What is the difference between "Fitness for Purpose" and "Calibration/Verification"?

placebo_master

Starting to get Involved
#1
I work at a development firm with ISO9001 and AS9100 accreditations (among others). An AS9100 auditor visited us last fall and issued us an NCR with the following evidence:

No defined process found for retaining documented information on fitness for use [i.e. purpose]. Equipment not calibrated is defined in the system and labeled as "Calibration not Required." [Qty.] 217 items with no documented information on fitness for use.

In response, our management addressed the first sentence of the auditor's comment. Our Calibration Procedure now has a clause that dictates the steps to execute a Fitness for Purpose assessment.

Now, I've been assigned the task of developing a plan to evaluate the 217 uncalibrated assets and list them in our master calibration file. (FYI: I do not have previous calibration system experience--I've only ever been the tool abuser user.) Where I'm struggling with this task is how do I differentiate a "Fitness for Purpose" evaluation from a "Calibration/Verification" evaluation?

At face value, Fitness for Purpose necessitates some kind of measurement using the equipment to confirm it produces a valid measurement--yet how is that different from Calibration/Verification? Is the primary difference that we are doing the work ourselves, as opposed to contracting the work to a calibration service with an ISO 17025 certification?
 
Elsmar Forum Sponsor

John C. Abnet

Teacher, sensei, kennari
Leader
Super Moderator
#2
I work at a development firm with ISO9001 and AS9100 accreditations (among others). An AS9100 auditor visited us last fall and issued us an NCR with the following evidence:

At face value, Fitness for Purpose necessitates some kind of measurement using the equipment to confirm it produces a valid measurement--yet how is that different from Calibration/Verification? Is the primary difference that we are doing the work ourselves, as opposed to contracting the work to a calibration service with an ISO 17025 certification?
Good day @placebo_master ;
Consider this...
If an organization is machining and finish grinding a metal surface to within a tolerance of 0.0001", then a yardstick (tape measure) would not be "fit(ness) ---...for its intended purpose" . This holds true EVEN if the yardstick/tape measure are calibrated .

There are very few prescribed requirements in ISO 9001 so your organization has considerable liberty in how this is determined (unless, of course, you have agreements/requirements with/from the customer).

A typical "minimum" (albeit statistically incomplete) method of ensuring "fitness for purpose" is to have a measuring device with a resolution that is a maximum of 1/10 of the total tolerance. Therefore, if a tolerance is 2", the expected resolution of the measuring device would be 0.2".

There are additional more substantial methods via MSA.

Hope this helps.

Be well.
 

placebo_master

Starting to get Involved
#3
This is a helpful start to the discussion. Thank you :)

We develop many things for many companies operating in several industries... suffice to say we have to be flexible with our processes. And I agree that what you described in ISO 9001 grants us flexibility. For example:
A typical "minimum" (albeit statistically incomplete) method of ensuring "fitness for purpose" is to have a measuring device with a resolution that is a maximum of 1/10 of the total tolerance. Therefore, if a tolerance is 2", the expected resolution of the measuring device would be 0.2".
In this example, I would select a measurement instrument for this specific application, then document why I think an instrument with X resolution is acceptable to measure a widget with Y tolerance. The document would then become an artifact I store in the repository of info we maintain for said widget.

What I've been asked to do (i.e. evaluating all of our uncalibrated equipment) seems to conflict with this example you gave. Maybe an example will better explain where my confusion stems from: how can I execute a fitness for purpose evaluation on a power supply if I don't know what it's end use will be? It could be used in a +/- 1 VDC application or a +/- 100 VDC application. IMO it seems like one shouldn't broadly evaluate the power supplies for its fitness, but should instead evaluate it on a case-by-case basis.
 

John C. Abnet

Teacher, sensei, kennari
Leader
Super Moderator
#4
This is a helpful start to the discussion. Thank you :)

We develop many things for many companies operating in several industries... suffice to say we have to be flexible with our processes. And I agree that what you described in ISO 9001 grants us flexibility. For example:


In this example, I would select a measurement instrument for this specific application, then document why I think an instrument with X resolution is acceptable to measure a widget with Y tolerance. The document would then become an artifact I store in the repository of info we maintain for said widget.

