Definition Verification vs. Validation - What are the differences?

Randy

Super Moderator
#31
To the best of my knowledge I did not post anything "against" other posters. I did argue some of others opinions, isn't this forum excatly for that?
To a certain degree it is, we just need to ba careful how passionate that agument is.

The 700 pages airfraft spec was a joke- there is really no need to comment on jokes.
You'll find that some folks have no sense of humor, so one has to be careful. (Most of them don't go to the bathroom either:lol:)

You provided pretty good comment, just hang in there and stay involved.
 
Elsmar Forum Sponsor

Kevin Mader

One of THE Original Covers!
Staff member
Admin
#32
More for the mix…

Customer need: chocolate cake

How will this be specified? Milk chocolate, dark or perhaps a blend? If I gather the ingredients, develop a recipe, mix the ingredients and bake the cake, will the customer be happy?

Through verification activities, I can confirm that my inputs are correct, the proportions correctly measured, and blending and bake times correct. There’s still no guarantee that the customer will like my chocolate cake, but I did create a chocolate cake.

Through validation, I need to confirm that I’ve met the consumer’s expectations (requirements). Did I use enough chocolate or too much? Was the cake moist and fluffy? All the bench testing (verification) in the world can’t confirm that I’ve met expectations. I must have the consumer’s (user’s) confirmation. It's the one that counts!!

Marketing specifications, which are usually vague, establish the general customer requirements. These must be refined into more detailed requirements (which can take the shape of product requirements) for engineering. Here, the engineering team establishes design requirements on how they plan to fulfill the customer needs. These might be thought of as to how the engineering team plans to mechanically and electronically establish the product the customer wants. These initial design requirements will undergo an iterative design process until final input requirements can be established. Often times, engineering test is performed during this development phase (sniff testing), but this is not yet design verification. Customer preference testing might also be done at this time to further dial in the product requirements. Design verification will occur later after the final design inputs have been established, generally in the form of a design requirements specification. Design Verification activities trace directly to the Design Inputs. Design Validation is performed to confirm that the customer/end user requirements (intended uses) have been met. As noted in Randy’s post, both activities are important, but separate.

Potential document/activity hierarchy:

· Marketing Specification (top level user needs sometimes called a User Needs Document)
· Product Specification (engineering translation of user needs)
· Design Requirements (initial design inputs, perhaps broken down into subordinate requirements documents for recognized disciplines such as mechanical, electrical, software, etc.)
· Design Requirements Specifications (final design inputs)
· Design Verification (confirmation of final design inputs)
· Design Validation (confirmation of user needs)

It has been my experience that mastery of these two concepts is difficult. I suspect that this is why there is disagreement amongst us Covers, especially for those folks who seldom engage in validation activities. The thing of it is that through continued discussion and debate, we continue to refine our theory of what verification and validation are. For me, I’m always in a curious mode to refine my own theory, so I enjoy these types of exchanges.

Enough for now, so back to the group…

Kevin
 

Helmut Jilling

Auditor / Consultant
#35
Simple but true>
Verification: Did we make the thing right?
Validation: Did we make the right thing?

That is very elegant and simple, but I think it misses the point.

Verification is checks (verifies) whether all the details have been addressed in the design.
Validation checks whether the "thing" actually works. There is a product or prototype at the validation stgae.

That is the fundamental difference.
 
M

madannc

#37
That is very elegant and simple, but I think it misses the point.
In what way?

Verification is checks (verifies) whether all the details have been addressed in the design.
In other words did we make the thing right - against requirements, design inputs. A prototype would be used here for testing usually in a QA controlled environment.

Validation checks whether the "thing" actually works. There is a product or prototype at the validation stage.
In other words did we make the right thing - The functionality that we have verified: meets customer requirements, can be used by the customer in customer environment, does it intergrate with configurations as expected, etc etc

This would not be a prototype unless it was manufacturing representive, as the result of the validation would be part of the data used to determine if it can be released as a product/"thing"
 

Helmut Jilling

Auditor / Consultant
#38
Our company just recently aquired a manufacturing division. This required me to remov our 7.3 exclusion(much to my dismay). I created a controlled form that requires all of the measurements to be checked and recorded after the product has been made. Can this be my validation? If so, how do I show verification?

Thanks in advance!
Susan
Out of curiousity, Susan, with all the opinions and replies this simple question attracted...have you gotten an answer to your original question? Or, is it still not clear?
 
W

w_grunfeld

#39
That is very elegant and simple, but I think it misses the point.

Verification is checks (verifies) whether all the details have been addressed in the design.
Validation checks whether the "thing" actually works. There is a product or prototype at the validation stgae.

That is the fundamental difference.
What is "checks/verifies"? Doesn't it involve any physical measurements just looking at papers/data? To verify that the "thing" actually works is to me design verification, and more often than not, involves a prototype too.
Validation is to check beyond the basic functionality that the "thing" performs as specified over its entire range of operational (use ) conditions .
This argument is never going to end as long as it remains theoretical. There is no definition that fits every type of product.
 

Jim Wynne

Staff member
Admin
#40
That is very elegant and simple, but I think it misses the point.

