"Verification of Calibration" - CMM arm

K

Kchnwtch

Hello!
I have a CMM arm that at the moment has a yearly calibration schedule, which means I have to pack it up, send it back to the mfr, and pay hideous fees both for shipping and for the calibration sticker.

I had a (certified) lab rep come to me today and say he can offer "calibration verification," i.e., he can verify whether the arm is in or out of calibration. If it's in, we're fine; if it's out, we have to send it back to the mfr. He can do it onsite while he's calibrating some of our other tools (like an optical comparator) that have to be calibrated on site.

My question: Would a verification like this meet the calibration requirement?

Thanks for your help!
 

normzone

Trusted Information Resource
Re: "Verification of Calibration"?

The answer is dependent upon your situation and regulatory constraints, but in general yes, that is just fine. People get all whupped up about calibration, since there's an adjustment implied in that useage, but in reality what the manufacturer does may be exactly what the alternative supplier you're considering is proposing.

Both of them look at the amount of error in the device and either pronounce it within acceptable limits (that's a can of worms, depending on your useage requirements) or not and adjust it.

Does the manufacturer provide "as received" and "as returned after adjustment, if any" kind of reports, or do they just tell you "it's been calibrated" ? Generally outfits charge more for more data, so sometimes you only get the latter.

Others with more professional statements should show up shortly. But if you have history of the device remaining in cal and not requiring adjustment, then that can be referenced to justify your decision to change the cal provider.
 
T

Tyler C

"Where necessary to ensure valid results, measuring equipment shall be calibrated or verified, or both, at specified intervals, or prior to use, against measurement standards traceable to international or national measurement standards; where no such standard exists, the basis used for calibration or verification shall be recorded."

As long as the lab rep is certified and can prove he is not making "adjustments that would invalidate the measurement result" and is performing against a traceable standard, or a documented standard of verification from your organization, there should be no issue. And, of course, you must maintain these records.

This is quoted from the ISO 9001:2008 standard.
 
J

JoShmo

Re: "Verification of Calibration"?

Hello!
I have a CMM arm that at the moment has a yearly calibration schedule, which means I have to pack it up, send it back to the mfr, and pay hideous fees both for shipping and for the calibration sticker.

I had a (certified) lab rep come to me today and say he can offer "calibration verification," i.e., he can verify whether the arm is in or out of calibration. If it's in, we're fine; if it's out, we have to send it back to the mfr. He can do it onsite while he's calibrating some of our other tools (like an optical comparator) that have to be calibrated on site.

My question: Would a verification like this meet the calibration requirement?

Thanks for your help!

Some wacky statements here, like "calibration verification" sounds like something they made up to sound good. "he can verify if...is in or out of calibration"? He can either calibrate it or verify it and that's all. You decide if it's calibrated, based on the results... It's your choice, really. It sounds like you're using a Faro arm or similar which seems to me to have both options
 

Ninja

Looking for Reality
Trusted Information Resource
Re: "Verification of Calibration"?

We do this on-site ourselves, FWIW. We bought a traceable standard and simply run a verification and get a report printed out. Only if the max error or RMS error are out of acceptable limits do we then have any adjustment made.
Historically we've made an adjustment every ~5yrs or so...the CMM holds tight pretty well in our usage level.

The standard itself is pretty hard to change without breaking it (etched glass) so we verify it at a much longer cycle.

HTH
 
N

ncwalker

Re: "Verification of Calibration"?

Step back and look at your bigger picture.

You told us that your optical comparator gets calibrated on site. Why?

The answer is most likely it's too big and bulky to ship out and have it done. (Also, it has to be shipped BACK. THAT's a scary thought... trusting it to a common carrier after it has been calibrated).

The better question becomes - if your comparator can be calibrated on site, why cannot your Faro arm be?

Answer - it absolutely can. And in reality, it SHOULD be if at all possible.

Again, look at it the other way.... If you purchase a new CMM/Faro arm and they test it on the factory floor, after they ship it, aren't you going to want it tested again at your location to make sure it has not been damaged in transit? Of course you are.

So why then would you not want the same care with your calibration? And if you can verify a new piece of equipment is OK on your floor, why can you not also calibrate current equipment?

We tend to fall into this trap of "do it this way because that's how it is done" with not enough thinking about "what" it actually is we are trying to do. :)
 
K

Kchnwtch

Re: "Verification of Calibration"?

Last year we had to ship it anyway for repairs along with calibration, but yes, I was thinking inside a very small box and making the assumption we needed to send it back again. This should save us a ton of money--thanks everyone for your help!
 
