What are the obligatory medical device reporting timelines for AIMD devices ?

Aphel

Involved In Discussions
Dear all,

Can someone explain to me - what are the obligatory timelines in EU for reporting incidents to authorities in case of active implant devices?

Is it similiar to class 1 and 2 devices?
2 days in case of public danger?
10 days in case of user death?
30 days for other mandatory reportable events?

Thanks a lot in advance!

BR Aphel
 

pkost

Trusted Information Resource
As per the MEDDEV (which is not a legal requirement but one which you would have difficulty justifying your deviation from) states that all incidents must be reported immediately where immediately is defined as:

without any delay that could not be justified

The timescales that you have provided are the upper limits i.e. it is not justifiable to wait longer than this in any circumstance
 
Y

yana prus

Essential Requirement 2 states: ?The solutions adopted by the manufacturer for the design and construction of the devices must conform to safety principles, taking account of the generally acknowledged state-of-the-art.? This means documents and standards that are safety related, relevant to the intended purpose and considered state-of-the-art must be taken into account and solutions adopted that are identical or conform to the same safety principles. MEDDEV guidances are considered as the state-of-the-art standards with regards to the essential requirements of Directive for Active Implantable Medical Devices (AIMD), 90/385/EEC, therefore MEDDEV 2.12-1 rev 8 Vigilance guidance must be followed for the determination of the timescale for the initial reporting.

TIMESCALE FOR THE INITIAL REPORTING OF AN INCIDENT
Upon becoming aware that an event has occurred and that one of its devices may have caused or contributed to that event, the MEDICAL DEVICE MANUFACTURER must determine whether it is an INCIDENT.
The following time lines apply in a case of:
Serious public health threat: IMMEDIATELY (without any delay that could not be justified) but not later than 2 calendar days after awareness by the MANUFACTURER of this threat.
Death or UNANTICIPATED serious deterioration in state of health: IMMEDIATELY (without any delay that could not be justified) after the MANUFACTURER established a link between the device and the event but not later than 10 elapsed calendar days following the date of awareness of the event.
Others: IMMEDIATELY (without any delay that could not be justified) after the
MANUFACTURER established a link between the device and the event but not later than 30 elapsed calendar days following the date of awareness of the event.
If after becoming aware of a potentially reportable INCIDENT there is still uncertainty about whether the event is reportable, the MANUFACTURER must submit a report within the timeframe required for that type of INCIDENT
 
Thread starter Similar threads Forum Replies Date
M Is a confirmation trial is obligatory in development of a product? Design and Development of Products and Processes 3
M Is a confirmation trial obligatory in development of a product? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Is double packaging obligatory for surgical single-use scalpels? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
J Are the Pre-Launch Control Plan and Pre-Launch PFMEA obligatory? FMEA and Control Plans 3
H Statistical process programs for PPAP: Expecting some to become obligatory? APQP and PPAP 9
E Is it obligatory that only trained personel can carry out calibration? General Measurement Device and Calibration Topics 1
M If there is any restriction on the transportation of Used Medical Device (with blood) from Canada to other countries? Canada Medical Device Regulations 4
M If there is any restriction on the transportation of Used Medical Device (with blood) from EU to other countries? EU Medical Device Regulations 2
adztesla Sterile packaging validation for medical device - ASTMD4332 and ISTA2A ISO 13485:2016 - Medical Device Quality Management Systems 3
P On-going stability study requirement for medical device US Food and Drug Administration (FDA) 2
K Medical devices for laymen and necesssary reprocessing Other Medical Device Related Standards 3
N Regulations for Medical Device in Jamaica Other US Medical Device Regulations 0
M Classification of device according to China medical device regulations China Medical Device Regulations 4
M Looking for kindly help... If there is any prohibits from contract manufacturing in US? About SR medical devices US Medical Device Regulations 0
adztesla Design change and Process change Class 3 medical device Medical Device and FDA Regulations and Standards News 2
K New set of medical devices - samples EU Medical Device Regulations 3
M Importing a custom component for our medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
FRA 2 FDA Reusable Vs Single Use Medical Devices in EU Other Medical Device and Orthopedic Related Topics 1
adztesla Adding new supplier/ Medical device design change ISO 13485:2016 - Medical Device Quality Management Systems 2
adztesla Medical device Drawing related change justification Other Medical Device Related Standards 0
O Medical Device With 2 Power Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Stoic Which medical device companies are using Recombinant Factor C (rFC) instead of Limulus Amebocyte Lysate (LAL) for endotoxin testing? Sustainability, Green Initiatives and Ecology 0
I Commercializing a non medical device software Medical Information Technology, Medical Software and Health Informatics 10
I Non-medical device software Preventive Action and Continuous Improvement 1
C California Prop 65 for medical device labels US Medical Device Regulations 3
adztesla GDP rules- redline drawings in medical device Medical Device and FDA Regulations and Standards News 2
R Instructions for Use supplied with Medical Device Components Medical Device and FDA Regulations and Standards News 4
M Deburring plastic medical device Manufacturing and Related Processes 1
W Training software medical device Medical Device Related Standards 0
S Modification of medical device EU Medical Device Regulations 2
I Commercialize a non-medical product ISO 13485:2016 - Medical Device Quality Management Systems 5
P Do Barcode scanners need validation in Medical device manufacturing? Manufacturing and Related Processes 6
F The use of 4G/Wi-Fi dongles in medical devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
B FDA Requirements for medical device to be used for training purposes only US Food and Drug Administration (FDA) 9
L Verification of wireless QoS and coexistence during FDA certification of medical devices Medical Device and FDA Regulations and Standards News 1
F Adding wireless communication to existing medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Medical device sizes and system and procedure packs EU Medical Device Regulations 0
U Does Medical Device training video falls under labeling requirement ? Other US Medical Device Regulations 4
dgrainger Informational Consultation: Medical devices – single identifier for similar highly individualised devices EU Medical Device Regulations 0
dgrainger Informational MHRA Guidance: Crafting an intended purpose in the context of Software as a Medical Device (SaMD) UK Medical Device Regulations 0
J Is ISO 13485 certification necessary for Class I (non-sterile) medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 7
L Directive 2012/19/EU for medical software EU Medical Device Regulations 2
S Sample size considerations in medical process qualification Reliability Analysis - Predictions, Testing and Standards 11
B Medical device disposables - Stored in a controlled environment for 4 months - Effects on Sterilization Medical Device and FDA Regulations and Standards News 1
W MOOPs in Medical Device Charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
adztesla Specification (Significant digits) - Measurement (Medical device) ISO 13485:2016 - Medical Device Quality Management Systems 3
S Accessory to medical device or not? EU Medical Device Regulations 1
M Looking for medical grade wall adapter with 50Hz power transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Microbial Barrier Testing for Medical Devices Qualification and Validation (including 21 CFR Part 11) 2
R Medical equipment selection for refurbishment US Food and Drug Administration (FDA) 1

Similar threads

Top Bottom