What do you use to Calibrate 40" Calipers
Jason S
Our company is going for ISO 13485 with an audit in March. One of the comments during Phase I was regarding the fact that some of our calibrations are not the complete range of the device, specifically our 40" calipers.
My question is, how or what do you use to calibrate digital calipers at 40"?
We use gage blocks up to 6" and have 12" and 24" standards for larger calipers.
I can't seem to find a standard to use for that length.
Jerry Eldred
Re: Calibrating 40" Calipers
The common practice is wringing blocks together by competent dimensional metrologist to obtain the 40 inches. Sounds like you have enough blocks to make 40 inches. Beware that the longer the block, the greater the temperature coefficient errors. So if you are not in a 68 degree room, the block errors will be bigger. But bottom line, if you have blocks totalling a combined 40 inches, you should, with proper technique, be able to calibrate 40 inch calipers, I believe.
Jason S
I don't have the blocks to do that. The blocks are a .05-4" set. The 12" and 24" standards are micrometer standards purchased individually.
Jason S
The blocks are only .05-4". We use individual 12" and 24" micrometer standards for our 12 and 24" calipers. I wouldn't have enough blocks to make up 40".
You should have the thing calibrated by an outside lab, or buy the standards necessary to do it in-house.
Jason S
I would like to keep this in house; however, I cannot find the correct length standard. I've seen 24" and 48".
Anyone run into or know where I can purchase larger size calibration rods?
What is the range of measurements you'll do with the caliper, and what and what kind of tolerances are you working with? Presumably your measurements will be >24" and ≤ 40".
Jason S
We deal with all the sizes in between. Occasionally even over 40" and we have to use a steel rule. Tolerances are typically .010" or larger
Mikishots
Once you have the cost of the rods I would then check this against the cost of sending the unit(s?) out. DO not forget your time and the cost of keeping the rods calibrated.
Here is a link to some sources for gauge rods of the length you are looking for.
http://longislandindicator.com/p58.html
http://www.starrett.com/metrology/m...urrentPage=1&displayMode=grid&sortBy=none/asc
:2cents:
Agreed. The cost of the rods and the investment in time is much higher than what an outside lab would charge. If one was measuring throughout the range, as indicated by the OP, one would want to at least measure every half a foot, so we're looking at 6 or 7 rods. Not cheap.
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
C Seeking a holder/jig to calibrate large OD micrometer calipers (up to 23") General Measurement Device and Calibration Topics 8
W ASTM Spec or what standard is used to calibrate calipers? General Measurement Device and Calibration Topics 2
F Is there a standard that I should use to Calibrate Calipers and Micrometers? General Measurement Device and Calibration Topics 8
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
I Can I Calibrate On Demand? General Measurement Device and Calibration Topics 9
F Crimper Calibration - How do you Control & Calibrate your crimpers? Calibration Frequency (Interval) 13
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
K New Kilogram - The process we use to calibrate the standards General Measurement Device and Calibration Topics 6
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
L Re-calibrate after software update? Calibration Frequency (Interval) 3
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T How to Calibrate Taper Gages General Measurement Device and Calibration Topics 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
D Lesser expensive to calibrate various test weights General Measurement Device and Calibration Topics 5
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
P How do you calibrate a VNA (Vector Network Analyzer) cal kit? General Measurement Device and Calibration Topics 4
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
bobdoering Who says you can't calibrate a steel rule! Funny Stuff - Jokes and Humour 14
S How can I Calibrate a Biosafety Cabinet? General Measurement Device and Calibration Topics 1
C How to Calibrate a Viscosity Tester General Measurement Device and Calibration Topics 3
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
T How do I calibrate NPT Thread Plug Gages? General Measurement Device and Calibration Topics 9
E Do you need a Certification to Calibrate Gauges? General Measurement Device and Calibration Topics 2
B Calibrate Hole Presence Detection Device? General Measurement Device and Calibration Topics 5
A Guideline to Calibrate Chamber Temp. & Pressure of a Medical Steam Sterilizer General Measurement Device and Calibration Topics 8
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
B ISO/TS 16949 Section 7.6.3 - Use of Manufacturer to Calibrate IATF 16949 - Automotive Quality Systems Standard 21
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
N Laboratory that is able to Calibrate Dew Point Meter up to 60?Cdp - Needed General Measurement Device and Calibration Topics 1
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
C How to calibrate or control patterns for fiberglass General Measurement Device and Calibration Topics 4
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
I Need help finding an Accredited Lab that can Calibrate Brunson 76-RH Survey Transits General Measurement Device and Calibration Topics 5
M Controlling Temperature and Time - Do I need to calibrate the thermostat? IATF 16949 - Automotive Quality Systems Standard 4
D How much Tolerance do I use to Calibrate a 150kgs capacity Scale? General Measurement Device and Calibration Topics 5
M To Calibrate or Not to Calibrate various Weight Scales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
N Thermocouple Inhomogeneity - Need to calibrate a roll of 50m thermocouple type K General Measurement Device and Calibration Topics 3
J Calibrate before use required? Inspection Gages and Measurement Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
S 3 Point Calibration or Calibrate to Extreme High and Low? General Measurement Device and Calibration Topics 9
S Is it OK to calibrate a measurement device tighter than manufacturer specifications? General Measurement Device and Calibration Topics 18