Thread starter Similar threads Forum Replies Date
C IATF 16949:2016 Major NC pertaining to 7.1.5.2.1 Calibration/verification records IATF 16949 - Automotive Quality Systems Standard 15
placebo_master What is the difference between "Fitness for Purpose" and "Calibration/Verification"? General Measurement Device and Calibration Topics 6
G IATF 7.1.5.2.1 Calibration/verification records :Program/software verification IATF 16949 - Automotive Quality Systems Standard 7
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
G Assigning a calibration tolerance - An x-y coordinate machine - Uncertainty as my verification tolerance General Measurement Device and Calibration Topics 4
J Calibration/Verification Records (IATF 16949 7.1.5.2.1) IATF 16949 - Automotive Quality Systems Standard 5
D How to set equipment calibration/verification criteria General Measurement Device and Calibration Topics 1
E IATF 16949 Cl. 7.1.5.2.1 - Calibration and Verification Records Requirements IATF 16949 - Automotive Quality Systems Standard 1
S Calibration and Verification requirement for foreign EASA Part 145 EASA and JAA Aviation Standards and Requirements 7
J IATF 16949 Cl. 7.1.5.2.1 - Gauges (Measuring Jigs) Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
M Clarification on Calibration/Verification Records 7.1.5.2.1d (IATF 16949) General Measurement Device and Calibration Topics 11
B Calibration/Verification Records - IATF 16949 section 7.1.5.2.1 f) IATF 16949 - Automotive Quality Systems Standard 6
P IATF 16949 Cl. 7.1.5.2.1 "Calibration/Verification Records" Interpretation IATF 16949 - Automotive Quality Systems Standard 13
J Pneumatic & Electrical Torque Tool - Calibration/Verification of 'Power Tools' General Measurement Device and Calibration Topics 3
Q Using Caliper A to verify Caliper B - Verification of Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Verification instead of calibration?(ISO 9001:2008) Calibration Frequency (Interval) 10
I Verification or calibration Fluke 8508a with 5720a? TUR not 4:1? Measurement Uncertainty (MU) 1
Q Number of Calibration Verification Points - Calipers, height gages, etc General Measurement Device and Calibration Topics 6
A Calibration/Verification of thermo-hygrometer in storage areas ISO 13485:2016 - Medical Device Quality Management Systems 9
G Is this Calibration or Verification ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
D How Working Standard Calibration result use in Verification of Equipment General Measurement Device and Calibration Topics 7
M Calibration Certification Qualification vs. Verification by Uncertified Staff General Measurement Device and Calibration Topics 6
A Torque Wrench Calibration/Verification General Measurement Device and Calibration Topics 24
G BRC (British Retail Consortium) - Calibration vs. Verification Food Safety - ISO 22000, HACCP (21 CFR 120) 4
G TS-16949 7.6.2 ? Calibration/Verification Records - Assessment of the Impact IATF 16949 - Automotive Quality Systems Standard 5
J Balance Calibration Verification question General Measurement Device and Calibration Topics 9
J Log Book for Analytical Balance Calibration Verification General Measurement Device and Calibration Topics 4
Uriel Alejandro How to Record Evidence for Verification/Calibration of Equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Establishing Ranges for Weights for Calibration Verification of a Micro-Balance Calibration Frequency (Interval) 1
H Analytical Balance Calibration and Verification General Measurement Device and Calibration Topics 10
D Precision Balance Verification and Calibration Process General Measurement Device and Calibration Topics 3
D Equipment Verification and Calibration Report Standard General Measurement Device and Calibration Topics 9
P Record Retention requirements of Calibration and Verification Service Providers ISO 13485:2016 - Medical Device Quality Management Systems 11
K Calibration and Verification Records - TS 16949 Clause 7.6.2 General Measurement Device and Calibration Topics 8
P Is it compulsory to send new measuring equipment for calibration / verification? Calibration Frequency (Interval) 3
D Calibration vs. Verification and Measurement Device Labeling General Measurement Device and Calibration Topics 10
M Basis used for Calibration or Verification of Measuring Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
D For Reference Only or Calibration/Verification - Tape Measures for cutting bar stock ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
M Methods for Verification/Calibration of Video (Optical) Extensometer General Measurement Device and Calibration Topics 2
C Updating our Laboratory Calibration / Verification Procedures IATF 16949 - Automotive Quality Systems Standard 5
R Verification vs. Calibration and what to do with the results of each General Measurement Device and Calibration Topics 3

Similar threads

Top Bottom