation - EO Residual Minimum Sample Requirement ISO 13485:2016 - Medical Device Quality Management Systems 2
S Sterilization Validation Accreditation - A mandated standard requirement? ISO 13485:2016 - Medical Device Quality Management Systems 5
C Software Validation Procedure per FDA requirement Qualification and Validation (including 21 CFR Part 11) 3
I CNC Machine Validation Requirement ISO 13485:2016 - Medical Device Quality Management Systems 20
I AS9100B Cl 7.5.1.3 - How to comply with validation of production tools requirement? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
H EU Validation Requirement vs. FDA Validation Requirement - EN 62304 IEC 62304 - Medical Device Software Life Cycle Processes 16
C Meeting Validation of Process Requirement - Stainless steel pipe & tube manufacturer ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Manix GM Requirement for Trim Clip Validation - Anyone know a specification number Customer and Company Specific Requirements 4
A Validation of Processes - How to meet requirement of TS 16949 para. 7.5.2.1? IATF 16949 - Automotive Quality Systems Standard 12
R System Requirement - FDA - Software Validation on various OSs ISO 13485:2016 - Medical Device Quality Management Systems 13
J 7.5.2.1 Validation of processes - supplemental requirement IATF 16949 - Automotive Quality Systems Standard 11
N Validation of special processes Oil and Gas Industry Standards and Regulations 2
L Validation without Tolerance Qualification and Validation (including 21 CFR Part 11) 0
shimonv Test Method Validation ISO 13485:2016 - Medical Device Quality Management Systems 10
R Debug mode in software/device validation IEC 62304 - Medical Device Software Life Cycle Processes 2
M Software verification and validation AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Melissa Process Validation of Rotary Heat Sealer Speeds Design and Development of Products and Processes 4
C Validation of process for releasing the UDI EU Medical Device Regulations 4
C Validation of process for production and servicing 5.7.1.5 API Spec Q1, 9th Edition Oil and Gas Industry Standards and Regulations 4
W Transport validation Qualification and Validation (including 21 CFR Part 11) 4
M Is validation required when consumables are changed Qualification and Validation (including 21 CFR Part 11) 6
C 8.3.4 Verification and Validation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
W Transport validation Qualification and Validation (including 21 CFR Part 11) 12
mustomutlu Process Validation Final Report Other Medical Device and Orthopedic Related Topics 2
P Software verification and validation procedure IEC 62304 - Medical Device Software Life Cycle Processes 6
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 8
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 1
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 3
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
T Laboratory Verification after validation ISO 17025 related Discussions 3

Similar threads

Top Bottom