that's great if it works for you.
Unfortunately I've had the opposite experience where people would feud via the CAR system.
Yup, I've seen exactly the same garbage in other organizations, particularly one specific defense contractor I once worked for. It was to the point where someone would write a CAR just so that the subsequent blamestorming would distract attention from their own deficiencies. Fortunately that sort of crap isn't tolerated where I now work.

Regards -John
 
Elsmar Forum Sponsor
L

lrowe

#12
I am going through a complete retooling of the CAPA process /procedure with my new company now. I've been here a little over a month and what I found here was Deja Vu from my previous company when I first took over ISO responsibilities with them. In short - if you don't have a fairly well defined CA criteria, people will call just about anything they don’t like a cause for "Corrective Action". This leads to a system that becomes over whelming in a hurry and CA are open for a long time, since some of the problems won't necessarily have a root cause or effective action plans.

It's my contention that if you use the term "Corrective Action" this indicates a problem that is serious in either it's cost, your company's reputation, customer satisfaction, etc. As such, this means there will be substantial resources applied to fix the problem. Other problems that have less impact should be called something else, like an internal Non conformance, Supplier reject, customer complaint etc. These types of problems are processed differently, in other words appropriate actions that won't require the intensity of the resources for a "real" corrective action.

So in a nutshell, I like to define the situations that warrant a true corrective action and make sure that definition is not so vague that any problem is processed as a corrective action. This makes the system much more manageable and has passed audit muster several times.

Larry
 
T

Tim Butler

#13
John,

That is a problem. If the QM and Management Team does not review these NCRs and the resulting ( CA/PA ) for:

1.) Accuracy - Internal Auditors should be well trained
2.) How and why it was found. If anyone can write a N/C ( CA/PA ), this could very well happen. Control who can write the N/C, allow anyone to issue a request to investigate the possible N/C.

If the procedure is that the QM and their (I.A.s ) Internal Auditors must investigate the NCR prior to issuance - this can prevent useless findings from being written and prevent the issue of employee bashing.

It is the QM and the Executive Management, that sets the tone and how N/Cs are written against the QMS. There are, in the majority of cases, good findings and I think it best, not to lose sight of the benfit from finding N/Cs and implementing ( CA/PA ) before the customers do. Provided they are valid.

I will agree that in some cases, there is a blame game that goes on and people use the N/C as a means of political ammo. This hurts the QMS and its' validity.

If this occurs during the Internal Audits by the internal auditors, it is up to the QM to prevent this. I have seen in somes cases where the QM didn't have the ability to prevent it and as a result, outsourced the I.A. to a third party to maintain objectivity.


Sincerely,
Tim

JFount,

My name is Tim Butler and I have 10 years of experience and have worked with 1000+ companies on their ISO programs: AS/ISO/EMS/CE/OHSAS/TL/RCMS/TS.

If you are looking for what should trigger a CAPA look to your companies' procedure and if there is a deviation in this procedure from what actually occurred, then a CAPA should be required. I would also make sure the procedure is lean - calling out only what is absolutely required to perform the specific task. Otherwise, you are requiring additional steps that could bog your QM system. If you need a checklist for any standard or anything else, please feel free to ask.

Sincerely,
Tim
 
Last edited by a moderator:
J

John Nabors - 2009

#14
Tim-

An earlier post places my second post in context-

...The CAR or PAR is reviewed by the managers of Quality, Production, Materials, Engineering and Maintenence. If the CAR is found to be inappropriate or spurious it is rescinded...
Regards -John
 
J

justncredible

#15
A CA should address the outliers and sporadic change.

A PA should address the normal occurring process waste.

Pa goes with a pareto for known causes of waste or scrap.

CA goes with a system breakdown, a flier, a outlier, a non-conformance that does not fall into known waste related issues.

The PA will eliminate process variation, based on knowns. The CA is a remedy for a unknown. Cost of quality analysis should be studied for a PA, whereas a CA involves customer risk. So the CA should be a weighted, and customer risk driven action.

But, then I just make this stuff up as I go along.........
 
R

Roland Cooke

#16
"The most important preventive action we ever took was to implement a Corrective Action system, just in the unlikely event that our up-until-now fully effective Preventive Action system let us down at any future point."


Joking aside, I really would recommend breaking out the CA system from the PA system. Don't even call it CAPA if you can get away with something that radical.

Anything that happens following a problem to stop it or something similar happening again is further corrective action.

There is an argument that lessons that can be applied to unrelated areas would be preventive actions. I do tend to subscribe to that theory, but from my perspective, I don't care what you call it as long as it gets done.


Most companies have a lot of Preventive Activity going on - they just don't recognise it as such. Action/alert levels on trending data, contingency planning, safeguards built in for larger / longer-term projects, etc etc.

Analysis (e.g. by Internal Audit) of why projects went bad, will usually indicate that the preventive action system wasn't effective.
 
