Validity of 7 Year Old Calibration Certificate
sarasheed
Our Laboratory is certified to ISO 17025:2005.We have procured a Master slip Gauge Box from Mutituyo on 15-10-2006 with calibration certificate traceable to Japan National Standard and put use as it was in good condition suitable for useing as a Master Standard.Date of certificte is 21-10-1999 and there is no indication about the validity of the certificate.The Auditor has raised a Major Non-confirmance since it has been calibrated 7 years back which is very long period.We have set a periodicity of calibration as 2 years in our system.
Is the Auditor's decision in raising the NC is correct?.We presumed that we can use it upto 15-10-2008 which is 2 years from the starting date of use .
sarasheed
Jen Kirley
Quality and Auditing Expert
Re: Validity of Test certificte
Even though 7 years is a long period, I am confused about why you were iussued a major NC. Can you tell us exactly what the NC says?
Jerry Eldred
Since the certificate is seven years old and you have a two year interval, you would have to justify that it is valid to begin your calibration interval on the first date you used it. And you would also have to justify what has happened to the unit for the years since 1999.
Since it was calibrated in 1999, you have to prove that nothing has changed in such a long time (since it is multiples of your entire calibration interval).
I unfortunately would have to agree that if it were my standard, I would definitely send it out and have it re-certified rather than take the risk of the amount of time passed since it was calibrated something had possibly changed.
The only exception would be if you just bought it new from the manufacturer still in it's original box, and they gave you written justification that it is still valid to use after all those years since the original calibration date.
Sorry to be the bearer of a difficult position.
BradM
I agree entirely with Jerry's reply. You have no idea where/how this standard has been stored, the use of it before you got it, etc. Was this brand new? How had it been stored? Etc.
Did you perform any type of cross-verification with other standards in your lab prior to use? Did you perform even a simple 1 to 1 check of some sorts?
You need to have this thing re-certified. The cost of recertification can be insignificant to the cost of making poor measurements with a potentially O.O.T. standard. Agreed?
Now, just my opinion, but I don't agree with the major NC on this one. But I do think it should be sent in.
L
lawriematkin
Re: Validity of Test certificate
As you procured in 2006. Proof that the measurement equipment still conformed to requirements is required. As this is used as a master calibrator the auditor may have taken the N/C as systemic as all other gauges calibrated to the master on the evidence be out of calibration to the master.
Purchase, without a currect calibration certificate (2006) means you have no evidence that supports conformity
BradM
Hello! Welcome to the Cove!! Thanks for the great first post!
You make a valid point concerning the NC. I would go with a major if there was no certificate or evidence of calibration whatsoever. To me (and I am not a 17025 expert) this would be a minor, for citing not having a current, valid certification. That is an opinion, and I would defer to those much more proficient in 17025 than I.
I hope to see you contribute and add to our knowledge base here!
K
Ken K
I'm rather surprised Mutituyo sold you a standard with a 7 year old certificate.
I'm not surprised you received a major for using it. You indicate you have used it for a Master Standard, yet you have not indicated you have performed any type of verification on it. Chances are it is within tolerance for your use, but you have to proof except for an out of date calibration certificate.
Personally, I would send it out for calibration and wait for the results. If it is out of tolerance, you have nearly a year of data which will be questionable that I suspect the auditor will take a good look at.
Good luck.
Hershal
Metrologist-Auditor
A corrective action is valid in this case, because the blocks had not been calibrated in 7 years, the original cert showed no evidence of validity, and your own system says two years. I would also have written a non-conformance.
The reality is, you cannot state that you know the value +/- the uncertainty value, of any block in the set. Temperature changes, skin oils, and many other factors will change the values.
As for the major/minor, each AB makes its own decision as to that grade system.....some use it, some do not.....I can't offer any real opinion on that.
Send it to an accredited lab and have it cal'd.....then staying on cycle is much easier.
Hope this helps.
TIMMYS - 2010
In reading all of the posts on the subject I did not find a mention that your procedures were a primary cause for the NC. Whether it is Major or Minor it still is a NC. Your procedures indicate a maximum frequency of 2 years. They did not account for the equipment being longer than 2 years. That should be the NC. An observation or recommendation for the item to be calibrated more frequently is the most an auditor should make without data on degree of usage or conditions of usage or storage.
Summary is that the NC is valid regarding the procedures vs practice. It is not a valid NC because of the long frequency.
Regards,
TIMMYS
bgwiehle
...If it is out of tolerance, you have nearly a year of data which will be questionable that I suspect the auditor will take a good look at.
And that, I suspect, is the reason the finding was deemed a major NC. Our registrars tell us up front that lapses that may lead to non-conforming product being sent to customers are candidates for major NCs.
