Agree with c.itch to a point. I DO generally see SRSs written prior to software development. In fact, the developers are typically begging for it! But that's sliding off topic a bit. Anyway, agree that the V model is not required. Neither is 13485 but pushing through regulatory without 13485 compliance is more challenging. Similarly, if you try to take a radical (i.e., not state of the art / common practice) approach to software development it will likely get higher scrutiny. I find SDDs to be marginally useful. If they are needed to lay out an algorithm or to document design decisions that aren't obvious, then they're useful. We take a minimalist approach and generally have, for design, an architecture diagram, maybe some state models, and anything that is not obvious. But as c.mitch points out, I would not submit source code as design. That, to me, is just waving a red flag in their face.
 
Elsmar Forum Sponsor

robert.beck

Involved In Discussions
#12
From a practical point of view, in my limited experience with about 8 510k projects that had significant software components and about an equal number with firmware as an important component, the software developers and project managers don't understand what an SDS is let alone what its value is. FDA should either drop it from the list of required documents, or do a much better job explaining what it's supposed to document. The other documents in the guidance make intuitive sense, or don't take much explaining to get their rationale across.

the SDS always stands out as an outlier - what it is, why do we need to do it, and can we just combine it with the SRS or the architecture?

I usually do the SDS after the fact, but keep it consistent with the architecture.

In response, to other comments on this thread, I don't see how source code can be part of SDS. source code is the medical device. it's the final output, not some intermediate step along the way. granted that source code should be expected to change in response to requirement changes and detected bugs, but this malleability doesn't make it either an SDS or an SRS.

I have observed this scenario often enough that it seems normal:

1) there is a discussion about what the software is supposed to do.
2) some coding happens.
3) somebody generates a flow chart or an architecture diagram that describes what this early version of the software does.
4) there is more discussion, and new requirements and constraints are spoken of.
5) the flow chart is updated.
6) then the final coding is done.
7) it's checked against the flow chart, and either or both are modified as necessary.
8) there are multiple rounds of testing until the software is done.
9) at this point, it's possible to write the SRS, architecture, hazard analysis. The overall device hazard analysis (assuming this is not a software-only device) has been long done, and it may be supplemented now that the software is completed. the product specifications were also done a long time ago, and they may be supplemented now that the exact role of the software is known. Steps 7 and 8 may take a long time and include many iterations.
 
#13
Hi Everyone,

This topic is very old, but I wanted to add a small comment to a statement mentioned by sagai that was probably overlooked. :2cents:

sagai said:
"... other topic but what if the unit test code is considered as a specification for the source code?"

Maybe you were talking about something different, but I think you are refering to "Test Driven Requirement"

(There are interesting articles in the web about this subject)

Do you see this happening anytime soon in the Medical Device Software Development world?
 
R

robertjbeck

#14
i have seen test driven development (TDD) recently for two simpler medical devices.

in one case, the developer called what he was doing TDD but it was not; he did not really understand what this meant. however, he did some other interesting things that worked very well for verification, and I don't know if they have a formal name.

in another case, the software group started writing test cases before requirements specifications. this got the quality/regulatory department upset because they thought it was an attempt to write software without having requirements, but it was actually a good process.
 
