Verification Activities - Clinical Medical device manufacturer

M

Mark Smith

#1
My company is in the process of moving to a new location and expecting to resume manufacturing in 2 months time. We are a medical device manufacturer that is currently developing devices and manufacturing for clinical evaluation only!!. We have several designs that we have performed verification testing on however, clinical evaluations intended for DESIGN VALIDATION are ongoing and are scheduled for at least 1 more year.
Since validation FOLLOWS successful verification, the manufacturing process as well as equipment used to manufacture the devices will not be validated until we are certain of the final design (AFTER we move to our new location and prior to distributing any approved product)
When we did verification testing at our current location, we required a number of pass / fail tests on 60 devices and allowed zero failures as evidence of 95% confidence/ 95%Reliability for the design.

I am aware that processes and equipment must be re-validated as the result of a relocation but do I need to repeat the design verification testing of these devices?
If so,
Will I need to repeat the sample size of 60 devices?
If not,
Can I significantly reduce the number of units to something around 10 devices?
 
Elsmar Forum Sponsor
D

Don Winton

#2
<font COLOR="#DF0218"><BLOCKQUOTE>I am aware that processes and equipment must be re-validated as the result of a relocation but do I need to repeat the design verification testing of these devices?</BLOCKQUOTE></font>

I would think that the design verification would not have to be if you can show that the process re-validation was satisfactory. Meanwhile, check out the links below for additional info:

http://www.fda.gov/cdrh/comp/designgd.html
http://www.fda.gov/cdrh/comp/designgd.pdf
http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfTopic/topicindex/topindx.cfm?alpha=d

Regards,

Don
 
Thread starter Similar threads Forum Replies Date
Q Sampling method for Distributor verification activities EU Medical Device Regulations 1
L Documentation of Medical Device Software Verification activities Other US Medical Device Regulations 7
J Design Inputs and Verification Activities CE Marking (Conformité Européene) / CB Scheme 5
P Risk Management Verification Activities for Implementation and Efficiency ISO 14971 - Medical Device Risk Management 3
P How to separate Product Design Review, Verification, and Validation Activities ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
R Delegation of Verification Activities to a Supplier - 7.4.3 Interpretation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
V QA Documentation Activities - Differentiate activities of "Review" and "Verification" Document Control Systems, Procedures, Forms and Templates 3
E AS9100 Interpretation - Delegation of Verification Activities to the Supplier AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 19
Q Determining Sample Size for FDA Verification and Validation Activities Inspection, Prints (Drawings), Testing, Sampling and Related Topics 41
M Using Statistics for corrective action verification activities Statistical Analysis Tools, Techniques and SPC 8
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
C 8.3.4 Verification and Validation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
F Compendial method verification Design and Development of Products and Processes 0
H Verification Process for the existing MDS Product IEC 62304 - Medical Device Software Life Cycle Processes 5
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 5
P Software verification and validation procedure IEC 62304 - Medical Device Software Life Cycle Processes 6
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 6
S How to perform verification of the Statistical Analysis Software? Qualification and Validation (including 21 CFR Part 11) 7
M Sample record for verification performed by importers before placing a device on the market EU Medical Device Regulations 0
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
M ISO14971:2019 - Verification of implementation and effectiveness of risk control ISO 14971 - Medical Device Risk Management 3
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
A Verification protocol example Design and Development of Products and Processes 2
D Design Verification Sample Size vs Repeats Statistical Analysis Tools, Techniques and SPC 9
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
T Laboratory Verification after validation ISO 17025 related Discussions 3
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 6
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
S Documenting Design Verification Test Results (ISO 9001) Design and Development of Products and Processes 1
C Documentation for items used for Design Verification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
W Seeking Guidance Verification Test Strategy for Class B Medical Devices IEC 62304 - Medical Device Software Life Cycle Processes 1
NDesouza Verification of Supplier RCCAs Nonconformance and Corrective Action 5
I IATF 2016 - 8.5.1.4 Verification after shutdown due to COVID19 Process Maps, Process Mapping and Turtle Diagrams 1
M IATF 16949 8.5.1.3 Verification of job set-ups - Do we need secondary check? IATF 16949 - Automotive Quality Systems Standard 7
G Supplier flowdown verification of tests and revisions Supplier Quality Assurance and other Supplier Issues 0
G Devices from IQ, OQ or PQ process to be used for verification, validation and summative? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Bev D Verification and Validation of Measurement Systems Misc. Quality Assurance and Business Systems Related Topics 0
A Our auditor told if we didn't have a patent we would have to do a validation or verification ISO 13485:2016 - Medical Device Quality Management Systems 6
N Design Verification & Process Validation - Statistical sample sizes Design and Development of Products and Processes 2
D Design Verification - Is testing required? Design and Development of Products and Processes 5
J Verification of purchased product / Alexa as a medical device ISO 13485:2016 - Medical Device Quality Management Systems 6
C Machine Verification Plan Manufacturing and Related Processes 2
A What to do when 100 % verification is not 100 % Manufacturing and Related Processes 5

Similar threads

Top Bottom