P

Phil Fields

#17
Is you company FDA regulated? If yes then follow 21 CFR Part 820 section 820.100 a (1). This section defines the areas that would trigger a CAPA
 
Thread starter Similar threads Forum Replies Date
A Should we assign the PRRC before the date of application of MDR (26 May 2021)? EU Medical Device Regulations 0
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Should I take an online course for a career in Occupational Health and Safety? Career and Occupation Discussions 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
K Should APQP/PPAP has its own section in a QM? Quality Management System (QMS) Manuals 1
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
P Should eIFU link per ISO 15223-1:2016 be added to labels out of scope of Reg 207/2012? EU Medical Device Regulations 1
S Which Sampling Plan(s) Should I Use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
A Document release vs its related training. Which should come first? ISO 13485:2016 - Medical Device Quality Management Systems 18
S Which department should prepare the control plan? could you show me a standard regarding to this matter. FMEA and Control Plans 17
J Help settle a disagreement: Should external providers of preventive maintenance be on your ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Master Samples - What should we be keeping? IATF 16949 - Automotive Quality Systems Standard 9
G Supplier delivered recent PPAP, should he deliver yearly layout inspection? IATF 16949 - Automotive Quality Systems Standard 4
John Broomfield Vote - Should ISO9004 Become a Requirements Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
A Capability Study - in the beginning of your career what should you have known about the tool Quality Tools, Improvement and Analysis 11
J Should Loading and Unloading be Included in Cycle Times? Lean in Manufacturing and Service Industries 14
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0
T 510(k) submission - Which name should I use in the submission? Other US Medical Device Regulations 3
N ISO 19011:2018 - 5.4.2 "...audit program should engage in appropriate continual development..." Training - Internal, External, Online and Distance Learning 4
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
DuncanGibbons Should the requirements FAA/EASA Part 21 be addressed within the QMS and AS9100D quality manual? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Should 510(k) Predicates be Actively Listed Devices? Other US Medical Device Regulations 12
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
J On PFMEA for danger labels - Label always should be assigned severity 10 ? FMEA and Control Plans 3
H Who should be listed as the manufacturer/distributor on the box? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 3
BeaBea How Many Processes should be created for each Department? Process Maps, Process Mapping and Turtle Diagrams 5
M Should volume of sales be factored into risk probability assessments? ISO 14971 - Medical Device Risk Management 33
MrTetris Should potential bugs be considered in software risk analysis? ISO 14971 - Medical Device Risk Management 5
S Should safety checks be included in the Control Plan? IATF 16949 - Automotive Quality Systems Standard 5
M Which incubation condition should be selected to recover both bacteria and fungus effectively Miscellaneous Environmental Standards and EMS Related Discussions 3
D Is there a specific location for PPE such as safety glass holders and glove dispensers should be mounted Occupational Health & Safety Management Standards 10
Robert Stanley Which Registrar Should I Choose for ISO 9001:2015 registration? Registrars and Notified Bodies 10
M Who should receive the bills from suppliers and vendors, account payable or procurement? Consultants and Consulting 4
V IATF 16949 8.4.1 Control of externally provided processes, products and services - Should the CB be on our Approved Supplier List? IATF 16949 - Automotive Quality Systems Standard 10
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Received a Major finding during IATF Surveillance audit for loss of BIQS Level 3 (more than 6 SPPS in 6 months)...how should we address SYSTEMIC CA? IATF 16949 - Automotive Quality Systems Standard 11
J Organization merger. Should we keep two separate ISO 13485 certificates? ISO 13485:2016 - Medical Device Quality Management Systems 6
S Companies that maintain your machine should be in ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Use of "Shall" versus "Should" in Procedures ISO 13485:2016 - Medical Device Quality Management Systems 26
D Class II medical device - When should a complaint be closed? Customer Complaints 6
Sidney Vianna IATF 16949 News Presentations from the latest IATF Stakeholder Event - Expectation that IATF 16949 certification should equate with product quality. Misguided? IATF 16949 - Automotive Quality Systems Standard 7
L Clause 0.4 of ISO 9001 and EHS - Where should I stop the inclusion of EHS in my QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Ed Panek Part 11 Self Certify Memo - What else should it cover? Qualification and Validation (including 21 CFR Part 11) 5
H Should I mention machine/Equipment password In SOP? Qualification and Validation (including 21 CFR Part 11) 4
D How long should we keep the spare parts available for our medical device, after we have stopped the production? ISO 13485:2016 - Medical Device Quality Management Systems 0
H Statistical Techniques Procedure - What should be included Document Control Systems, Procedures, Forms and Templates 4
Q How should I analyze measurement correlation between me and customer? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Sidney Vianna Interesting Discussion ISO 9001:2024 - What should be changed in the next Edition of ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 114

Similar threads

Top Bottom