I have an interesting situation with a Complaint and I'm not sure if it's reportable or not. A user engaged a consumer medical device, a single piece of molded plastic, in a normal manner but for some reason the device got stuck and required the use of forceps at an urgent care facility to remove. No injury was reported despite being in the user's body for over 24 hours.

I'm going through the reportability criteria and since there was no injury it doesn't satisfy those criteria. Looking at the device malfunction side, I could see how if this happened again it might cause serious injury due to infection given the area of use. However, the device didn't "malfunction" in the sense that it failed or broke or even didn't do what it was supposed to do. It's entirely possible the user could have put it right back in their body and not had this happen again--same piece of plastic, nothing different at all.

The FDA defines "malfunction" in 803.3 as:
(k) Malfunction means the failure of a device to meet its performance specifications or otherwise perform as intended. Performance specifications include all claims made in the labeling for the device. The intended performance of a device refers to the intended use for which the device is labeled or marketed, as defined in 801.4 of this chapter.
The thing is here, the device did exactly what it was intended to do (it never stopped performing its intended use). Sure it's supposed to come out easily too, but there's no obvious aspect of the device that created this situation. It's basically some freak anatomical coincidence. Is this a "malfunction" of the device?

We're just the private-label distributor so I'll report it to the actual Manufacturer, but I'm curious what everyone thinks.

Thanks,
~Josh
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
Hi Josh,
For sure, you have to report it to the legal manufacture. The device itself did not break but it could be that its design is inherently flawed, and thereby not achieve its intended purpose. It could also be an isolated event caused by user mishandling.
The manufacture will need to review the records, consult their medical director, and take action as needed.

Shimon
 
Thread starter Similar threads Forum Replies Date
C Medical Device Malfunction - Customer Complaint - Response vs. Correction Other US Medical Device Regulations 3
C Medical Device Malfunction during misuse - Does this need to be reported to the FDA? Other US Medical Device Regulations 5
L Device shows an Error message - Malfunction or not? Other Medical Device and Orthopedic Related Topics 12
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 1
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 0
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
M Sample record for verification performed by importers before placing a device on the market EU Medical Device Regulations 0
S IVD Device History Record ISO 13485:2016 - Medical Device Quality Management Systems 3
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
K How to handle GTINs for different configurations of one device with one part number? Other Medical Device Related Standards 0
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
C Multiple owners for one listed device Medical Device and FDA Regulations and Standards News 1
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
J Promoting and marketing of a non approved device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Russian Device to US US Food and Drug Administration (FDA) 5
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
S DHF/DMR/MDF for a software-only, cloud-based, single-instance device Medical Information Technology, Medical Software and Health Informatics 2
V MDSAP Requirements for Device Designer Canada Medical Device Regulations 1
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
E Procedure ( SOP) for Device Master Record ( DMR ) and for Device History Record (DHR)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
J US Device Listing when not the 510(k) owner Medical Device and FDA Regulations and Standards News 2
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
G Voluntary Class I Device Filing Other US Medical Device Regulations 4
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
N Is this a single integral drug device combination product EU MDR CE Marking (Conformité Européene) / CB Scheme 1
S Medical Device MRI Compatibility EU Medical Device Regulations 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
L Medical device HIPAA compliance in encryption Medical Information Technology, Medical Software and Health Informatics 1

Similar threads

Top Bottom