Weight Calibration Uncertainty

D

Debbie

#1
Our calibration lab has initiated uncertainty measurements of weight standards.
They are responsible for calibrating around 60 different weight sets (around 25 weights per weight set) and around 200 individual weights.

They just showed me what they were doing to calculate uncertainty -- it looked close and might be OK -- but I thought I'd check to see what more experienced people thought.

They calculated uncertainty based on the uncertainty of the lab standard weight and the standard deviation of three reps of ABBA sequences (A= standard ; B= unknown). THey calculated a standard deviation based on six differences (A's - the adjacent B's).

Is this OK -- or would it be better to pool the standard deviation from like weight calibrations? THey have no historical data.
 
Elsmar Forum Sponsor
D

DICKIE

#2
Debbie,
How did they "calculate" the uncertainty? How are the measurements being performed? What is the resolution of any instruments? Are instruments analog or digital? What are the uncertainties of all of the components of the measuring system?
 
D

Debbie

#3
Thank you very much for your response. The balance that is used is digital. It is considered to be the "Standard" balance for weight calibrations. For a given weight calibration, the same balance is always used, i.e., for all 1 gram weights, a particular Mettler or Sartorius balance is used.

Calculations at this point are based on three repeats of six comparisons of working standard to lab standard. Here's an example of the data


1g
A 0.99997
B 1.00015
B 1.00021
A 0.99998
A 0.99995
B 1.00020
B 1.00019
A 0.99987
A 0.99993
B 1.00019
B 1.00014
A 0.99991
A&B DIFF 0.00024


MEAN A 0.99994
MEAN B 1.00018

Uncertainty was calculated by determining the standard deviation of the lab standard's weight (not the uncertainty of the differences between the lab standard and the reference) and combining that with the standard deviation of the reference.

I see referenced in EA-4/02 that the difference within each ABBA grouping should be considered a single observation. If I directed the lab in that direction, there would only be three observations. Since they calibrate many similar sets of weights to the same reference, can I pool the standard deviations from similar comparisons (e.g., all similar 1 gram weight references)?

Sorry this is so long, I really appreciate comments and responses that I have gotten.
 
D

DICKIE

#4
Debbie,
I'll give you some of the details of a 1 gram analysis I just did.

Subject unit is a class F 1 gram mass.

Measurement parameter is 1 gram reference mass calibrated with a stated uncertainty of 0.01832 mg from the calibration source.

five readings on the class F mass yield standard deviation of .005 mg at 95.0 confidence level.

next I use uncertainty=(xmass*(1-(air density/ref mass density)))/(1-(air density/mass density\))-.001

combining these values gives uncertainty of .018mg with infinite degrees of freedom.

root sum square of the component uncertainties give combined uncertainty of .0259mg at 95% confidience level and 348 degrees of freedom.

Hope this helps

Greg
 
T

tomvehoski

#5
Try downloading the freeware proficiency calculator from http://www.proficiency.org. It is in the downloads section. There are examples for mass and scale calibration, along with many others. It was very useful on a recent 17025 audit for a scale calibration company I consulted for.

Hope this helps,

Tom
 
Last edited by a moderator:
D

Debbie

#6
Tom,

I downloaded the software -- looks real good.
I have a few things to figure out -- (I'm referring to the UC mass 10 kg nominal.unc file. For example 'comparator linearity'.

This example and the replay from Greg have helped me a lot. We were initially going to ignore boyancy -- but I can see that I need to take that into account.

I'm sure we will be using this software. I like that examples are included.

Debbie
 
D

Debbie

#7
Greg,

Thanks for your help. I've looked up a lot of information on boyancy -- I see that I can't ignore it. I'll have to go down to the lab and check out conditions to figure out the air density. I already know temperature and humidity.

I really appreciate you taking the time to review the information in my post and reply.

Thanks again.

Debbie
 
T

tomvehoski

#8
You might be able to get more help from your state department of weights and measures. In Michigan, the Dept. of Agriculture certifies most of the calibration standards in use by scale companies. They might be more willing to help than a commercial lab. I found a link to all of the different state departments here:

*** Dead Link Removed ***

You can try http://consumeraction.gov/
 

howste

Thaumaturge
Trusted Information Resource
#9
Try downloading the freeware proficiency calculator from http://www.proficiency.org. It is in the downloads section. There are examples for mass and scale calibration, along with many others. It was very useful on a recent 17025 audit for a scale calibration company I consulted for.

Hope this helps,

Tom
Tom, is there any chance you still have some of the freeware downloads from the site? They no longer have a download section on their website.
 
