What edition of the IEC 60601-1 standard is required?
ganglai
Registered Visitor
#1
I was trying to get my head around what edition of the IEC 60601-1 standard is required on European Commission\harmonised standards for medical devices page
There are the following notes:
Addendum to Note 1 and Note 3 concerning dates of cessation of presumption of conformity when applying EN 60601-1:2006.
The date of cessation of presumption of conformity when applying EN 60601-1:2006 is 31/12/2017. However the Annex ZZ to EN 60601-1:2006 ceases to specify the presumption of conformity with the Essential Requirements of Directive 93/42/EEC on 31/12/2015. As from 01/01/2016, only the clauses and sub-clauses of EN 60601-1:2006 corresponding to the clauses and sub-clauses referred to in Annex ZZ to EN 60601-1:2006/A1:2013 provide presumption of conformity with the Essential Requirements of Directive 93/42/EEC, to the extent indicated in the Annex ZZ to EN 60601-1:2006/A1:2013.
Thanks for your help in advance!
Elsmar Forum Sponsor
H
hcrook
#2
IEC 60601-1 3.1 Edition (harmonized as EN 60601-1:2006/A1:2013) is best to apply since its transition date for the EU is December 31, 2017. That is to say, after the transition date, compliance with IEC 60601-1 3.0 edition (EN 60601-1:2006) will no longer be accepted as Evidence of Conformity with the Essential Requirements of the MDD.
#3
In your opinion, does this mean that for medical devices in conformity with IEC60601-1 ed.3, already on the market in Europe, the IEC60601-1 tests must be updated according to the latest version?
ganglai
Registered Visitor
#4
I would say Yes is the answer. We have had experience that CE marked product had to remove CE marking because of not being able to demonstrate compliance to latest harmonised standard.
But I would love to know how many companies are strictly following the latest harmonised standards.
Ronen E
Problem Solver
Moderator
#5
That sounds a little strange to me.
If there are more than one version that are harmonized, you should be able to use any, not necessarily the latest amongst them. Once an older version ceases to be harmonized, you must of course move on to a newer one that is still on the list.
Theoretically a company can also be in compliance (and eligible for CE marking) while applying an older, non-harmonized standard (especially one that was harmonized not long ago), because the directive's requirement is compliance with the ERs, taking consideration of the state-of-the-art, rather than a strict requirement to apply harmonized standards; but I can see how in practice that would be difficult to pull off in the context of 60601.
ganglai
Registered Visitor
#6
Standards always allow for transition period and everyone understands. It could've clarified the word "latest" in the reply....
I agree with the theory. But in practice, it could end up CE mark being removed. Or it could also be ok.
marmotte
Diapason Consulting
#7
Interested in this discussion too.
Hi Ronen :)
My understanding is that compliance to the current/latest harmonized standard gives you a presumption of compliance to the ER/EP without the need for further documentation.
If this is not achieved, i.e. if there is only compliance/certification to a legacy harmonized standard, (or even to no standard at all), then the manufacturer must justify/demonstrate compliance with the EP through documented evidence.
60601-1 (and whole series) is about safety and efficacy which is a very comprehensive topic. Without at least the legacy standard I think it would be pretty much impossible to demonstrate that safety and efficacy is up to par (except maybe for some very specific extremely simple products which are considered electrical for whatever reason but not really relevant to 601 - not that I could think of anything ???? ). Anything with a power plug or a battery would definitively need to comply with / be certified to the 2nd edition of 601 + to document through a gap analysis that the new/modified requirements of the 3rd (or 3.1) do not raise any concern compared with the 2nd ed.
At least that has been my understanding so far - and the way one of the products I am in charge of is documented ... So any comments are welcome....
Thread starter Similar threads Forum Replies Date
T IEC 60601-1 vs IEC 60601-1-2 4th Edition Other ISO and International Standards and European Regulations 1
Ed Panek Transition to IEC 60601 4th Edition - Risk Analysis and test submissions CE Marking (Conformité Européene) / CB Scheme 2
O IEC 60601-1-2 4th edition transition deadline Dec 31 2018 CE Marking (Conformité Européene) / CB Scheme 1
T IEC 60601 3.1 edition Gap Analysis wanted IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T What is PESS? Cited in Edition 3.1 of IEC 60601, clause 14.1 note 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T Is a new edition of the IEC 60601 in preparation for 2019 (Amendment 2)? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Using IEC 60601-1 2nd Edition for "Evaluations" IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-2-25 Reference to 60601-1 2nd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 4th Edition - Is a "Red-line" version available? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P FDA's deadline for IEC 60601-1 3.1 edition compliance - Aug 1, 2016 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E IEC 60601-1 3?EDITION IN VENEZUELA and COLOMBIA IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L IEC 60601-1 Amendment 1 3rd Edition and Brazil ABNT NBR IEC 60601-1:2010 IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
M Reducing cost of IEC 60601 3rd edition testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
B Did UL 60601-1 change when the IEC 60601-1 standard changed to the 3rd edition? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Recognition of and Transition to IEC 60601-1 3rd Edition in Korea IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J IEC 60601 Third Edition Labeling Requirements - Section 5.2.2.1 (g) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Which IEC 60601-1 Edition applies in different countries IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B When is Compliance required for 4th edition IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E FDA Implementation Date for IEC 60601 - 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Infusion Pump and IEC 60601 2nd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
N IEC 60601-1 , 3rd edition and the FDA - Special 510k to my updated Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 40
S Are you struggling with your IEC 60601-1 3rd edition testing lab? IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
A Essential Performance in IEC 3rd edition of the 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 62133 to satisfy 60601-1 3rd edition - What happen to 60086-4 Other Medical Device Related Standards 18
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
thisby_ IEC 60601-1 3rd edition Transition Period - US vs. Europe US Food and Drug Administration (FDA) 5
M Rework Devices placed on the market - IEC 60601-1 Requirements 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
T IEC 60601-2-27 - Comparison of 2nd to 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
W Is there an IEC 60601 3rd Edition 1st Amendment? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B IEC 60601-1 3rd Edition adoption - I cant find this standard on the ISO page Reliability Analysis - Predictions, Testing and Standards 9
I Acceptability of Existing Devices complying to IEC 60601-1 2nd Edition after Jun'2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Switching Power Supplies Certified to IEC 60601 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Current Products on the Market and IEC 60601 3rd Edition Safety Re-Test Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
M 60601-1 3rd Edition Compliance with Particular Standards & Equivalency IEC/EN stds EU Medical Device Regulations 6
E Retrofitting Risk Management for IEC EN 60601 3rd Edition Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
J Acceptance of IEC EN 60601-1 3rd edition in China China Medical Device Regulations 8
J IECEE Guidance Document for ISO 14971 Risk Analysis - IEC EN 60601-1:2005 3rd edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M UL (Underwriters Laboratories) Publishes IEC 60601 3rd edition FAQs IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H IEC 60601-1 3rd Edition Certification - New Products Only? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
B IEC 60601 3rd Edition - Test Report Form UL uses for testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
J IEC EN 60601 3rd edition CF vs. CFD IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L When IEC 60601-1 2nd edition can not be used for CE mark medical device application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1-3 2nd Edition (X-ray Radiation Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C Relationship of IEC 60601-1-6:2006 vs. IEC 60601-1 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C New edition of IEC 60601-1 - Risk assessment to determine the liquid - 11.6.3 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1 3rd Edition Recognition in Canada IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
T Harmonization date of IEC 60601-1 , 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 135
Similar threads
Top Bottom