Thread starter Similar threads Forum Replies Date
M Source code and explanation for computer simulation in camera calibration, HELP Calibration and Metrology Software and Hardware 2
J Can someone give a hint about FMEA about software source code FMEA and Control Plans 7
J Starting a Weibull Analysis: Source on how to perform Weibull Reliability Analysis - Predictions, Testing and Standards 2
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L IEC 62368-1 / Electrical Energy Source vs. Power Source? CE Marking (Conformité Européene) / CB Scheme 2
F Electrical Metrology (Source/Generation and Measurement) General Measurement Device and Calibration Topics 1
R Supplier related drawings and verification of process requirements - Source Inspection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R Supplier Controls we can place on Single-Source Suppliers ISO 13485:2016 - Medical Device Quality Management Systems 2
N Time source for paper-based documentation (research nurses) US Food and Drug Administration (FDA) 1
C Looking for source of silicone contamination in ISO Class 8 Cleanroom Other Medical Device Related Standards 2
Ron Rompen Source needed for Spline Form gauge (GO and NOGO members) General Measurement Device and Calibration Topics 3
PastorBee13 Source for the blades for Reef Snap Gages General Measurement Device and Calibration Topics 2
S Source of QMS templates including templates for design verification/design validation Document Control Systems, Procedures, Forms and Templates 2
Ed Panek Sensitive Supplier Issue - Single source supplier ISO 13485:2016 - Medical Device Quality Management Systems 6
H Gamma Irradiation - dose audit and source replenishment Other Medical Device Related Standards 13
I Uncalibrated source to calibrated meter, then compare to UUT General Measurement Device and Calibration Topics 4
T Definition Sole source VS. Single Source - Definitions Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 2
P Datron/Wavetek 4808 Option 70 Wideband Source - User and Service Manuals needed General Measurement Device and Calibration Topics 2
Q Source of practice to Evaluate Risks? (ISO 9001:2015) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M Contacting a "Must Use" (aka Sole Source) Supplier's Registrar Supplier Quality Assurance and other Supplier Issues 6
E Source for The Enforcement Regulations in English - Article 228-20 Japan Medical Device Regulations 8
R FAI (First Article Inspection) on source-controlled parts AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
E PMDA Act, Nov 2014 revisions - source in English? Japan Medical Device Regulations 3
A Reference Source for Polarity Symbol Other ISO and International Standards and European Regulations 1
J Legal source to purchase a copy of the 9001:2015 Draft ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Crusader Calibration Source not ISO 17025 accredited ISO 17025 related Discussions 6
D Product Testing Source - Specific testing on an electronic parking brake IATF 16949 - Automotive Quality Systems Standard 1
K Ideas for an Open Source Calibration Database Software Calibration and Metrology Software and Hardware 3
R What can I expect in a Government Source Inspection? Quality Manager and Management Related Issues 13
A How to understand the source of variability and "stream-to-stream variation" Six Sigma 8
B Convenient source for finding a listing of changes to BS EN ISO standards Various Other Specifications, Standards, and related Requirements 5
M KANO Model - Changing the source Quality Tools, Improvement and Analysis 2
G Six Sigma Case Studies - Does anyone have a good source for LSS case studies? Six Sigma 8
S Source of Online or Home Based HACCP - FSSC22K Training Food Safety - ISO 22000, HACCP (21 CFR 120) 2
H Source for Compiling and Maintaining an EHS Legal Register Miscellaneous Environmental Standards and EMS Related Discussions 4
C Nuclear Industry--Unqualified Source Material CMTR Requirements Various Other Specifications, Standards, and related Requirements 1
N Boeing Portal - ASAP for Source Inspection Customer and Company Specific Requirements 4
W Seeking a Freelance Mechanical Inspector - Source Inspection in Bangalore, India AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W X-ray based NDT Test Equip. Mfger. seeking CE Mark - Must X-Ray Source be included? CE Marking (Conformité Européene) / CB Scheme 5
G Source of a Clear Fixture Block on a x-y Video Microscope - Rafaelco? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
M TS 16949 7.4.1.3 - Difference between Approved Source and Customer Designated Source IATF 16949 - Automotive Quality Systems Standard 2
P Source Inspection vs. Right of Access clause in a Purchase Order - Differences AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
N Outside source to provide Basic Measurement Training - Louisville, KY Training - Internal, External, Online and Distance Learning 1
M Incoming Source of Variation & Source of Variation in Process Flow Chart IATF 16949 - Automotive Quality Systems Standard 3
E Faro Laser Tracker Calibration Source Needed General Measurement Device and Calibration Topics 5
S Separate Medical Device 12V Power Source - What inspection and testing is required? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Open Source Software - OpenOffice vs. LibreOffice After Work and Weekend Discussion Topics 1
Marc Google Wave going Open Source After Work and Weekend Discussion Topics 1
A Second Source for a similar Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3

Similar threads

Top Bottom