I know that this topic has been touched on before but I haven't seen a difinitive answer on exactly what devices ISO 9001 and 13485 require to be controlled to the level of the standards. We are getting inundated with requests to calibrate equipment and I have been commisiond to define the criteria for determing which devices to include in the control system. Please comment on the following statements. Agree, disagree or am I all wet?
Control of "monitoring and measuring devices" is required by ISO standards and FDA regulations. ISO 9001:2000 and 13485:2003 require "devices needed to provide evidence of conformity of product to determined requirements" be identified and controlled to ensure valid measurement results. The CFR Title 21, Part 820, Subpart G, Section 820.72 stipulates "inspection, measuring, and test equipment" be "suitable for its intended purpose" and "capable of producing valid results." Both ISO Standards and FDA Regulations specify equipment calibration as a component of the control system and outline the necessary requirements to ensure the device measurement system achieves the intended purpose.
ISO, in previous releases of 9001, specified IMT&E "used by the supplier to demonstrate the conformance of product to the specified requirements" must be controlled, and periodic calibrations performed. ANSI/ASQC Q9001-1994 element 4.11 CONTROL OF INSPECTION, MEASURING, AND TEST EQUIPMENT, directs the supplier to refer to ISO 100012 QUALITY ASSURANCE REQUIREMENTS OF MEASURING EQUIPMENT for guidance implementing a control system. ISO 100012 is very specific in scope regarding measuring equipment. It is applicable to measuring equipment "used in the demonstration of compliance with a specification"… "it does not apply to other items of measuring equipment".
FDA has provided guidance for implementing a calibration program in Chapter 7 EQUIPMENT AND CALIBRATION, of HHS Publication FDA 97-4179 MEDICAL DEVICE QUALITY SYSTEMS MANUAL: A SMALL ENTITY COMPLIANCE GUIDE. The Guide indicates that "Some instruments need only to be checked to see that their performance is within specified limits, while others may require extensive calibration to a specification." It continues "Sometimes equipment used only for monitoring a parameter need not be calibrated but should be identified (e.g., for monitoring)."
These clarifying statements by FDA along with those by ISO, indicate that:
a. Only devices used to demonstrate product conformance must be controlled to the level outlined in the standard/regulations.
b.Other devices may be checked, verified, or tested based on utilization.
Your comments please,
Jim
Jim