What is the ISO 17025 Certified Laboratory Requirement for External Calibration?

S

Sean Kelley

We are planning to make a change to our procedure and allow calibration of safety and environmetal equipment by non 17025 sources when the equipment is determined not to impact product quality. We will encourage them but in some cases it is difficult to find a certified calibration service for certain devices. Examples of equipment that do not affect product quality and would not fall under this accredited requirement include emission equipment, safety related confined space / atmospheric, monitoring equipment, industrial hygiene sampling equipment and medical testing equipment.

:thanx:
 

AndyN

Moved On
We are planning to make a change to our procedure and allow calibration of safety and environmental equipment by non 17025 sources <snip>

Before you make a blanket statement about being difficult to find a source, have you contacted the accreditation bodies to see if they can help locate a lab? A significant part of calibration of any equipment is knowing how well it was done - hence accreditation. You might want to ask yourself - if getting an accurate result isn't so important, why spend the money with a supplier who can't demonstrate an accurate result? If the labs you are using don't want to be accredited (who knows how many other customers like you are asking?) then make them deliver their uncertainties etc. so you can have some confidence in their results...
 
T

TTE Pipette guy

We are planning to make a change to our procedure and allow calibration of safety and environmental equipment by non 17025 sources <snip>

Hey Sean,

As an accredited pipette calibration lab in XXXXXXX. Be careful of labs accredited to ISO 17025. In summary, ISO 17025 requires labs to have a process and follow it. ISO 17025 can leave labs room enough to design their own PROCEDURES that may or may not comply with the way that certain instruments are to be appropriately and correctly serviced. When shopping for a company, be sure to look over their actual procedures, and choose a reputable company based on this.
 
Last edited by a moderator:

AndyN

Moved On
Hey Sean,

As an accredited pipette calibration lab in XXXXXXX. Be careful of labs accredited to ISO 17025. In summary, ISO 17025 requires labs to have a process and follow it.

To be accredited to ISO/IEC 17025 is more than "say what you do, do what you say"! The lab must demonstrate competency in the discipline. The accreditation auditor is competent/experienced in the calibration/testing done by the lab and must have determined their measurement uncertainty etc.

I can't agree with your statement that they can have a process/procedure and that's all!
 

BradM

Leader
Admin
Before you make a blanket statement about being difficult to find a source, have you contacted the accreditation bodies to see if they can help locate a lab? A significant part of calibration of any equipment is knowing how well it was done - hence accreditation. You might want to ask yourself - if getting an accurate result isn't so important, why spend the money with a supplier who can't demonstrate an accurate result? If the labs you are using don't want to be accredited (who knows how many other customers like you are asking?) then make them deliver their uncertainties etc. so you can have some confidence in their results...

Along with Andy's good suggestion, you can check some accredited calibration sources to see their range of who they have site-sourced. Sometimes they may know of vendors who can perform services that may not be known to you. :)
 

jelly1921

Quite Involved in Discussions
We are planning to make a change to our procedure and allow calibration of safety and environmetal equipment by non 17025 sources when the equipment is determined not to impact product quality. We will encourage them but in some cases it is difficult to find a certified calibration service for certain devices. Examples of equipment that do not affect product quality and would not fall under this accredited requirement include emission equipment, safety related confined space / atmospheric, monitoring equipment, industrial hygiene sampling equipment and medical testing equipment.

:thanx:


If you re-read 7.6, you will find "The organization shall determine the monitoring and measurement to be undertaken and the monitoring and measuring equipment needed to provide evidence of conformity of product to determined requirements"

All sub-clauses under it have the same purpose. I, personally, don't think 7.6.3.2 is a mandatory requriement for EHS equipment, unless your procedure want to. But there are some statutory and regulatory requirements for EHS equipment calibration you have to follow.

Jelly
 

Jen Kirley

Quality and Auditing Expert
Leader
Admin
We are planning to make a change to our procedure and allow calibration of safety and environmetal equipment by non 17025 sources when the equipment is determined not to impact product quality. We will encourage them but in some cases it is difficult to find a certified calibration service for certain devices. Examples of equipment that do not affect product quality and would not fall under this accredited requirement include emission equipment, safety related confined space / atmospheric, monitoring equipment, industrial hygiene sampling equipment and medical testing equipment.

:thanx:
TS 16949 allows using a non-accredited calibration lab if the customer approves it. You could rightfully claim safety system and emission devices are outside the scope of your system under TS16949, but instruments used in your QMS would be expected to follow the accredited lab requirement unless you get approval not to.
 