What I've been asked to do (i.e. evaluating all of our uncalibrated equipment) seems to conflict with this example you gave. Maybe an example will better explain where my confusion stems from: how can I execute a fitness for purpose evaluation on a power supply if I don't know what it's end use will be? It could be used in a +/- 1 VDC application or a +/- 100 VDC application. IMO it seems like one shouldn't broadly evaluate the power supplies for its fitness, but should instead evaluate it on a case-by-case basis.
As stated, there is no requirement within ISO 9001 as to HOW sufficiency is determined. ISO 9001 heavily (rightly so) focuses on RBT (risk based thinking), therefore, your organization can simply analyze (don't overthink that....since it sounds from your original post that the current method(s) have been in place and apparently ?? working well for some time. [By the way...this begs the question as to why the CB only NOW decided the current approach warranted an NC ??] ) the risk and then apply the measurement device that your organization deems within an acceptable level of risk.

Hope this helps.
Be well.
 

qualprod

Trusted Information Resource
#5
I work at a development firm with ISO9001 and AS9100 accreditations (among others). An AS9100 auditor visited us last fall and issued us an NCR with the following evidence:

No defined process found for retaining documented information on fitness for use [i.e. purpose]. Equipment not calibrated is defined in the system and labeled as "Calibration not Required." [Qty.] 217 items with no documented information on fitness for use.

In response, our management addressed the first sentence of the auditor's comment. Our Calibration Procedure now has a clause that dictates the steps to execute a Fitness for Purpose assessment.

Now, I've been assigned the task of developing a plan to evaluate the 217 uncalibrated assets and list them in our master calibration file. (FYI: I do not have previous calibration system experience--I've only ever been the tool abuser user.) Where I'm struggling with this task is how do I differentiate a "Fitness for Purpose" evaluation from a "Calibration/Verification" evaluation?

At face value, Fitness for Purpose necessitates some kind of measurement using the equipment to confirm it produces a valid measurement--yet how is that different from Calibration/Verification? Is the primary difference that we are doing the work ourselves, as opposed to contracting the work to a calibration service with an ISO 17025 certification?
The first question it means, that what is used is appropriated for that, example, to measure a piece you have to use a caliper or micrometer, can’t be used a flexible tape or a multimeter , the second one, calibration , to ensure your equipment is validated against international standards, and verification to ensure your equipment is able to measure , comparing it against a ( 1 meter, 1 feet, 1 pound) , without the need to calibrate it.
 

dwperron

Trusted Information Resource
#6
Now, I've been assigned the task of developing a plan to evaluate the 217 uncalibrated assets and list them in our master calibration file. (FYI: I do not have previous calibration system experience--I've only ever been the tool abuser user.) Where I'm struggling with this task is how do I differentiate a "Fitness for Purpose" evaluation from a "Calibration/Verification" evaluation?

"Fitness for their purpose" is different from "suitable for the specific type of monitoring and measurement activities being undertaken ". Calibrations show that the tool is "fit" for the purpose of using it to make a measurement, those calibration records are the "documented information" which is required.

What your auditor seems to be questioning is the use of "Equipment not calibrated". That is certainly allowed, but many times an organization fails to differentiate between when calibrated tools are required or not. Your organization is required to:
When measurement traceability is a requirement, or is considered by the organization to be an essential part of providing
confidence in the validity of measurement results, measuring equipment shall be:
a. calibrated


Once the tool is deemed to require calibration it is placed in the "register of the monitoring and measuring equipment".

If your auditor is stating that the 217 tools that you have deemed to not require calibration somehow require that you document their "fitness" then he is wrong. 7.1.5.2 clearly states that not all equipment requires calibration. The two classes of tools need to be treated separately. Your auditor may be questioning your decisions as to which tools require calibration.
 

placebo_master

Starting to get Involved
#7
(don't overthink that....since it sounds from your original post that the current method(s) have been in place and apparently ?? working well for some time. [By the way...this begs the question as to why the CB only NOW decided the current approach warranted an NC ??] )
I didn't mean to give the impression that we have a process that has been working: this conversation is motivated by the fact that, previously, we had no process for evaluating fitness for purpose. We only just recently implemented the process (as a result of the NC the auditor gave us), and since its implementation it hasn't been used by our engineering staff. That's a whole other issue :)

If your auditor is stating that the 217 tools that you have deemed to not require calibration somehow require that you document their "fitness" then he is wrong. 7.1.5.2 clearly states that not all equipment requires calibration. The two classes of tools need to be treated separately. Your auditor may be questioning your decisions as to which tools require calibration.
This is the root of my confusion, so I'm glad you mentioned this. I'm not sure if the auditor is wrong, or if the auditor's message was misconstrued in the CAPA we initiated. I wasn't in the room when it happened, so I don't have first-hand knowledge of what the auditor said. Obviously you won't have the answer either, but this discussion is nonetheless valuable to me because it'll help me better frame the problem with my manager.
 
