We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Question

A

amyjpowers

#1
Hi, quality folks,
I work for a design and development/R&D facility that develops a natural medicinal salve for treatment of wound care. It's considered a class II medical device. We are an organization of less than 15. I came on board 2 months ago and was tasked with setting up a QMS where no formal systems exist.

As we continue to develop our QMS in preparation for ISO 13485 registration in the future I've run into a couple of snags....or maybe I'm over thinking.....like that's NEVER happened! :rolleyes:

We design and develop. The output of our design system is manufacturing specifications that will be sent to our (pending) contract manufacturing organization (CMO). We'll have a Feedback System in place and our lab will be involved with testing/approvals before final manufacturing can take place. Since all of our manufacturing is outsourced, and without "shirking" any responsibility, is it necessary for us to maintain systems for:

  • Control of Production and Service Provision
  • Monitoring and Measuring of Product
  • Control of NC Product
It appears to me that the CMO will have it's own requirements regarding these systems. I'm wondering if a "blurb" of some kind in our Quality Manual will satisfy any external parties.

All thoughts are greatly appreciated.
 
Elsmar Forum Sponsor

sagai

Quite Involved in Discussions
#2
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

13485 certification alone makes not too much sense, for EU you have to have a CE marking approval (in US the 51ok market clearance) next to it in order to be able to sell your classII device.
Where are you planning to market this device?
Is this device will be marketed under the name of your company?
Regards
Sz.
 

DannyK

Trusted Information Resource
#3
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

Based on your post I am assuming that the product will be labelled with your company name.

If that is the case, you have to indicate in your quality system the way you control the outsourced processes.
Do you perform audits at your CMO?
Does your CMO have ISO 13485 registration?
If there are customer returns, which company will be responsible?
 
M

MIREGMGR

#4
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

21 CFR 820 and ISO 13485 have somewhat similar requirements for each of the QS elements you listed. It would be difficult to not have such functionality in your QS and still be able to pass an FDA inspection that reviewed your control of product realization, given that you are the Manufacturer even if you utilize a contract manufacturer to do the actual formulation and packaging.
 

somashekar

Staff member
Super Moderator
#5
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

Hi, quality folks,
I work for a design and development/R&D facility that develops a natural medicinal salve for treatment of wound care. It's considered a class II medical device. We are an organization of less than 15. I came on board 2 months ago and was tasked with setting up a QMS where no formal systems exist.

As we continue to develop our QMS in preparation for ISO 13485 registration in the future I've run into a couple of snags....or maybe I'm over thinking.....like that's NEVER happened! :rolleyes:

We design and develop. The output of our design system is manufacturing specifications that will be sent to our (pending) contract manufacturing organization (CMO). We'll have a Feedback System in place and our lab will be involved with testing/approvals before final manufacturing can take place. Since all of our manufacturing is outsourced, and without "shirking" any responsibility, is it necessary for us to maintain systems for:

  • Control of Production and Service Provision
  • Monitoring and Measuring of Product
  • Control of NC Product
It appears to me that the CMO will have it's own requirements regarding these systems. I'm wondering if a "blurb" of some kind in our Quality Manual will satisfy any external parties.

All thoughts are greatly appreciated.
This is a nice business thinking and a very good one for application of control over outsourced processes. Since the manufacturing and associated activities as you have indicated are all of clause 7 and 8 (minus design and development) it will do good to make a very good quality agreement with your future CMO, where you will list all the details of your controls exercised, your responsibilities and liabilities and the CMO responsibilities and liabilities. You may also consider putting in the quality agreement his acceptance to be audited by you and your CB / NB towards your ISO 13485 or CE requirements. On a broader level you need to carefully draft a quality agreement and discuss the same internally, with your CB/NB and with your future CMO before signoff and implementation.
 
A

amyjpowers

#6
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

Sagai,
We are setting up our systems to be compliant with ISO 13485 and the FDA QSR. My focus is the controlled systems. Another co-worker is on board and investigating the CE marking, etc. We are well aware of the FDA 510K market acceptance. It's being built into our Design and Development system. This device will be marketed under the company name and we intend to go world wide (long term).

Danny and Somashekar....GREAT considerations! Thanks! I believe I have my answer!

Gosh....I love the Cove!! :thanks:
 

Roland Cooke

Quite Involved in Discussions
#7
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

Per Clause 1.2, you can only deem elements from Clause 7 as non-applicable, so you will definitely need systems for monitoring/measuring and also for control of NCP, even if this is restricted to ensuring those activities are properly established at your subcontractor, and perhaps dealing with returned product (complaint investigation etc).
 
Last edited:

MRWardell

Involved In Discussions
#8
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

I know this thread is a little old, but seemed a good place to start. We are CM for a small company that developed a Class II device which received 510(K) approval. This company is the manufacturer of record, although I know we have shared responsibility for complying with QSR. My question - does our customer (the developer of the device) require a QMS per 21 CFR 820? Thanks for your thoughts.
 

