What specifically must be calibrated per ISO 9001:2000?

Eredhel

Quality Manager
#61
You've opened up the fun "steel rule and tape measure debate" :). We reference only ours unless they're used for "final acceptance" (might be a term you're looking for). But we tend to avoid doing some things unless we absolutely have to. And so far we've never had a complaint with reference only for in-process checks. We've thought about getting a master steel rule to set up somewhere in the future, but we haven't needed to yet.
 
Elsmar Forum Sponsor
T

Tyler C

#62
Yes, I know this can be quite a difficult debate. And "Final Acceptance" was exactly what I was looking for! Thank you!

You have confirmed my thought on this. We are also looking at "reference only" for any in process checks (with a few exceptions). This is a heated debate in my company as well, should we or should we not get a "golden" ruler. We have finally settled on a "golden caliper" to use to verify other calipers with a gauge block, and we can use it to verify tapes, rules, etc.

Glad to hear you have not had any issues with this.
 

Eredhel

Quality Manager
#63
We do calibrate/verify, different language argument there :), our calipers. We keep standards that are sent out yearly to a calibration lab and then do all our own in house checking of mics, calipers, et cetera. Our CMM, Comparator, and surface plates may be the only things done by an outside party.
 

THamlett

Starting to get Involved
#64
I went through this several years ago with a Quality Director who told us that every micrometer,caliper, gauge block, pin gauge, pressure gauge, clock, etc in the company had to be calibrated and traceable to NIST. We told him he was nuts. Only the devices that had a direct effect on product quality needed to be calibrated or if a specification required it. We had an air compressor feeding blow-off guns that had a pressure regulator and gauge but the gauge was a reference only type of device. No calibration needed.

My rule is, if a device reading is recorded on a document that some one signs, then the device needs to be calibrated.

Get yourself a lot of "Reference Only" Stickers.
 
#65
Get yourself a lot of "Reference Only" Stickers.
Better to get "Calibration Not Required". It confuses people to use "Reference Only". Why would anyone reference something if it can't be trusted?

ref·er·ence
ˈref(ə)rəns/Submit
noun
1.
the action of mentioning or alluding to something.
"he made reference to the enormous power of the mass media"
synonyms: mention of, allusion to, comment on, remark about
"his journal contains many references to railroads"
2.
use of a source of information in order to ascertain something.
"popular works of reference"
 
L

ljreimer

#66
I tell clients during audits if they are using rulers for a measuring device for acceptance of product it should be calibrated once to verify that the cardinal markings on the ruler is correct. Future calibration is not required, only verify that that the device is still in working condition. Recalibration is not adding any value since the ruler will not produce inaccurate readings in the future.

Any monitoring or measuring device that is used to made decision to accept product at different stages of the manufacturing process needs to be calibrated if the company determines a risk in meeting customer requirements. If only calibrated devices are used at final inspection that meets the intent of the standard. However, Risk base thinking, if during the manufacturing processes inspection activities are conducted and specific product characteristics are only reviewed during these stages and not at final that could be issue if the measuring device is not calibrated.

It should also be taken into consideration the risk of identifying nonconforming product at the final stages, This could impact delivery and order fulfillment. if more effective inspection program with calibrated devices are use at earlier stages of the manufacturing process, identifying nonconforming product earlier into the manufacturing stages saves time and money.
 
#67
I tell clients during audits if they are using rulers for a measuring device for acceptance of product it should be calibrated once to verify that the cardinal markings on the ruler is correct. Future calibration is not required, only verify that that the device is still in working condition. Recalibration is not adding any value since the ruler will not produce inaccurate readings in the future.

Any monitoring or measuring device that is used to made decision to accept product at different stages of the manufacturing process needs to be calibrated if the company determines a risk in meeting customer requirements. If only calibrated devices are used at final inspection that meets the intent of the standard. However, Risk base thinking, if during the manufacturing processes inspection activities are conducted and specific product characteristics are only reviewed during these stages and not at final that could be issue if the measuring device is not calibrated.

It should also be taken into consideration the risk of identifying nonconforming product at the final stages, This could impact delivery and order fulfillment. if more effective inspection program with calibrated devices are use at earlier stages of the manufacturing process, identifying nonconforming product earlier into the manufacturing stages saves time and money.
Welcome! Could you clarify which type of audit you do, when you advise your clients - some visitors here may not realize that different "types" of auditors shouldn't be offering advice.

I understand the standard to be about prevention so I'm not certain your statement; "If only calibrated devices are used at final inspection that meets the intent of the standard." is accurate. Waiting until final to discover parts were made wrong, when the operator was using a worn-out gauge, isn't prevention.
 

howste

Thaumaturge
Super Moderator
#68
I don't know why anyone would ever measure something if there was doubt if the measuring equipment was reliable. If we don't care about in-process measurements being reliable then why should we even bother to do them?
 

Golfman25

Trusted Information Resource
#69
I don't know why anyone would ever measure something if there was doubt if the measuring equipment was reliable. If we don't care about in-process measurements being reliable then why should we even bother to do them?
I don't think it is necessarily about reliability but trying to manage a whole lot of gages and avoiding unnecessary dings from auditors. We have all been there, on the shop floor, trying to run some part, when "Johnny" pulls out of his toolbox some gage nobody has seen in years to measure the feature. He checks it, makes a process adjustment, checks again, and hands the part off to QA for buyoff. Part is approved and production starts.

Should we make sure Johnny's gage is calibrated? Probably. Is it absolutely necessary? No, given that the buyoff is actually done by QA.

