What are the limits on manufacturing changes and substantial changes?
Ed Panek
QA RA Small Med Dev Company
Additional line of production under the same QMS and same building?
Different buildings on the same campus?
Differing addresses on the same campus?
planB
Ed,
I would say, all 3 examples you are mentioning are _not_ beyond limits re. manufacturing changes. Discussing the intended change with your notified body certainly helps.
planB
Ideally, there are supplier contract and specifications in place that require the contract manufacturer to notify and seek approval by the legal manufacturer prior to implementing the requested change.
Ed Panek
Found this MDD for NBOG
Changes that are substantial
b) of changes to EC-approved quality systems (MDD Annex II.3, V, VI; IVDD Annexes IV and
VII respectively):
· Change of a critical material supplier; critical component/process subcontractor
· A change in the manufacturing process whereby a new technology is introduced
· Buying a product design from a subcontractor, that falls within the manufacturers own
approved product range, but which was not designed under the manufacturer´s own
design control system
· Making substantial changes to the sterilization process/cycle that would necessitate a
full new validation (e.g. going form EtOH sterilization to gamma sterilization)
· Changing the sterilization contractor
· Building a new clean room or changing an existing cleanroom’s classification
· Making substantial changes to the environmental monitoring program or then environmental
control systems
· Moving a production line from in-house to an outside facility (manufacturer´s own or a
contractor); introducing a production line into manufacturer´s own facility from an outside
facility
planB
Ed,
if you are looking for a decision tree for assessing whether a change has to be assessed as being significant (and thus, has to be submitted to your notified body), this is quite complex these days for MDD/AIMD devices due to the MDR transition period until May 26 2021.
Depending on the nature of your change NBOG 2014-3 or MDCG 2020-3 may apply.
Hope this helps,
RobertvanBoxtel
If the identified changes are done with regard to MDD/AIMDD certified products, the referenced NBOG 2014-3 should be used to determine notification and NB approval requirements.
However, per Date of Application (DoA) - May 26, 2021 - everything changes. The Commission has written a guidance document how Article 120.3 should be applied by manufacturers and NB's for products falling under not yet expired MDD/AIMDD certificates.
Additionally, for MDR compliant products, the requirements on change notifications in that Regulation will become relevant.
And always, please check your contract with your NB. There you might find some additional notification requirements, on top of mentioned documents.
D Different limits for new and worn gauges General Measurement Device and Calibration Topics 2
B Setting flexible packaging specification limits ISO 14971 - Medical Device Risk Management 4
S Laboratory Environmental Conditions Outside of limits ISO 17025 related Discussions 4
J How to Set Microbial Limits in Food Supplements Food Safety - ISO 22000, HACCP (21 CFR 120) 3
M Targets / limits for PPM/ Scrap Manufacturing and Related Processes 2
S Tolerance limits for micro balance General Measurement Device and Calibration Topics 1
samer Acceptable limits for Spills - Tracking Hydraulic Spills ISO 14001:2015 Specific Discussions 3
P Dropping ISO 9001 limits the scope of the ISO 13485 audit? ISO 13485:2016 - Medical Device Quality Management Systems 6
N Apply control limits to a non-normal distribution Statistical Analysis Tools, Techniques and SPC 13
D IEC 60601-1 - Performance limits for essential performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 - Limits of agreement as Essential Performance IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Calibration certificates avowing accuracy beyond the uncertainty limits? Measurement Uncertainty (MU) 7
L Gage R&R >10% - Decrease Machine Limits? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
G I-Chart and X bar Chart Control limits Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
N Converting ETO Residual Limits from PPM to mg ISO 13485:2016 - Medical Device Quality Management Systems 6
S GR&R using 10 Parts with different Specification Limits Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P Setting Acceptance Quality Limits (AQL) for New Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
V Correlating Maximum 'Range' of Values for a given "RSD" (Control Limits) Statistical Analysis Tools, Techniques and SPC 3
H Control limits for varying subgroup sizes Statistical Analysis Tools, Techniques and SPC 1
G Procedure to define Signing Authority for Procurement Limits ISO 13485:2016 - Medical Device Quality Management Systems 2
T What to do if parameters are found to be Outside of Control Limits Document Control Systems, Procedures, Forms and Templates 5
M Roundness & the Limits of Size - GD&T Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
S What is the Legal Limits of Fungi in Poultry Feed Food Safety - ISO 22000, HACCP (21 CFR 120) 2
E How to use the Standard Error and how to set the error limits? Reliability Analysis - Predictions, Testing and Standards 2
L How do I set a limits to cfu in an ISO class 7 Clean Room? ISO 13485:2016 - Medical Device Quality Management Systems 2
C Sewage Discharge Consents - Proving within Limits ISO 14001:2015 Specific Discussions 7
L Registrar Audit Scope and Limits ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
B Process Capability - Changing Limits to Improve Cpk or find Root Causes? Capability, Accuracy and Stability - Processes, Machines, etc. 6
K Historical Data to establish Control Limits Manufacturing and Related Processes 4
B Cleanroom Control Limits: Cumulative or individual? Statistical Analysis Tools, Techniques and SPC 11
S Product Bio-burden Limits and Sublethal Cycle Requirements ISO 13485:2016 - Medical Device Quality Management Systems 4
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
D Calculating Lower and Upper 3 Sigma Control Limits Statistical Analysis Tools, Techniques and SPC 2
V Setting Specification Limits for UOD (Uniformity) for Live Bacteria or Spores Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
B How to set Cleanroom Alert and Action Limits as part of Microbiological Evaluation Quality Assurance and Compliance Software Tools and Solutions 8
K Advice on Calculating Control Chart Control Limits Statistical Analysis Tools, Techniques and SPC 13
B Non Acceptable % Tolerance Result - Decrease Sorting Limits Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
Q Setting Specification Limits Quality Tools, Improvement and Analysis 11
R ISO 14644-1 maximum particulate concentration limits concern Other ISO and International Standards and European Regulations 2
K Determining Control Limits from a desired Cpk Using Minitab Software 19
M Critical Limits for CCP in a food storage warehouse Food Safety - ISO 22000, HACCP (21 CFR 120) 4
T Calculation of Unified Thread Limits - Non-standard thread 1 ? 36 UNS-2A Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
P Critical Limits for Air Blower as CCP (Critical Control Point) Food Safety - ISO 22000, HACCP (21 CFR 120) 6
I Shewhart Constants vs Central Limit Theorem in calculating Control Limits Statistical Analysis Tools, Techniques and SPC 18
S Is a Stable Process (within Control Limits) required for Calculating Pp, Ppk? Capability, Accuracy and Stability - Processes, Machines, etc. 6
H Johnson Transformation in Minitab - How upper & lower spec limits are recalculated Using Minitab Software 1
D Are Four Sigma Control Limits based on Shewhart's work Statistical Analysis Tools, Techniques and SPC 3
H Specification Limits outside of Johnson Transformation Capability, Accuracy and Stability - Processes, Machines, etc. 15
W Touch Current Measurement and Limits of 42.4V peak or 60Vdc IEC 60601 - Medical Electrical Equipment Safety Standards Series 4