Thread starter Similar threads Forum Replies Date
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
A Calibration of Disposable Weight Scales General Measurement Device and Calibration Topics 2
smryan Weight Calibration Standards - External Calibration Frequency General Measurement Device and Calibration Topics 11
K Calibration Standards - Master weight having more tolerance than our gram scale General Measurement Device and Calibration Topics 3
K Weight Calibration to meet the Z17025 Standard ISO 17025 related Discussions 1
F Calibration Reference Standard for Mass Weight Calibration General Measurement Device and Calibration Topics 2
J Classes for Hydraulic Calibration - Dead Weight Tester or Automated Controller General Measurement Device and Calibration Topics 1
S Selecting weights for daily Calibration Check - Using just one weight? General Measurement Device and Calibration Topics 10
K Recording Weight Values on Calibration Certificates General Measurement Device and Calibration Topics 13
C Verification vs. Calibration of Dial Weight Dynamometer General Measurement Device and Calibration Topics 4
G Count By Weight Scale - Calibration within specification, but not linear General Measurement Device and Calibration Topics 8
T Class "F" Weight "Certificate of Traceability" Requirements - Calibration General Measurement Device and Calibration Topics 7
B Weight class required for scale calibration General Measurement Device and Calibration Topics 2
J UPS owned shipping carton weight scales - Include this in my Calibration system? General Measurement Device and Calibration Topics 9
S Dead Weight Tester Calibration General Measurement Device and Calibration Topics 2
C Help and Advice Requested: Calibration of Weight Sets General Measurement Device and Calibration Topics 3
S MSA clause 7.1.5.1.1 - Statistical studies on counting weight scale Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
S CQI-23 - Molding System Assessment - Control of part weight IATF 16949 - Automotive Quality Systems Standard 5
M Informational FDA discussion paper – Consideration of Benefit-Risk Approaches for Weight-Loss Devices Medical Device and FDA Regulations and Standards News 0
D Measuring fine tubing (< 0.100") "Wall by weight" General Measurement Device and Calibration Topics 11
Q FDA 3500A Reports - Trip and Fall - Weight scale with a fold out platform Other US Medical Device Regulations 1
P Calculating a weight for a machine for the CE label CE Marking (Conformité Européene) / CB Scheme 0
M Informational US FDA final guidance Characterization of Ultrahigh Molecular Weight Polyethylene (UHMWPE) Used in Orthopedic Devices Medical Device and FDA Regulations and Standards News 0
Y How to fix weight tolerance for plastic injection molded part? Does it vary with material groups? Manufacturing and Related Processes 1
M Perform a coating weight test within 3 hours of processing the parts Manufacturing and Related Processes 2
A IMDS - Changing the Locked Weight Value RoHS, REACH, ELV, IMDS and Restricted Substances 1
N PFMEA Potential Failure Mode for Weight FMEA and Control Plans 5
S Powder Weight Proportion Measurement Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
J Calculating impact force, expressed in lbs, of a 5 lb weight dropped 48 inches Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R Weight Sets for Checking Table Balances Quality Assurance and Compliance Software Tools and Solutions 5
T Calculation Question - Calculate the runner weight for a part Manufacturing and Related Processes 2
T Calculating Plating/Coating Weight for IMDS RoHS, REACH, ELV, IMDS and Restricted Substances 3
H FDA Classification of a Body Weight 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C Verifying In House Laboratory Weight Sets General Measurement Device and Calibration Topics 2
F Injection Molding Weight Control Systems Document Control Systems, Procedures, Forms and Templates 2
P Pressure Corrections and associated MU for Dead Weight Testers Calibration and Metrology Software and Hardware 1
B SPC on Weight of Plastic Injection Molded Parts - What variables affect my process Statistical Analysis Tools, Techniques and SPC 2
J Galvanizing and ASTM A90 - Coating Weight Test Compliance Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
I Connection of Secondary Circuits to Medical Weight Scale Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Stijloor Best way for girls to lose weight.... Funny Stuff - Jokes and Humour 2
T Required ISO 17025 Traceability - The specific ID/SS number of the test weight ISO 17025 related Discussions 6
V What has more weight: Master Print or Part Submission Warrant? APQP and PPAP 4
S What is the 10d in the owner's manual for a balance's minimum weight? General Measurement Device and Calibration Topics 5
T Weight Load Testing Methods for Surgical Operating Tables ISO 13485:2016 - Medical Device Quality Management Systems 8
R Timber Harvest Weight - Combining Error between Model & Sample in JPM Design and Development of Products and Processes 2
M GRI (Global Reporting Initiative) En 1 Materials used by Weight or Volume Sustainability, Green Initiatives and Ecology 1
G IMDS (International Material Data System) - Is there a minimum weight? RoHS, REACH, ELV, IMDS and Restricted Substances 17
M Calibrating a Weight Scale in a Laboratory Calibration Frequency (Interval) 4
Randy Loss of gravitational attraction (Weight loss for you PhD's) Coffee Break and Water Cooler Discussions 29

Similar threads

Top Bottom