What EMC 60601-1-2 version to use with 60601-2-47
We are recertifying for MDR with our portable ECG holter device (currently our device has medical CE certification under MDD).
Since we have a ambulatory electrocardiographic system, we need to use the latest IEC60601-2-47:2012.
Now we have a discussion with our notified body on which version of IEC60601-1-2 we must use.
In according to Clause 201.2 (Normative references) of 60601-2-47:2012, we need the IEC60601-1-2:2007 version.
Our notified body states we need the latest state of the art. Since IEC60601-1-2:2207 is withdrawn, our notified body states that we need to use the harmonized IEC60601-1-2:2014.
Can we legally still use IEC60601-1-2:2007, since this is the normative reference for IEC60601-2-47:2012?
Our device has passed all the IEC60601-1-2:2007 tests. Testing against IEC60601-1-2:2014 would implement extra time and costs and potential design changes, so this is not our preferred route.
Any help would really be great, since we could not find any rational on how to handle this situation.
Thanks,
Sander
yodon
Apologies for the delay replying.
"Legally" is probably not the right term. (In theory) Compliance to a particular standard is optional. That aside, there is the concept of 'state of the art' and with -1-2 being both superseded by a later revision AND the fact that :2014 is the harmonized version, I'd say compliance to the :2007 version would be a tough row to hoe. You *might* be able to do a gap analysis and try to justify why the results from :2007 tests are suitable but that's a gamble at best.
Benjamin Weber
  1. IEC does not really withdraw old standard revisions, they are marked as revised. There also usually no transitional periods given in IEC versions. This is usually only part of national versions, such as EN. If the national legislation still require compliance with such a revised standard, IEC cannot "prohibit" that. (Think of 60601-1 Ed. 2.0 which was required by China until 2020 or 2021, not sure about the exact year)
  2. IEC standards are usually not harmonized with European directives/regulations, but EN versions are, which can have substantial national differences. (But: I don't think that there are such differences between the IEC and EN versions of 60601-1-2:2014).
  3. Neither the IEC nor the EN versions of 60601-1-2:2014 are harmonized with MDR (in contrast to MDD).
  4. If a standard refers to a dated version of another standard, this dated version must be applied to be compliant with the referring standard.
  5. If the notified body refers to the "state-of-the-art" version of a standard, they definitely can do so. Espacially if there is no harmonized version of the standard for MDR!
What does this mean in summary:
  1. If the particular standard refers a (out-)dated version of 60601-1-2, you still have to apply the old version.
  2. If the notified body requires you to be compliant with the latest version, you should comply with the new version.
  3. As yodon said: Perform a gap analysis and identify the more stringent requirements for each test. Than test according these worst-case-conditions nad test levels. Eventually you will comply with both standard versions!
  4. If there are conflicting requirements, apply your risk management process to identify the requirement which will lead to the lowest risk. Discuss those requirements with your notified body!
Thanks Yodon and Benjamin, your answers are really helpful. I'll perform a gap analysis and apply our risk management process where needed.
After that, probably some discussions with our notified body is needed.
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Question: Do we consider 60601-1-2 (4th ed) EMC testing a mode for a diagnostic port? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-2-2 - EMC Requirements - Surgical instrument manufacturer IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
C EMC Test Report Requirements to IEC 60601-1-2:2009 ISO 13485:2016 - Medical Device Quality Management Systems 4
P New EMC standard IEC 60601-1-2:2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 60601-1-2 EMC and WiFi IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
C An ongoing EMC test in accordance with IEC/EN 60601-2-47 EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S EMI/EMC Failure Cause Analysis per IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Do HF accessories require EMC testing according to IEC 60601-1-2? IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
S Does a 12V DC device need EMC (IEC 60601-1-2) Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 60601-1: EMC problems: Confused about 915MHz ISM band in Europe IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Particular standard IEC 60601-2-40 references obsolete EMC standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601-1-2 - EMC Retesting Rational - Minor Component Changes IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
amjadrana Medical Device EMC requirements - Canada - Testing to FCC part 18, EN 55011, EN 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
L IEC60601-1-11 Home Healthcare + IEC60601-1-2 EMC (Functional Earth allowed?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M EMC Directive and product safety standards CE Marking (Conformité Européene) / CB Scheme 1
B Evaluation of Basic Safety during EMC Immunity or Climate Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Does EMC Directive 2014/30/EU cease to apply when the Machine Directive 2006/42/EC is applicable? CE Marking (Conformité Européene) / CB Scheme 4
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F EMC testing for a system that is provided with a computer IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
O EMC - EFT (Electrical Fast Transients) on SIP couples erroneously with other cable CE Marking (Conformité Européene) / CB Scheme 1
K Is a 3V torch classified as medical electrical equipment needing EMI/EMC testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Arguing with the test lab about failing in an EMC test IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
E Airborne Medical Equipment EMI/EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Portugal ISO 13485 Certification and EMC Test Houses Other Medical Device and Orthopedic Related Topics 3
O Medical Device EMC Risk Management CE Marking (Conformité Européene) / CB Scheme 4
O Guide on EMC for Functional Safety CE Marking (Conformité Européene) / CB Scheme 3
T CE Marking & Compliance to MDD, EMC, RoHS, and GPSD REACH and RoHS Conversations 4
D EMC Testing for a Production Fixture CE Marking (Conformité Européene) / CB Scheme 0
Q IVD vs EMC OJ (Official Journal of the European Union) Harmonised Standard Conflicts EU Medical Device Regulations 1
A EMC Emission Issues - Communication between PC and Monitor IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Is EMC + Radio Re-Test Required? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Particular standards, essential performance in EMC Immunity Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F EMC (Electromagnetic Compatibility) Conformity Charts - IFU Requirements Misc. Quality Assurance and Business Systems Related Topics 3
G CE Marking - Low Voltage Directive and EMC Directive CE Marking (Conformité Européene) / CB Scheme 4
T Firmware version in EMC report / DoC CE Marking (Conformité Européene) / CB Scheme 1
W EMC, ESD for battery operated apparatus Other Medical Device and Orthopedic Related Topics 1
N New EMC IEC60601-1-2 4.0 Ed. Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B What is our requirement for maintaining the EMC series of standards? IEC 61000 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P EMC requirements for an accessory to a medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T EMC for Medical Devices in Home Environment Other US Medical Device Regulations 4
J EMI/EMC Test differences for Radiography Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M EMC Testing for a US-only, Home-Based Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Safety & EMC certificate updated when MCU (MIcro-Controller) is changed? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2