We have many gages that are calibrated, but essentially reference, on our shop floor. During setup, they will use a caliper to get a "looks good" measurement before handing off to QA who has the final say. QA will use a CMM to verify and approve. Is it a good use of resources to manage all of the "reference" gages thru calibration? Each company will have to make that decision.
 

Rich Shippy

Starting to get Involved
#70
Just an observation, 7.1.5.2 a gives the option of "calibrated OR verified OR both", verification is much quicker than calibrating on a lot of measuring equipment. So technically for compliance, it would appear verification is acceptable.
 
Thread starter Similar threads Forum Replies Date
M Labelling - Is there a guidance or regulations that specifically defines the 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Question regarding ECO process, specifically for Life Science products and defining form fit and function ISO 13485:2016 - Medical Device Quality Management Systems 1
N ISO 22716 Cosmetics - Hygiene (hair restraints specifically) Other ISO and International Standards and European Regulations 2
G Procedure Creation specifically for dimensional (caliper, micrometer, dti) Document Control Systems, Procedures, Forms and Templates 8
X Customer Service FMEA - Specifically Customer Care / Severity Rankings Service Industry Specific Topics 3
Q NADCAP NDT (specifically RT) - Calibration of eye loupe reticle & optical comparator AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
Q NADCAP NDT (specifically RT) - Checklist AC7114 Rev. K, section 8.1.3 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Plating Shop - Forums specifically related to the plating / metal finishing industry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D UV Absorption Test - What is it's purpose and what specifically is it looking for? Other Medical Device Related Standards 9
T Specifically calling out the use of lower level documents in higher ones. IATF 16949 - Automotive Quality Systems Standard 16
P Electronic Record Control (Specifically editable records not fixed documents!) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Help needed for Fighting a specifically identified Organism found during Dose Audit Miscellaneous Environmental Standards and EMS Related Discussions 2
D ISO 13485 Help - Determining what processes I specifically need to document ISO 13485:2016 - Medical Device Quality Management Systems 9
N Customer Property Logs - Tracking customer property, specifically production tooling IATF 16949 - Automotive Quality Systems Standard 5
Q Are there any threads specifically concerned with designing with AutoCAD? Design and Development of Products and Processes 1
C Documentation of training and retraining procedures, specifically how training should Document Control Systems, Procedures, Forms and Templates 6
G AMS 2750 - Tell me what troubles, specifically, you have had with AMS 2750 Various Other Specifications, Standards, and related Requirements 4
S Method Validation for qualitative tests (specifically Absence/Presence) General Measurement Device and Calibration Topics 2
T Non-Conformance Reports - Does the standard specifically say what an NCR is called? Nonconformance and Corrective Action 28
Tim Folkerts NCAA football polls, BCS, etc - Specifically the USA Today Coaches Poll Coffee Break and Water Cooler Discussions 8
CarolX Does ISO 9001:2000 Specifically Require Process Maps? Process Maps, Process Mapping and Turtle Diagrams 19
S Corrective Actions - Specifically closing out and verifying effectiveness Nonconformance and Corrective Action 30
D Gage R & R - Do we specifically have to do Gage R&R on every gauge? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
A What must be recorded? (ISO 9001:2015, subclause 10.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
John C. Abnet Must COPQ always be quantified as a monetary ($) amount? IATF 16949 - Automotive Quality Systems Standard 2
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
B IATF16949 audit requirement - Auditor request UCL and LCL must be show Xbar-R, IATF 16949 - Automotive Quality Systems Standard 7
A Read instruction manual - What Graphic Symbol must I use? US Food and Drug Administration (FDA) 7
M What kind of instrument that must be calibrated? General Measurement Device and Calibration Topics 3
Q Must product name be listed the same name in FURLS, UDI, GUDID and Company Website? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J Customer VDA Audit - We must provide refresher training? VDA Standards - Germany's Automotive Standards 4
J Any diverging opinions regarding audit findings or conclusions must be resolved no later than the closing meeting ISO 13485:2016 - Medical Device Quality Management Systems 13
I Sampling processes - Who must define the AQL level? AQL - Acceptable Quality Level 9
M Which documents must be updated upon product validation? Document Control Systems, Procedures, Forms and Templates 1
Q IATF 16949 - Product audit - Must each product be audited? IATF 16949 - Automotive Quality Systems Standard 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
R Changing Document Control software. Must I transfer EVERYTHING? Document Control Systems, Procedures, Forms and Templates 3
T Must all Disposable Medical Devices be DEHP Free for MDR? EU Medical Device Regulations 1
W Product Audit - It must be a separate and distinct activity (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 4
P Quality objectives - must they include CAPA and internal audit topic? ISO 13485:2016 - Medical Device Quality Management Systems 28
S Record Retention - How long must a company keep the following records? Records and Data - Quality, Legal and Other Evidence 17
I Is risk acceptability really needed if all risks must be reduced as far as possible? ISO 14971 - Medical Device Risk Management 6
A ISO 9001:2015 Clause 9.3.2 - Management Review Inputs must be Documented? Management Review Meetings and related Processes 15
N Must Bioburden Sampling Procedures be Documented? Other Medical Device Related Standards 5
M Contacting a "Must Use" (aka Sole Source) Supplier's Registrar Supplier Quality Assurance and other Supplier Issues 6
T Auditor states my Oven Thermocouples must be Calibrated General Measurement Device and Calibration Topics 18
K Must I update the Quality Policy? (ISO 9001:2008) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
B Must a Design Outsourcing Company have ISO13485? ISO 13485:2016 - Medical Device Quality Management Systems 3
U Must every procedure be proceeded by a policy? Document Control Systems, Procedures, Forms and Templates 2

Similar threads

Top Bottom