Verify Software System Testing: Verify that results meet the Pass/Fail Criteria
sope01
I'm new to IEC 62304 and medical device software development.
Chapter 5.7.4 "Verify SOFTWARE SYSTEM testing" in the 62304 standard states:
"The manufacturer shall verify that
a) The VERIFICATION strategies and the test procedures ...
b) ...
c) ...
d) test results meet the required pass/fail criteria
My question is: How to interpret this last item ?
Does this mean that results have to be reviewed AFTER tests were executed ?
Or would a review of test case's steps and expected results BEFORE execution be sufficient
Marcelo
Hello sope01 and welcome to the Cove!
First you have to define what and why you are testing. Then you need to define how to you verify that the test is passed (pass/fail criteria).
Then you perform test. Then you analyze the results and concludes if the test is passed or not.
This in fact is not only applicable to software, but to anything that needs to be verified.
K Verify Software Architecture - supporting interfaces between items IEC 62304 - Medical Device Software Life Cycle Processes 2
R IEC 62304 - 5.2.9 Verify Software Requirement IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 62304 Section 5.3.6 Verify Software Architecture IEC 62304 - Medical Device Software Life Cycle Processes 4
D How to verify a process that uses software to accomplish certain tasks? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
J Using current gauge to verify masters General Measurement Device and Calibration Topics 1
T What does "fully verify" mean exactly in process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R To Close the RCCA or Verify Effectivenessof RCCA first? Nonconformance and Corrective Action 14
Q Gage Pin Sets? Calibrate/Verify them all or just a sample size? General Measurement Device and Calibration Topics 22
D Verify Audit Trail of SaaS system Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 5
qualprod D5 of 8D clarification, how to verify root cause Problem Solving, Root Cause Fault and Failure Analysis 26
C Where to draw the line for "sufficient evidence" to verify safety/performance of a device? CE Marking (Conformité Européene) / CB Scheme 2
M How to verify profile gages provided by customer General Measurement Device and Calibration Topics 5
E Layouts : Can you "verify" things that are not visible (such as loctite present, internal dimensions) in a layout ? APQP and PPAP 2
K How to verify the Storage and Transport environmental condition for Medical Device Miscellaneous Environmental Standards and EMS Related Discussions 3
J How to verify the actual capacity of suppliers Supplier Quality Assurance and other Supplier Issues 3
V Regrinding Material - Verify whether a product contains any regrinding material Manufacturing and Related Processes 13
Q Using Caliper A to verify Caliper B - Verification of Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Tie-wrap pull test to verify contact retention Manufacturing and Related Processes 6
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
A How to Verify Supplier I-9 Forms for ITAR Requirements Other ISO and International Standards and European Regulations 2
O Creating a Tool to Track & Verify Mistake Proofing Devices Document Control Systems, Procedures, Forms and Templates 5
N What is the easiest way to verify we have the correct customer supplied materials? Misc. Quality Assurance and Business Systems Related Topics 5
B Verify GB/T 16886 is the same as ISO 10993 US Food and Drug Administration (FDA) 5
A How to verify and validate the effectiveness of CAPA with regards to employees ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R LKV Value: Can not verify Reliasoft Case Study Reliability Analysis - Predictions, Testing and Standards 2
L Gauge to test and verify a Set Square Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
D What to verify on Test Setup for Environmental Testing General Measurement Device and Calibration Topics 7
A When to verify and when to validate Design Changes 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Since NIST # can not be used to verify Traceability - What an Auditor should do AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
C I need to verify the surface profile of a part - Sine plates General Measurement Device and Calibration Topics 4
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
optomist1 CMM Shortcomings - GD&T - Inability of CMMs to verify the MMC situation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
J How do I continually Verify Compliance to my Processes? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R A way to verify 7075 T0 Aluminum has been successfully Heat Treated to T6 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
Ron Rompen Recommended Rockwell 30T Scale Hardness Tester to verify Incoming Product General Measurement Device and Calibration Topics 1
X Cannot Verify until Customer uses Product - Clause 7.5.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
C Surface Plate Calibration - How to verify the flatness of a granite surface plate General Measurement Device and Calibration Topics 1
E How Do You VERIFY Customer Property? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
R Verify Accuracy for Validation of CNC Starrett Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
L Verify, Inc. Source Inspection experience? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
H Definition Inspect, Review, Verify and Witness - What are the definitions and differences? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 5
B How to inspect Mylars to verify dimensions, grid lines and grid points Inspection, Prints (Drawings), Testing, Sampling and Related Topics 17
D How do I verify the Cold Room temperature in Restaurant Kitchen Food Safety - ISO 22000, HACCP (21 CFR 120) 7
C Bridgeport Mill as a Gage - Verify gage dimensions using a Bridgeport Mill Measurement Uncertainty (MU) 4
A Test Method - DIN ISO 1133 for melt flow needed to verify supplier results Manufacturing and Related Processes 1
F How would you Verify Repeatability and Reproducibility of Measurement Systems? Quality Tools, Improvement and Analysis 5
S How to Calibrate or Verify Screw Pitch Gages General Measurement Device and Calibration Topics 16
J Work Instruction to Calibrate or Verify a Combination Square General Measurement Device and Calibration Topics 15
H Hygrometer Calibration - Best way to verify accuracy of a 'reference' Hygrometer? General Measurement Device and Calibration Topics 11