My employer is 13485 certified. However, lots of our suppliers (providers of parts and/or raws) are only 9001 certified. There are validation requirements in both 13485 and 9001. The wording in 9001 is "weak" but the basics seems the same. What would be the reasonable requirement to these 9001 suppliers in process/equipment validation? Thanks! (We make cable assemblies for medical devices).
 
Elsmar Forum Sponsor
G

Gert Sorensen

#2
Re: validation requirement: 9001 vs 13485?

My employer is 13485 certified. However, lots of our suppliers (providers of parts and/or raws) are only 9001 certified. There are validation requirements in both 13485 and 9001. The wording in 9001 is "weak" but the basics seems the same. What would be the reasonable requirement to these 9001 suppliers in process/equipment validation? Thanks! (We make cable assemblies for medical devices).
The requirement to validate a process is usually:

*That you cannot verify the quality afterwards.
*That the process is considered critical
*That you want to use the validated state for marketing purposes.

IMO you only validate processes where there is a very good reason to do it. Validation is an expensive and tiresome process, and you can practically break a company if you want to validate everything.

If your suppliers provide you with 1 or 2 of the above you should require validation, if not you should have a good description of the quality that you expect. :bigwave:
 
Last edited by a moderator:
M

maxwell

#4
As long as you can verify the resulting output by measurement thier is no need to validate. For cables I would think physical measurements and functional testing would suffice.
 

RCW

Quite Involved in Discussions
#5
(We make cable assemblies for medical devices).
Do you perform any soldering functions during the manufacturing of your cable assemblies? Many companies consider soldering a special process which would fall under the validation umbrella.

Just providing food for thought.
 
T

treesei

#6
RCW,

Yes we do.

But is soldering a process that I can verify the quality afterwords?

Thanks, the food being delicious.:)
 

RCW

Quite Involved in Discussions
#7
But is soldering a process that I can verify the quality afterwords?
I have had to debate this one before from both sides of the fence.

Yes, you can visually inspect for the contour of the solder fillet and for any solder defects such as voids or blow-holes. This is usually a good indication of a proper solder joint.

No, you cannot visually inspect the intermetallic bond, where the true solder bond occurs. You could do destructive testing by cutting open a solder joint to visually inspect it but is there really a need to go that far? (I vote no.)

That's probably why some (but not all) customer companies want you to validate your soldering processes. It really depends on your customer(s) and how you want to handle it.
 
D

DRDDO

#8
Dear Friend,
I recommend you read: GHTF/SG3?N99-1-: 2004, QMS -process validation Guidance,
By Google, type: GHTF/SG3?N99-1-: 2004
or Type GHTF-Process validation,
Very good Document,
:thanx:
DRDDO
 
P

phenol

#9
RCW,

Yes we do.

But is soldering a process that I can verify the quality afterwords?

Thanks, the food being delicious.:)

Hi,

Some throughts on verifying the quality afterwords.

Just give you an example for products like cosmetics and toiletries in which a critical process (Mixing- e.g. time & temperature controlled) is involved will serious affect the stablility of the products, in this case it is difficult to measure the safety level & quality of product immediately until after certain period of time, there's a need to validate the critical process.

To decide whether process validation is needed it's depending on those processes are seriously affecting customer or product requirements.

Phenol
 

harry

Super Moderator
#10
My employer is 13485 certified. However, lots of our suppliers (providers of parts and/or raws) are only 9001 certified. There are validation requirements in both 13485 and 9001. The wording in 9001 is "weak" but the basics seems the same. What would be the reasonable requirement to these 9001 suppliers in process/equipment validation? Thanks! (We make cable assemblies for medical devices).
You can't really compare the two. ISO 13485 is industry specific whereas as ISO 9001 is purposely being vague because it has to cater to industries ranging from high tech and complex to low tech and simple ones.

Then there are product certifications to consider also. An ISO 9001 certified manufacturer of cables would have different processes, validation methods and testing methods for the different type of 'certified' cables it produce. Thus, 'VDE' certified cables had different testing and validation requirements than say a 'JIS' certified cable.
 
