What Clinical Data is required for Drug Delivery System

Y-Spi

Registered
#1
Hello Colleagues :bigwave:,

Does anyone know what clinical data is required for submission of a drug delivery system, to obtain a CE mark (see explanation below)?
Can the intended use be general, and not clinical (e.g., the device is intended for delivery of medicinal substances via..."?

We developed a drug delivery system that was tested (in-vitro as well as clinically) with only one drug. The device is not pre-filled but can be used only with pre-filled cassettes. Currently the cassettes contain only one drug, but we intend to produce cassettes with other drugs in the future. The device in intended for re-use.

Many thanks!
Y
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
Does anyone know what clinical data is required for submission of a drug delivery system, to obtain a CE mark (see explanation below)?
I suggest you start by looking at the MEDDEV guidances on clinical evaluation and clinical investigation Guidance - European Commission

Can the intended use be general, and not clinical (e.g., the device is intended for delivery of medicinal substances via..."?
It really depends in the specifics of your device... but I am doubtful when it comes to a drug delivery system.
Start with the clinical evaluation / investigation part and it will become clearer to you.

Shimon
 
Thread starter Similar threads Forum Replies Date
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 6
U Do we need clinical trial data for Class IIa medical device under MDR EU Medical Device Regulations 7
CPhelan Using clinical trial safety data for evidence for CE marking EU Medical Device Regulations 7
CPhelan Do you require MDSAP for CE Marking of a Medical Device or is ISO13485:2016 with clinical data sufficient? CE Marking (Conformité Européene) / CB Scheme 5
M Informational Health Canada begins release of clinical data Medical Device and FDA Regulations and Standards News 0
M Informational USFDA Final Rule – Human Subject Protection; Acceptance of Data From Clinical Investigations for Medical Devices Medical Device and FDA Regulations and Standards News 0
M IVD - Generating clinical data for IVD’s in the EU EU Medical Device Regulations 2
W ISO 14155 - Electronic clinical data systems - Definition Other Medical Device Related Standards 0
MDD_QNA Clinical data requirements - Changing a Class I Medical Device to Class II EU Medical Device Regulations 5
D Use of clinical data in a 510(k) submission Other US Medical Device Regulations 6
bio_subbu USFDA issues draft Guidance on the Acceptance of Medical Device Clinical Data Other US Medical Device Regulations 1
R FDA's guidance - Evaluation of Sex-Specific Data in Medical Device Clinical Studies Other US Medical Device Regulations 3
B Health Canada Acceptance Policy of Foreign Clinical Trial Data for Devices Canada Medical Device Regulations 2
M Category I Medical Device Clinical Data Requirements EU Medical Device Regulations 3
C Retention of Clinical Trial Data after a Clinical Trial has ended Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 5
C Key Links (Library) to search for updated Clinical Data (Is & IIa) EU Medical Device Regulations 1
E Need Corrective Action for Class III Device with no Clinical Data CE Marking (Conformité Européene) / CB Scheme 9
S Clinical or Non-Clinical Data in 510(k) Submission for a type II Medical Device US Food and Drug Administration (FDA) 1
D Registration of Class IV Device -Clinical Data and ISO Certificate Canada Medical Device Regulations 3
A Collecting Clinical Data using CE Marked Devices EU Medical Device Regulations 4
H 510(k) Filing and Predicate Device Clinical Data 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Technical File as per ANNEX VII and Sample Size for clinical data CE Marking (Conformité Européene) / CB Scheme 12
C Clinical Data to Prove Equivalence Other US Medical Device Regulations 4
N Clinical Data for Class 1 Devices? CE Marking (Conformité Européene) / CB Scheme 6
I 510k Approval - Does the FDA look for Clinical Data 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
L Clinical Data Report - Literature Route - MEDDEV guidance (MEDDEV 2.7.1) EU Medical Device Regulations 28
S Australia TGA Clinical Evaluation Report (CER) Other Medical Device Regulations World-Wide 0
L Clinical research - Request from ANVISA ISO 14971 - Medical Device Risk Management 1
E Use of OQ cleared batch for stability and clinical study use ISO 13485:2016 - Medical Device Quality Management Systems 0
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
M Clinical Development Plan - Advice Requested EU Medical Device Regulations 9
T Clinical evaluation of a new medical device EU Medical Device Regulations 0
dgrainger Informational MDCG 2020:13 - Clinical evaluation assessment report template EU Medical Device Regulations 0
shimonv Minimum 60601 requirements for a clinical trial IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Need for a cleanroom in the manufacture of a medical device for a clinical trial EU Medical Device Regulations 4
R Impact of changing manufacturer post clinical investigation CE Marking (Conformité Européene) / CB Scheme 3
C MDCG 2020-06 Clinical evidence legacy devices EU Medical Device Regulations 3
M Clinical Evaluation Plan vs. PMCF Plan EU Medical Device Regulations 19
S Selling Clinical Practice Advice Other US Medical Device Regulations 7
S FDA's E submitter - For clinical electronic thermometer Medical Device and FDA Regulations and Standards News 6
M Good Clinical Practice (GCP) - ISO 14155 vs. ICH Other ISO and International Standards and European Regulations 10
S Article 61 - MDR Clinical Investigation Requirement EU Medical Device Regulations 5
J Are DAM tools, Mobile application, clinical dashboard, etc. considered accessories to medical device or supportive functions? Manufacturing and Related Processes 4
M Can someone share a scrubbed version of Clinical Evaluation Plan (CEP) EU Medical Device Regulations 4
R Do Medical Device Complaints include Clinical Trials? Medical Device and FDA Regulations and Standards News 3
S EU MDR Annex XIV - Clinical Evaluation Plan - What do these methods mean? EU Medical Device Regulations 12
dgrainger Informational MHRA updated guidance: Clinical investigations of medical devices – guidance for manufacturers Medical Device and FDA Regulations and Standards News 0
vickyva Checklist for Clinical Investigation compliant to EU-MDR EU Medical Device Regulations 4
M Risk Classification For Supplier - Clinical Research Organisation (CRO) Supply Chain Security Management Systems 3

Similar threads