Thread starter Similar threads Forum Replies Date
A Difference between general PMCF and PMS EU Medical Device Regulations 0
F Difference between "misuse" and "off-label use" EU Medical Device Regulations 9
C Need to understand difference between an actual Calibration vs. a Calibration check General Measurement Device and Calibration Topics 3
H API monogram 4F, The difference between PSL-1 & PSL-2 Oil and Gas Industry Standards and Regulations 5
G What's the difference between 1.32 and 1.33 Cpk? Capability, Accuracy and Stability - Processes, Machines, etc. 18
W What is the difference between TYPE B and TYPE BF? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T The difference between ISO 14644-3:2005 and ISO 14644:2019 Other Medical Device Related Standards 2
Q Terminal Lugs sizes - Difference between 225/24 vs. 275/24 lugs Manufacturing and Related Processes 2
M Difference between "Production Trial Run" and "Run at Rate" IATF 16949 - Automotive Quality Systems Standard 8
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 19
A What is the difference between Design Process, Process Design and Design Control? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R What's the major difference between Green Belt and Black Belt in term of training and project Six Sigma 3
T Difference between a subcontractor and a supplier ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 21
H Difference between Stainless Steel 316 ASTM F899 and ASTM A276 Other Medical Device Related Standards 3
A Exact terms for a plating failure and difference between rejection rate and failure rate Manufacturing and Related Processes 9
M Difference between MSA and MSE? General Measurement Device and Calibration Topics 1
gramps What is the difference between discrete and continuous variables? Problem Solving, Root Cause Fault and Failure Analysis 3
JoCam Difference between Approval and Registration - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
S Difference between EU-MDR Annex IX and the Annex-combo X&XI EU Medical Device Regulations 4
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
Jimmy123 What is the difference between Error Proofing and Controls? ISO/IATF 16949 - Control Plans FMEA and Control Plans 16
C IEC 60601-1-8, difference between table 4 and annex D IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A What is the difference between Basic UDI-DI and UDI-DI? EU Medical Device Regulations 6
Q What is the difference between AS9100D 9.3.2.f and 9.3.3.a AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
M Aluminum - What is the difference between 6061-T6 and 6061-T651 (both per ASTM B211)? Manufacturing and Related Processes 4
P What is the exact difference between Risk and Opportunity in context of ISO 27001? IEC 27001 - Information Security Management Systems (ISMS) 7
S Difference between Surface Finish (Ra) and Flatness (GD&T) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
O Air Flow - Which is the operational difference between LAF (vertical and horizontal) and RLAF? Manufacturing and Related Processes 2
S What the difference is between Stub Acme & Acme thread? Oil and Gas Industry Standards and Regulations 1
S DO 178B - What is the difference between review and verification? Federal Aviation Administration (FAA) Standards and Requirements 1
T The difference between SOP and Kaizen Standardization Lean in Manufacturing and Service Industries 2
K Difference between intended purpose and intended use of the device EU Medical Device Regulations 9
Q What is the difference between normal and licensed internal auditor? VDA Standards - Germany's Automotive Standards 9
J What is the difference between Process Variation and Tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
O Difference Between PFMEA & Control Plan FMEA and Control Plans 3
S Difference between an Advisory Notice (ISO 13485) and a Field Safety Notice? ISO 13485:2016 - Medical Device Quality Management Systems 3
T Difference between "data analysis" and "management review" ISO 13485:2016 - Medical Device Quality Management Systems 4
qualprod What is the difference between 7.4.1 (2008) and 8.4.1 (2015)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Difference between uncertainty and expanded uncertainty of measurement General Measurement Device and Calibration Topics 1
S EASA Part 145 - The difference between non-certifying staff and Certifying staff Federal Aviation Administration (FAA) Standards and Requirements 2
S The difference between a Medical Device Accessory and Component Canada Medical Device Regulations 2
S What is the difference between a service request and a complaint? ISO 13485:2016 - Medical Device Quality Management Systems 2
I What is the difference between OOS/OOT Imported Legacy Blogs 3
A Definition Difference between Quality System Procedure and Standard Operating Procedure (SOP) Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 4
Q Difference between Monitor & Measurement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T QA vs. RA - The difference between QA and RA Coffee Break and Water Cooler Discussions 6
A Difference between Discrimination and Least Count IATF 16949 - Automotive Quality Systems Standard 2