Thread starter Similar threads Forum Replies Date
R CNC Software Validation requirements as per ISO 13485:2016 Other ISO and International Standards and European Regulations 8
N ISO 14644-1 Clean Room Occupancy State Validation Requirements Manufacturing and Related Processes 3
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
R ISO 13485 Validation and CAR Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
R ISO 13485 Software Validation Requirements - Help needed ISO 13485:2016 - Medical Device Quality Management Systems 4
I Work Instruction for Validation of Packaging to EN ISO 11607-2 Requirements Document Control Systems, Procedures, Forms and Templates 3
F ISO 13485 & FDA Requirements - What kinds of software require validation? ISO 13485:2016 - Medical Device Quality Management Systems 2
S ISO 17025 Validation Requirements - "calibration using reference standards or..." ISO 17025 related Discussions 5
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
M Validation Requirements - ISO 13485 / MDD Technical File ISO 13485:2016 - Medical Device Quality Management Systems 7
Sidney Vianna ISO 14065:2007, Greenhouse gases ? Requirements for greenhouse gas validation Sustainability, Green Initiatives and Ecology 5
E Contract manufacturer seeking help understanding ISO 13485 Validation requirements ISO 13485:2016 - Medical Device Quality Management Systems 9
R Design Validation and Clinical Evaluations to meet ISO 13485 requirements Design and Development of Products and Processes 3
Q Validation of New Process Equipment - What are the ISO 9001 requirements? Misc. Quality Assurance and Business Systems Related Topics 6
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
F 21 CFR Part 11 - Implicit requirements - Validation plan for a Software as a Service Other US Medical Device Regulations 1
R FDA Requirements - Printing Equipment Software Validation Qualification and Validation (including 21 CFR Part 11) 1
Ed Panek Verification & Validation requirements - Patch that adheres to skin ISO 13485:2016 - Medical Device Quality Management Systems 2
N China's CFDA - Medical Device Validation Requirements China Medical Device Regulations 8
D User Requirements Tracing - Verification and Validation Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S COTS (commercial off-the-shelf) Validation FDA Requirements Software Quality Assurance 4
R Document Management Software : Validation and other requirements Medical Information Technology, Medical Software and Health Informatics 8
A Products for Design Validation - FDA Requirements US Food and Drug Administration (FDA) 3
N Do new requirements always need a validation? IEC 62304 - Medical Device Software Life Cycle Processes 5
C Saboraud's Agar Incubation Temperature and Validation Requirements Other Medical Device and Orthopedic Related Topics 10
K What are the minimum requirements for Process Validation (Software)? ISO 13485:2016 - Medical Device Quality Management Systems 5
R TS 16949 Process Re-Validation Frequency Requirements IATF 16949 - Automotive Quality Systems Standard 3
E How to write requirements for Medical App Validation and Verification IEC 62366 - Medical Device Usability Engineering 7
T "Special Process" Validation Requirements (21 CFR Part 820) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
Z Final Validation Unit Requirements - Annex D of IEC 62366 IEC 62366 - Medical Device Usability Engineering 6
E FDA Requirements for Implantable Medical Device Software Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
I Minimum and Maximum Validation Batch Size Requirements Qualification and Validation (including 21 CFR Part 11) 3
X EtO Steriliser Software (PACS) Update Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
A Eto Sterilisation Re-Validation Requirements Other Medical Device Related Standards 4
E What are Validation Requirements for Vendor (Supplier) Processes Supplier Quality Assurance and other Supplier Issues 5
A IVD Label Requirements - For Validation Only Other Medical Device Regulations World-Wide 4
P Sterile Packaging Validation Requirements - Accredited Laboratory or not? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R Only Commissioning Software? COTS, GAMP Category 3 Software Validation Requirements IEC 62304 - Medical Device Software Life Cycle Processes 7
L Validation Plan Requirements when moving to a new Plant - Medical Device Industry Qualification and Validation (including 21 CFR Part 11) 19
Q Software Validation - FDA Regulations and Requirements US Food and Drug Administration (FDA) 4
P Sterile Packaging Validation - FDA Requirements for Statistically Valid Tests 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
A Software Validation Requirements for Class I Active Medical Device EU Medical Device Regulations 4
M CCP Validation and Pre-Requisite Program Reassessment Requirements Food Safety - ISO 22000, HACCP (21 CFR 120) 4
B Validation Method Requirements - 21 CFR Part 820 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 14
R CLIA Equipment or Software Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5
R Clean Room Validation Requirements ISO 13485:2016 - Medical Device Quality Management Systems 7
R Equipment Cleaning Validation Requirements - Medical Device Class II or Class III Other Medical Device and Orthopedic Related Topics 5
V "Value" and "Overhead(s)" of implementation of "CSV - Validation requirements" Qualification and Validation (including 21 CFR Part 11) 1
M Microsoft Sharepoint as Document Control Software - FDA Validation Requirements Document Control Systems, Procedures, Forms and Templates 15
R Swab, Rinse and Visual Cleaning Validation Requirements Qualification and Validation (including 21 CFR Part 11) 5

Similar threads

Top Bottom