Validation vs. Calibration of Centrifuges

#1
I've read up on the Calibration vs Validation FAQ, but it didn't clarify my question. In my previous job, which was in the blood banking sector, they validated our centrifuges. They were separating the red cells from the plasma. At my current employment, medical device world, we only calibrate our centrifuges. The centrifuges are used to separate one component from another. A validation has never been performed on the centrifuges. Is calibration of speed acceptable in this case or would it need to be validated? Thanks in advance for your assistance.

G
QA Manager
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#2
Hello there! Welcome to the Cove!:bigwave:

Essentially, I would consider calibration a subsection of validation. Validation is a bit more of a formal process of verifying suitable use, installation, performance, assessment against functional criteria, etc.

The matter of whether you require a validation would fall more under a regulatory requirement issue and/or a requirement to fulfill any organizational validation guidelines.

But a calibration to assure proper use is definitely a minimal verification.
 

raghu_1968

Involved In Discussions
#3
I am in the process of qualifying the VFD (Variable frequency Drive) associated with the centrifuge.

I also have the same confusion that VFD has to be calibrated or qualified.

Expeted some clarification on this.

Regards / Raghu
 

Ajit Basrur

Staff member
Admin
#4
I've read up on the Calibration vs Validation FAQ, but it didn't clarify my question. In my previous job, which was in the blood banking sector, they validated our centrifuges. They were separating the red cells from the plasma. At my current employment, medical device world, we only calibrate our centrifuges. The centrifuges are used to separate one component from another. A validation has never been performed on the centrifuges. Is calibration of speed acceptable in this case or would it need to be validated? Thanks in advance for your assistance.

G
QA Manager
Hi KingQA,

Both the terms viz. Validation and Calibration are applicable at different stages.

If you go with the definition of Validation, it says, " A documented evidence that provides a high level of assurance that a specific process will consistently produce a product that meets its predetermined specifications and quality". Therefore, The Validation (which includes the Equipment Qualification for Installation Qualification (IQ), Operational Qualification (OQ)and Performance Qualification (PQ) are performed before putting the equipment into use. In case of exisiting equipment, a retrospective Qualification / Validation can be performed to have consistency in operations. Validation (besides Calibration of any indicators) shall include various aspects like Imbalance detector accuracy, Speed, Safety features like automatic lid lock, Imbalance detector etc.

While, Calibration is performed on measuring or monitoring systems like gages etc. If there is a rpm (revolutions per minute) indicator, that shall be calibrated.

Thus calibration is a subset in the "umbrella" of validation.

Hope it clarifies.
 

BradM

Staff member
Admin
#5
Good answer, Ajit. I like the thumbnail definition. As far as the VFD, I would consider it as a functional part of the centrifuge. To me, it would be validated with the centrifuge.

As far as calibration, it's hard to give a definite answer. But in general, I would think it's operation (proper adjustment of the motor speed) would be verified when you calibrate the instruments of the centrifuge.
 

jkuil

Quite Involved in Discussions
#6
You calibrate the measuring and monitoring functions of the equipment: e.g. determine whether the equpment meets the accuracy and precision specs.

You validate the equipment against its user requirements: is the centrifuge consistently seggregating all red cells from your plasma (or whatever specification you have defined). In order to meet the use requirement, it needs to be calibrated.
 
J

Juan Dude

#7
You calibrate the equipment, you validate the process, one does not necessarely exclude the other. Ideally you want to perform both.
 
J

JaxQC

#8
One way to view it is that (made up numbers) the calibration verifies the machine used to spin blood spins at 12K rpm +/- 300 rpm.

Validation would verify that spinning actual samples of blood, at that speed, on that machine, actually separates it into a solution 98% +/- 1% pure with maximum 3% loss due to damaged cell structure (it’s a made up specification but the point is that the spec the Customer is looking for is not the 12K rpm, it’s that the blood is separated). The 12K rpm is just an one indicator that the process is still working.

The calibration just verifies it meets the physical requirements while validation tells if it’s doing the job you want it to do which should hopefully match. ie Prove it’s doing the job.
 
Thread starter Similar threads Forum Replies Date
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
K Validation of an Analytical Method that relies on Multi-Variate Calibration General Measurement Device and Calibration Topics 7
S Can Calibration of an instrument substitute for Validation (IQ, OQ,PQ)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
P Clean Room Calibration (Validation) Other ISO and International Standards and European Regulations 4
V Image Analysis - Particle Measurement (Characterization) - Validation - Calibration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
W Is an Oven I/O/Q Validation Required or is Calibration Sufficient? Qualification and Validation (including 21 CFR Part 11) 7
S ISO 17025 Validation Requirements - "calibration using reference standards or..." ISO 17025 related Discussions 5
V "Calibration" or "Validation with CMM machine" General Measurement Device and Calibration Topics 3
R Validation of outsourced processes - Calibration - How do we validate that? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
C Software Validation - GageTrak Calibration Software - Please Help Calibration and Metrology Software and Hardware 6
S Verification vs. Validation vs. Calibration - What is the difference? General Measurement Device and Calibration Topics 38
A Calibration Software Validation - Are there any Testing Tools available? Calibration and Metrology Software and Hardware 8
T Anyone have an example of a calibration method validation procedure? General Measurement Device and Calibration Topics 4
Marc Brute Force Validation - Difficult to show 10-year calibration cycle is sufficient General Measurement Device and Calibration Topics 0
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 2
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 2
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
T Laboratory Verification after validation ISO 17025 related Discussions 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 6
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 2
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
T Annual Validation as a detection mode on a PFMEA? FMEA and Control Plans 5
B TMV - Selection of TM's for Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4

Similar threads

Top Bottom