Hershal

Metrologist-Auditor
Trusted Information Resource
We are planning to make a change to our procedure and allow calibration of safety and environmetal equipment by non 17025 sources when the equipment is determined not to impact product quality. We will encourage them but in some cases it is difficult to find a certified calibration service for certain devices. Examples of equipment that do not affect product quality and would not fall under this accredited requirement include emission equipment, safety related confined space / atmospheric, monitoring equipment, industrial hygiene sampling equipment and medical testing equipment.

:thanx:

Sean, although you may have the requirement for some fields, accredited labs exist in several of the fields listed. However, calibration will typically focus on disciplines, NOT specifics. Therefore, the focus must be on the discipine involved, not specific equipment or even specific application.
 
Thread starter Similar threads Forum Replies Date
D How to handle suppliers that are not ISO 17025 certified ISO 17025 related Discussions 5
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
M We are ISO 9001 certified - Any guide on how to acquire ISO 17025 accreditation? ISO 17025 related Discussions 7
Q Internal Laboratories in TS 16949 companies - Certified vs. ISO 17025? ISO 17025 related Discussions 5
gard2372 ISO 17025 Certified Calibration Laboratory Has Moved Facilities ISO 17025 related Discussions 6
A Definitions in ISO 17025 - Standard reference, Certified reference material, etc. ISO 17025 related Discussions 1
O When do we need to be A2LA/ISO 17025 certified? ISO 17025 related Discussions 6
D ISO/IEC 17025 required for a calibration lab to be ISO certified? ISO 17025 related Discussions 6
V Does an internal calibration laboratory need to be certified on ISO 17025? ISO 17025 related Discussions 3
G ISO 17025.2017 Clause 7.8.4.3 Certificate of Calibration Label ISO 17025 related Discussions 5
GStough ISO 17025 Lead Auditor Training - Punyam Academy ISO 17025 related Discussions 6
K ISO 17025 Query Relating to Monitoring of Equipment ISO 17025 related Discussions 3
A Help required in establishing Laboratory Information Management System as per ISO 17025, Cl.7.11.2 ISO 17025 related Discussions 0
R Do I need to get calibration certificate from ISO 17025 for IATF Audit IATF 16949 - Automotive Quality Systems Standard 8
V ISO 17025 Accreditation Consulting Needed ISO 17025 related Discussions 1
D Is Good Laboratory Practice (GLP) Required in ISO 17025? ISO 17025 related Discussions 4
W Looking for IATF 16949 (and ISO 17025) QMS software Suggestions Quality Tools, Improvement and Analysis 8
R Process Maps for ISO 17025 ISO 17025 related Discussions 2
K ISO 17025 Method Validation and Verification for Test Lab ISO 17025 related Discussions 5
R ISO 17025 Lead Assessor Training ISO 17025 related Discussions 0
J ISO 17025 Calibration Other Medical Device Related Standards 2
E ISO 17025 Key Performance Indicators (KPIs) ISO 17025 related Discussions 5
R Basic Lab Practices - related to ISO 17025 testing labs ISO 17025 related Discussions 1
D Training for ISO 17025 ISO 17025 related Discussions 2
Quality Specialist Quick question on the sequence of ISO 17025 external audits ISO 17025 related Discussions 14
K ISO 17025 Facilities and environmental conditions SOP ISO 17025 related Discussions 6
M ISO 17025:2017 Clause 6.4 - Measuring Equipment Validity ISO 17025 related Discussions 2
B ISO 17025:2017 Clause 7.1.3 - Decision Rule ISO 17025 related Discussions 2
E Organisational Chart-ISO 17025 Laboratory ISO 17025 related Discussions 3
Casana ISO 17025 - Contractor certification? IATF 16949 - Automotive Quality Systems Standard 5
B Sanctioned interpretation #10 - ISO 17025 IATF 16949 - Automotive Quality Systems Standard 2
G Compliance with ISO 9001-2015 for ISO 17025 Accredited Labs? ISO 17025 related Discussions 8
D Management reviews ISO 17025 ISO 17025 related Discussions 3
I ISO 17025:2017 / ANAB 3125 - Articulating / Communicating Risks vis-a-vis Audit Findings ISO 17025 related Discussions 2
W What is okay to use as a reference material for ISO 17025? ISO 17025 related Discussions 4
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 8
P Filled in F48/F49 for internal audit ISO 17025:2017 Internal Auditing 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 4
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4