B.G. Wiehle
M Validity of Old Biocompatibility Data (1999) - Catheter Other Medical Device and Orthopedic Related Topics 3
lanley liao The validity of API Spec Q1 certificate Oil and Gas Industry Standards and Regulations 2
J Validity / outcomes measure for custom made medical device ISO 13485:2016 - Medical Device Quality Management Systems 2
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
M Informational EU – Application of transitional provisions concerning validity of certificates issued in accordance to Directives 90/385/EEC and 93/42/EEC Medical Device and FDA Regulations and Standards News 2
D MDD Certification Validity after May 2020 EU Medical Device Regulations 12
E ISO 13485 7.6 Control of monitoring and measuring equipment - Assess the Validity ISO 13485:2016 - Medical Device Quality Management Systems 4
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
T CE Certification Strategy - Expanding the validity of certificate EU Medical Device Regulations 1
M CE mark validity after Brexit, 29th March '19 EU Medical Device Regulations 6
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
M Is there an Industry Norm for Expiry of Sterilisation Validity? EU Medical Device Regulations 6
somashekar About NB suspension and CE certificate validity EU Medical Device Regulations 12
J EU CE Mark Validity in Non EU countries - Brazil. Taiwan, China, Japan, Mexico EU Medical Device Regulations 1
alonFAI ISO 9001 Quality Manual Validity Period ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
K Validity of Using NCR's to evaluate Performance Nonconformance and Corrective Action 7
G Validity for Thailand Medical Device License and can change Distributor thereafter? Other Medical Device Regulations World-Wide 3
M Validity of Calibration outside the Calibration Range General Measurement Device and Calibration Topics 7
G Validity of Extending Temperature Cycling Various Other Specifications, Standards, and related Requirements 3
D Assessing the Validity of Previous Measuring Results? General Measurement Device and Calibration Topics 8
C Validity of a 5 year ISO 13485 Certificate ISO 13485:2016 - Medical Device Quality Management Systems 3
A Validity of Internal Auditor Certificate/Training Quality Manager and Management Related Issues 14
A Expiration of Validity Date for EN ISO 14971 ISO 14971 - Medical Device Risk Management 8
A Validity of IEC 60601-1 and IEC 60601-2-27 - An intermittent ECG event recorder IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R Validity of Calibration Stickers on Measuring Tapes and Rulers General Measurement Device and Calibration Topics 26
rob73 Length of validity of a self certified Declaration of Conformity (DoC) for CE mark EU Medical Device Regulations 4
S Validity of "Ship to Control" - Do you endorse this approach? Statistical Analysis Tools, Techniques and SPC 16
C Validity Duration of USP / FDA Certificate? Other ISO and International Standards and European Regulations 5
V What is the Validity of IEC Test Reports? What about TRF Compliance? Other Medical Device Regulations World-Wide 2
J Length of Validity - Test Results Various Other Specifications, Standards, and related Requirements 2
K ISO 13485 Registration Certificate Life-Time (Validity Period) ISO 13485:2016 - Medical Device Quality Management Systems 21
Q Uncontrolled Hard Copy (Paper) Document Validity Period Document Control Systems, Procedures, Forms and Templates 12
J AS9100 Rev. B Certificate Validity AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
T CE Mark Validity despite expired 13485 Certification EU Medical Device Regulations 13
D IATF portal for TS 16949 Certificate Validity Check IATF 16949 - Automotive Quality Systems Standard 2
S Validity of Craig Hane's Global Local Index for X-bar R Chart Assessment Capability, Accuracy and Stability - Processes, Machines, etc. 3
P Validity of NB-MEDs (MDD / CE marking) EU Medical Device Regulations 2
E Validity of ISO 9001:2000 Certificates - How to handle supplier ISO 9001 certificates ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
V Criteria used to assess the Validity of a Complaint - Insurance Company Customer Complaints 4
L How to check for validity of registration scope during ISO9001:2008 external audit Quality Manager and Management Related Issues 3
A Validity of ISO 9001 Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Time frame for validity of ISO 9001:2000 Certifications IATF 16949 - Automotive Quality Systems Standard 3
C ISO 9001:2000 Certificate - How to have a 3 year validity instead of 1? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
C Ever come across an ISO 9001 certificate with more than 3 years validity period? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 35
D Validity of ISO 17025 calibration waivers due to proprietary procedures ISO 17025 related Discussions 4
E COA's (Certificate of Analysis) from suppliers - Beginning to question validity Supplier Quality Assurance and other Supplier Issues 4
P CE certificate validity - Signing employee has left the company EU Medical Device Regulations 7
M 7.6 - Validity of the previous measuring results - Is this what validity means? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
T Variables Gage R&R - Minitab v14 - F statistic Calculation - ANOVA - Methods validity Using Minitab Software 2
Marc Customer Surveys - Validity? Fraud? Are they really of any value? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5