Verification is checks (verifies) whether all the details have been addressed in the design.
Validation checks whether the "thing" actually works. There is a product or prototype at the validation stgae.

That is the fundamental difference.
Alll of the confusion stems from the fact that validation is a type of verification, and vice versa and that validation isn't necessarily a standalone activity--it's comprised of a series of verifications.

Think in terms of efficacy--the property of being able to perform as intended. In our business, "validation" is confirmation of the efficacy of a design--that design intent has been met. Validation comes about through various types of verification--dimensional, cosmetic, performance, etc. After we're satisfied through verification that design intent has been met, the product/process/whatever has been validated.
 
Thread starter Similar threads Forum Replies Date
D Medical Device Design Verification and Validation Differences 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 16
T Definition Verification and Validation Definitions and Differences Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 9
Q Differences between Sample Sixes for Design Verification vs. Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
I The differences between D&D Review, D&D Verification and D&D Validation? Design and Development of Products and Processes 8
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
T Laboratory Verification after validation ISO 17025 related Discussions 3
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
G Devices from IQ, OQ or PQ process to be used for verification, validation and summative? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
Bev D Verification and Validation of Measurement Systems Misc. Quality Assurance and Business Systems Related Topics 0
A Our auditor told if we didn't have a patent we would have to do a validation or verification ISO 13485:2016 - Medical Device Quality Management Systems 6
N Design Verification & Process Validation - Statistical sample sizes Design and Development of Products and Processes 2
D 510K and Changes to Verification and Validation US Food and Drug Administration (FDA) 2
Ed Panek Label verification and validation US Food and Drug Administration (FDA) 5
V Which batches should or could be considered for design validation and design verification? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
I Medical device Validation/Verification template ISO 13485:2016 - Medical Device Quality Management Systems 1
M Does anyone has a good verification and validation plan template? ISO 13485:2016 - Medical Device Quality Management Systems 3
S Source of QMS templates including templates for design verification/design validation Document Control Systems, Procedures, Forms and Templates 2
Ed Panek Verification & Validation requirements - Patch that adheres to skin ISO 13485:2016 - Medical Device Quality Management Systems 2
M Bringing blending in-house - Verification or full validation? (Disinfectants) ISO 13485:2016 - Medical Device Quality Management Systems 3
A Level of details required for Class IIb Product Verification and Validation CE Marking (Conformité Européene) / CB Scheme 1
S CE Mark Approval Design Verification and Validation Questions CE Marking (Conformité Européene) / CB Scheme 8
S What is the difference between Verification and Validation of Test Reports? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M IEC 62366-1:2015 Usability Verification / Validation Plan example wanted IEC 62366 - Medical Device Usability Engineering 20
shimonv System Verification & Validation in HC Class III Submission Canada Medical Device Regulations 2
G Verification or Validation for Glue Dispenser Units ? Qualification and Validation (including 21 CFR Part 11) 4
T Validation and Verification - Catheter into a tyvek pouch ISO 13485:2016 - Medical Device Quality Management Systems 2
D User Requirements Tracing - Verification and Validation Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Validation before Verification. Why not? ISO 13485:2016 - Medical Device Quality Management Systems 3
H Is design verification and validation required for samples accompanying the quotes ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
E How to write requirements for Medical App Validation and Verification IEC 62366 - Medical Device Usability Engineering 7
Chennaiite Cause Analysis for Verification/Validation Failures - Seeking Opinions Problem Solving, Root Cause Fault and Failure Analysis 3
F Usability Objectives and Verification & Validation Plan IEC 62366 - Medical Device Usability Engineering 6
S Verification and Validation of Post Market Design Change Design and Development of Products and Processes 2
M Validation/Verification Trace matrix - examples or recipe? Document Control Systems, Procedures, Forms and Templates 1
T Class II Software Device 510k V&V (Verification and Validation) Criteria and Results 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
V Utility (power production) Company - Mitigating risks in Verification and Validation. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A Should the FSMS have a Documented Generic Procedure for Verification and Validation Food Safety - ISO 22000, HACCP (21 CFR 120) 2
P What is Validation, Verification and Improvement of FSMS in ISO 22000 Clause 8? Food Safety - ISO 22000, HACCP (21 CFR 120) 2
P How to separate Product Design Review, Verification, and Validation Activities ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
T Medical Device Software Verification & Validation for Client Specific Configurations Design and Development of Products and Processes 2
Q Verification & Validation Project Time Line Qualification and Validation (including 21 CFR Part 11) 5
M Design and Development Outputs Approval, Verification and Validation IATF 16949 - Automotive Quality Systems Standard 4
G Legacy Product - Do we need follow QSR 820? Design Verification and Validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 12
S Design and Development - Review vs. Verification vs. Validation Design and Development of Products and Processes 15
G Sample Size for Design Verification and Validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
A Does the V&V (Verification and Validation) belong to Bench Testing for 510k? Qualification and Validation (including 21 CFR Part 11) 2
K Verification, Validation or Checking? Definition of the word "Checked" Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
B Is Ship Test considered Design Verification or Design Validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
D Design Input/Output and Design V&V (Verification and Validation) Interpretations 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 14