Ronen E

Problem Solver
Staff member
Moderator
#9
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

I know this thread is a little old, but seemed a good place to start. We are CM for a small company that developed a Class II device which received 510(K) approval. This company is the manufacturer of record, although I know we have shared responsibility for complying with QSR. My question - does our customer (the developer of the device) require a QMS per 21 CFR 820? Thanks for your thoughts.
Hi,

Unless your contract explicitly states otherwise, all regulatory responsibilities rest with your customer (assuming you are not involved in distribution activities). If their device is not GMP exempt then they would have to meet part 820 requirements applicable to specification developers. You have no direct responsibility to comply with part 820, though such compliance may be beneficial to all involved parties, and anyway your customer will probably not be able to meet their regulatory obligations without your support.

Cheers,
Ronen.
 

MRWardell

Involved In Discussions
#10
Re: We Design and Develop - All of our Manufacturing is Outsourced - ISO 13485 Questi

Hi,

Unless your contract explicitly states otherwise, all regulatory responsibilities rest with your customer (assuming you are not involved in distribution activities). If their device is not GMP exempt then they would have to meet part 820 requirements applicable to specification developers. You have no direct responsibility to comply with part 820, though such compliance may be beneficial to all involved parties, and anyway your customer will probably not be able to meet their regulatory obligations without your support.

Cheers,
Ronen.
So, unless the contract says otherwise, we have no obligation to manufacture the device in accordance with the GMP or 21 CFR 820? That is the way I read the regulations, but that seems so bizarre to me that as the contract manufacturer we would not have some obligation or need to have the records necessary to pass an FDA audit, should it occur. Thanks for your input.
 
Thread starter Similar threads Forum Replies Date
M ISO 9001 Requirements for associated service - Design and develop software Design and Development of Products and Processes 7
S Product Development Program - Design, develop, approve and manufacture new products Design and Development of Products and Processes 6
A What is the difference between Design Process, Process Design and Design Control? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
S Documenting Design Verification Test Results (ISO 9001) Design and Development of Products and Processes 1
DuncanGibbons Understanding the applicability of Design of Experiments to the IQ OQ PQ qualification approach Reliability Analysis - Predictions, Testing and Standards 0
S Requirement to Conduct New Shelf-life Testing? (re-do testing for design change) EU Medical Device Regulations 3
A Sample Agreement available for Outsourcing Medical Device Design activity? ISO 13485:2016 - Medical Device Quality Management Systems 1
DuncanGibbons How is the arrangement between Design and Production organisation envisaged? EASA and JAA Aviation Standards and Requirements 4
L Design & Development of a SERVICE Service Industry Specific Topics 13
C Documentation for items used for Design Verification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
A AS9102B - 3.6 Design Characteristics and form 3 AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 3
P Design FMEA - Detection Rating criteria ISO 14971 - Medical Device Risk Management 3
U Medical Device Design finalization testing ISO 13485:2016 - Medical Device Quality Management Systems 2
S MDR Delay - MDD design Change? (before new MDR DOA) EU Medical Device Regulations 8
J Iterative design and production for custom made products ISO 13485:2016 - Medical Device Quality Management Systems 3
T Design Input detail & specificity ISO 13485:2016 - Medical Device Quality Management Systems 4
J Design file for pre-existing products - Inputs and Outputs ISO 13485:2016 - Medical Device Quality Management Systems 5
D Design Transfer Template capturing Customer Specific Requirements Other Medical Device Related Standards 3
T Design Control Procedures later in the Development Process ISO 13485:2016 - Medical Device Quality Management Systems 6
M Looking for a Presentation on Design for Excellence (DfX) Manufacturing and Related Processes 2
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
R Design and Manufacture Guidelines for Surface Mount Technology Misc. Quality Assurance and Business Systems Related Topics 8
optomist1 Design Exclusion, but now we might have an outsourced Product Design ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Q Relabeler for patent expired product - design control responsibilities? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
B Supplier of design and manufacture process ISO 13485:2016 - Medical Device Quality Management Systems 10
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18
A Design process goal for ISO 9001 Manufacturing and Related Processes 23
Z Definitive definition of design? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A UDI and Design Controls - Labeling change via the Design Control process 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
P Design FMEA for Industrial Machinery FMEA and Control Plans 3
M Design Development MDR Design and Development of Products and Processes 0
L Contracted Manufacture Company wanting to be able to design and manufacture own product. 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
C Essential Design Output(s) and Design transfer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
O How can I justify excluding the R&D group and the design and development clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
N Design Verification & Process Validation - Statistical sample sizes Design and Development of Products and Processes 2
D Design Verification - Is testing required? Design and Development of Products and Processes 5
D DHF Responsibility after design handover Document Control Systems, Procedures, Forms and Templates 4
K Completing Design Controls after 510(k) Clearance Other US Medical Device Regulations 11
C Design and implementation of process audits as defined within IATF 16949 IATF 16949 - Automotive Quality Systems Standard 2
R Design Verification Documentation ISO 13485:2016 - Medical Device Quality Management Systems 19
C Design Control - Ways of capturing inputs from manufacturing Other Medical Device Related Standards 5
R Design verification for interim design outputs - sampling rationale ISO 13485:2016 - Medical Device Quality Management Systems 2
C IATF 16949 8.3 Exclusion - Manufacturing process design and development IATF 16949 - Automotive Quality Systems Standard 5
R No design history file or device master record ISO 13485:2016 - Medical Device Quality Management Systems 5
R Suggestions for putting together a DHF (Design History File) ISO 13485:2016 - Medical Device Quality Management Systems 3
M One clinical design: Russia and Europe Other Medical Device and Orthopedic Related Topics 0