What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration
Hello All,
Working in production of medical devices we do assembly of purchased subcomponents (to our design).
For this we do the occasional measurment during production. We have our internal master maesurement devices that are calibrated/verified by external labs and we use these verified master tools to verify our own tools. (With our own procedure for that)
Sometimes there are discussions on what we need from certification point of view for our laboratories that calibrate these tools.
Of course the ISO 17025 jumps in immediately.
However we notice that on a lot of the calibration reports the companies confirm that they work according to the ISO 17025 but do not mention they are accreddited or certified (most of they are I believe) so in the past we have had some discussions on the subject. To have them certified according ISO 17025 there will be more costs for the calibratio certificates.
What we can or cannot do basically comes down to what risks there are in using labs that work according but not accredited.
I would like your oppinions in this how we can make a thourough risk analysis that we can be sure we are within the expectations of our field.
The only uncertain aspect might be for incoming goods inspection where we confirm the quality of the delivered products with these measurement devices.
Thanks for your help in this.
John C. Abnet
Teacher, sensei, kennari
Hello All,
Good day Janthero;
As you seem to be aware, there is no requirement in ISO 13485 to utilize laboratories that are accredited (FYI...the term "certification" does not apply to such labs. The correct term for ISO 17025 is indeed 'accreditation') to ISO 17025.
There is, however, the requirement ..."against measurement standards traceable to international or national measurement standards"
I would council that you clearly define the requirements your organization (also considering your customers), NEEDS to mitigate any risks associated with calibration and then hold your organization nd the providers accountable to those requirements.
Also, ensure any calibration service requirements applicable regulatory governance of your organization may specify.
Hope this helps.
Be well.
dwperron
I can give you an example of where accredited calibrations can make a difference. It depends on who you need to comply with, who will be auditing you.
I have worked with companies who were seeking UL approval for their products. UL does not require 17025 accreditation of the test and measuring equipment being used, but they are very specific that if you choose not to get accredited calibrations (very different from using a 17025 accredited lab or a 17025 "compliant" lab) then you will need to be able to demonstrate traceability to a National Lab, such as NIST in the United States. That will require you to demonstrate records of all calibrations in the chain from your equipment through every calibration instrument used in each lab for your calibration - including measurement uncertainties - then continuing this process all the way to NIST through every calibration lab involved for every piece of equipment that is part of the reverse traceability for your instruments.
Not impossible, but it can be a daunting task.
If you have 17025 accredited calibrations then it is automatically assumed to have a valid traceability chain back to NIST.
A Operational Risk - Quality system procedure should cover all eventualities Misc. Quality Assurance and Business Systems Related Topics 8
B Risk Management Procedure updates needed for 14971:2019 ISO 14971 - Medical Device Risk Management 5
M What is the Risk of Using Obsolete Versions of C=0 & ANSI/ ASQ Z1.4 Sampling Plans? ISO 13485:2016 - Medical Device Quality Management Systems 8
D AS9100D 8.4.2 Note 2 Significant Operational Risk AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
A Calculating Risk Estimation ISO 14971 - Medical Device Risk Management 28
M Intended Use vs Actual Use and Scope of Risk Management EU Medical Device Regulations 8
S IDCB 0129/0160 Clinical Risk Management ISO 14971 - Medical Device Risk Management 2
H At what level (harm, hazardous situation, seq. of events, etc) is "risk" estimated? ISO 14971 - Medical Device Risk Management 12
A Risk Management Team IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
S Risk Management File - Procedure Packs ISO 14971 - Medical Device Risk Management 3
B ISO 14001 Risk assesment ISO 14001:2015 Specific Discussions 4
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
S What is your favorite Usability Risk Analysis tool? IEC 62366 - Medical Device Usability Engineering 5
T Assessing risk where harm is indirect - Generic devices / accessories / intermediates ISO 14971 - Medical Device Risk Management 8
K Do you have separate clinical risk management group or experts in your manufactures? EU Medical Device Regulations 4
W IATF 9.2.2.1 Internal Audit how to determine risk IATF 16949 - Automotive Quality Systems Standard 12
S Risk control through Information for safety ISO 14971 - Medical Device Risk Management 8
A Derive Risk Acceptance Matrix from Risk Policy ISO 14971 - Medical Device Risk Management 8
B ERP software validation - risk assessment vs validation scope ISO 13485:2016 - Medical Device Quality Management Systems 11
I Estimation of overall residual risk. How to? EU Medical Device Regulations 11
Sidney Vianna ISO Practical Guide on ISO 31000:2018 - Risk Management Other ISO and International Standards and European Regulations 0
T IEC 62304 : Risk control for SaMD IEC 62304 - Medical Device Software Life Cycle Processes 8
T Risk Assessment and Management Misc. Quality Assurance and Business Systems Related Topics 0
P Scenario based risk assessment IEC 27001 - Information Security Management Systems (ISMS) 1
Q KPI risk assessment - Criteria for the given score IATF 16949 - Automotive Quality Systems Standard 3
S Foreign Risk Notification Canada Medical Device Regulations 2
J HELP NEEDED ! Risk Management Exercise ISO 14971 - Medical Device Risk Management 12
O Should a Covid vaccine and testing policy be included as part of ISO9001 or AS9100 risk management? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
M Does 4.5 - Alternative RISK CONTROL apply to the Particular Standards? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Q Measurement Equipment Revocation - Looking for a Disposal Form with Risk Assessment IATF 16949 - Automotive Quality Systems Standard 10
B ISO13485 Risk managment implementation for suppliers ISO 14971 - Medical Device Risk Management 2
Moncia Chemical risk assessment / COSHH Manufacturing and Related Processes 5
Enghabashy Supply chain main policies ,scope, risk assessments & relavant KPI Supply Chain Security Management Systems 2
D Use Error Risk Controls and Control Verification ISO 14971 - Medical Device Risk Management 6
J Risk Assessment of Lithium Ion Batteries FMEA and Control Plans 3
Melissa Risk Management Process, How far do I need to go? ISO 14971 - Medical Device Risk Management 13
D Does Risk Management apply to re-labeler (MDR) EU Medical Device Regulations 1
H Risk Management Plan in agile process ISO 14971 - Medical Device Risk Management 14
H Risk Analysis and Probability of Occurrence ISO 14971 - Medical Device Risk Management 3
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
P Benefit risk analysis on pFMEA ISO 14971 - Medical Device Risk Management 10
B AS9102 - 3D printing a special tool required for assembly (counterfeit risk?) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
K Defining risk control measures IEC 62304 - Medical Device Software Life Cycle Processes 14
U Supply risk management Manufacturing and Related Processes 4
T Biological Evaluation (10993) & Risk Management ISO 14971 - Medical Device Risk Management 9
D Cybersecurity and Risk Management: Loss of confidentiality IEC 62304 - Medical Device Software Life Cycle Processes 5
Q FMEA and Risk assessment in Microsoft Access FMEA and Control Plans 6
I Realization processes input into overall risk ISO 14971 - Medical Device Risk Management 2
thisby_ Post Market/Production Risk Assessment ISO 14971 - Medical Device Risk Management 0