What is satisfactory & acceptable for Coliform - Staphylococcus - Hands and Surfaces

D

DAWELAH

Dear Friends
Please what are the satisfactory, acceptable, and dissatisfactory limits for below tests in HANDS and Surfaces
- Coliform
- Staphylococcus

Thanks
 

Stijloor

Leader
Super Moderator
Re: What is satisfactory & acceptable for Coliform - Staphylococcus - Hands and Surfa

A Quick Bump!

Can someone help?

Thank you very much!!
 
D

DAWELAH

Re: What is satisfactory & acceptable for Coliform - Staphylococcus - Hands and Surfa

Dear Friends
Please what are the satisfactory, acceptable, and dissatisfactory limits for below tests in HANDS and Surfaces
- Coliform
- Staphylococcus

Thanks

please any reply
 
Thread starter Similar threads Forum Replies Date
P Acceptable quantity levels/Sigma levels for reducing the frequency of quality tests. Six Sigma 7
R Using RPN to Confirm Risk Reduced to an Acceptable Level Risk Management Principles and Generic Guidelines 12
Eva lee Which is the range for the validated batch size(for example ±5%, 10%) is acceptable? Manufacturing and Related Processes 8
S Examples of FDA acceptable Software Design Specification (SDS) Medical Device and FDA Regulations and Standards News 6
Q Acceptable Tolerance - how to derive? Measurement Uncertainty (MU) 16
W Misinterpretation of requirement acceptable as root cause? Problem Solving, Root Cause Fault and Failure Analysis 19
J Standard used to determine if check fixture gauge is acceptable Reliability Analysis - Predictions, Testing and Standards 2
M Measurement Error - How to determine what is acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
M Are Fungal counts acceptable in class 1000 clean rooms? Other ISO and International Standards and European Regulations 2
T ISO 14971-2019 doubt - Evaluate if estimated risks are acceptable ISO 14971 - Medical Device Risk Management 9
samer Acceptable limits for Spills - Tracking Hydraulic Spills ISO 14001:2015 Specific Discussions 3
V Acceptable maximum RSD (relative standard deviation) for an sample size Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Gage R&R Acceptable (10-30%), deduct Total Variation from Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Q Acceptable calibration accuracy of a 60" linear measuring device General Measurement Device and Calibration Topics 16
F It is acceptable moving remote locations staff to manufacturing plant for auditing? IATF 16949 - Automotive Quality Systems Standard 3
Q Is it acceptable to mix components from two different lots into an assembly? ISO 13485:2016 - Medical Device Quality Management Systems 1
C Acceptable NDC for %GR&R part inspection to Tolerance (%Tolerance) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
D Risk Register - have we considered enough and is the format acceptable? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
K Operator Checks - How to show that they were completed on a checklist as acceptable IATF 16949 - Automotive Quality Systems Standard 2
C Example Work Instruction/Procedure for AQL (Acceptable Quality Level) AQL - Acceptable Quality Level 4
S Dates on Labels acceptable to the USA - GS1 General Specification 3.4.4 Other US Medical Device Regulations 3
E Root Cause Analysis - Is Insufficient Understanding an acceptable Root Cause? General Auditing Discussions 9
W Unattached (stand-alone) Forms acceptable for AS9100C? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
V Why Gauge R&R 10% is acceptable for variable instruments ? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
X Acceptable methods to store and archive records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Is it acceptable to audit to the Nuclear Principles? Quality Manager and Management Related Issues 9
I Is it acceptable it is to list new products under an existing 510(k)? Other US Medical Device Regulations 14
S If this EN ISO 11137 certificate acceptable for Contract Sterilization? ISO 13485:2016 - Medical Device Quality Management Systems 2
S Is a Dynamic Controlled Form acceptable per 21CFR820 Document Control Systems, Procedures, Forms and Templates 3
C Lowest Acceptable Grade for Master Gage Set Calibration and Metrology Software and Hardware 1
B Digital Signature Solutions acceptable to use in an FDA Medical Device Environment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Is a ndc less than five is also acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is it acceptable to sterilize products without claiming and labeling "sterile" ? ISO 13485:2016 - Medical Device Quality Management Systems 6
M Quality Objectives - Acceptable Levels and Functions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 37
C What is an acceptable bioburden level or count when working in cleanroom ISO class7 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Is embedding documents within a Contract an acceptable practice ? Contract Review Process 9
B Non Acceptable % Tolerance Result - Decrease Sorting Limits Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
F ANSI ASQ Z1.4 - Is the following process acceptable? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C Maximum acceptable Tolerance - Uncertainty Ratio (TUR) General Measurement Device and Calibration Topics 2
S Is it acceptable to have an ASL for Critical suppliers alone? Supplier Quality Assurance and other Supplier Issues 4
S Is this Quality Policy acceptable in the context of ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 24
A Help with acceptable screw gage values per standard BS 919 for Go/ No Go gage General Measurement Device and Calibration Topics 4
S What are typical Acceptable Quality Levels (AQL) adopted in Food industry? Food Safety - ISO 22000, HACCP (21 CFR 120) 1
somashekar Who can provide a WHO GMP certificate in India acceptable to the CDSCO Other Medical Device Regulations World-Wide 6
D Root Cause and Corrective Action - Is our Practice Acceptable? ISO 13485:2016 - Medical Device Quality Management Systems 5
W Application of AQL (Acceptable Quality Level) Concept AQL - Acceptable Quality Level 15
H Acceptable Time to provide proof of CAR (Corrective Action Report) Nonconformance and Corrective Action 4
A Sample Size for Mathematical Model with acceptable Confidence Level Using Minitab Software 7

Similar threads

Top Bottom