Good day,
I am hoping someone can help explain exactly what the impact is of %GRR/TV
being at such a high percent? This gage source has historically had issues supplying GRR data with product, and I am having trouble getting the gages to pass a gage r&r in-house using their form.
It is a simple(?) load a part, check with an indicator in one spot fixture.
It seems to datum ok, but a gage R (1x10) fails after several attempts.
Thanks for the help.
Danny
 

Attachments

Elsmar Forum Sponsor

Miner

Forum Moderator
Staff member
Admin
#2
To what use will this gauge be applied? It is acceptable for use as an inspection gauge. It does not appear to be acceptable for use in SPC or for statistical studies such as capability. However, this last evaluation is highly dependent on the ten parts used in the study. If these parts are representative of the actual process variation, the results are accurate, if they do not represent actual process variation (e.g., they were 10 consecutive parts) then the results may be greater than actual.
 

Danny Hoover

Involved In Discussions
#3
Thank you Miner,
The gage is to be used to collect on-going SPC data for capability during production runs, one point to monitor length. My biggest trouble I have with these "borderline" as received studies is attempting to resolve the dispute when my Lab team can't get the same, or passing results. I suspect this may have been one part checked 2x2x10x, which is why I am suspicious of the high %GRR/TV. I need to repeat and pass these studies from a production run tomorrow and I am trying to get a easy way to explain to the engineer what is happening.
I suspect many times the samples I get for layout are consecutive, not 30/300 or more, and we have stability issues when running a 4th edition MSA study.
Any help is appreciated.
Thanks,
Danny
 
B

Barbara B

#4
The part could be the same for all values or the parts are too similar or the measurement system uncertainty (GRR%) is too high. Nothing within the data could answer the question why the GRR%(study variation) is big, it could only be said that GRR% is too high compared to study variation.

Attached you'll find a Gage R&R (crossed) analysis done in Minitab which provides additional informations about the similarity of values (gage run chart), the GRR% compared to study variation and tolerance width as well as a GRR% compared to a ficititous historical process standard deviation of 0.2 (std dev in Gage R&R study: 0.3).

GRR%(study variation): 100% -> way too high
GRR%(tolerance width): 11.1% -> good
GRR%(ficititious process standard variation): 16.6% -> acceptable

Do you have informations about the usual process standard variation (aka historical standard variation)? To evaluate whether the part selection for the Gage R&R was made properly you could compare study variation (here: 0.03) and historical standard variation. They should be roughly similar if the part selection was made according to MSA requirements.
 

Attachments

Danny Hoover

Involved In Discussions
#5
Thank you Barbara for the information.
The gage I am asking about is new, going to be used today on a 300 pc run sample to validate GRR using the SPC port and supplying that data as part of the PPAP submission.
We are not the most strict on part collection during run-offs, I suspect most may be 30 pcs consecutive(I am not invited to the run-off). My stake is to prove the gage is repeatable and accurate for data collection, and I feel many times it may not be.
I have been digesting all the theory I can on the MSA requirements and best practices, the Cove is a great resource.
Thanks for all the help!
Danny
 
D

Dtwo5

#6
Good day,
I am hoping someone can help explain exactly what the impact is of %GRR/TV
being at such a high percent? This gage source has historically had issues supplying GRR data with product, and I am having trouble getting the gages to pass a gage r&r in-house using their form.
It is a simple(?) load a part, check with an indicator in one spot fixture.
It seems to datum ok, but a gage R (1x10) fails after several attempts.
Thanks for the help.
Danny
Hello Danny - I think "using their form" is the key here. I have the AIAG format on disk and I can assure you this has been modified.
When I get a moment I will run it using your numbers. What is the upper and lower specification? My AIAG format does not have "Total Tolerance".
 

Danny Hoover

Involved In Discussions
#7
Dtwo5,
The total tolerance is ?0.9mm, for a 1.8 total range. I have dropped this data in a couple different forms, but I have found out is really a check of 1 part 10x by 2 operators show that is part of the issue.
Thanks for the reply, I am trying to hammer out details for future requirements.
Danny
 
Thread starter Similar threads Forum Replies Date
M Defining High/Medium Risk Impact Misc. Quality Assurance and Business Systems Related Topics 11
Judy Abbott The impact of laser power on biocompatibility of the products Manufacturing and Related Processes 6
R IATF 16949 Certification for new site with transferred product--what is the impact with CSR and scorecards? IATF 16949 - Automotive Quality Systems Standard 0
V Laser Welding Process - Impact on Electrical Properties Reliability Analysis - Predictions, Testing and Standards 4
L Impact on 510(k) to use an additional manufacture US Food and Drug Administration (FDA) 0
R Impact of changing manufacturer post clinical investigation CE Marking (Conformité Européene) / CB Scheme 3
Ed Panek Corona Virus impact on Supplier Audits and Received Parts ISO 13485:2016 - Medical Device Quality Management Systems 4
F Will the EU postpone 2017/745 MDR Implementation - COVID-19 Impact Medical Device and FDA Regulations and Standards News 11
R Revalidation machine line and Stability Studies impact. Medical Device and FDA Regulations and Standards News 3
M MDR Impact on MDSAP Countries Other Medical Device Regulations World-Wide 16
Watchcat Impact of MDR on Availability of Medical Devices in EU EU Medical Device Regulations 5
Judy Abbott Can any one help me with an urgent BIOCOMPATIBILITY Impact assessment? Internal Auditing 9
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-1 Impact test on glass touch controller IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
qualprod The Perfect audit? External Audit causes a significant negative impact in a company General Auditing Discussions 9
Ashok sunder Is it possible to reduce Risk likelihood and impact Post control Ranking after corrective action taken for risk? FMEA and Control Plans 1
E MDR Article 117 / Impact to New Marketing Authorization Applications Other Medical Device Regulations World-Wide 1
C Lithium battery powered Medical Device - 15.3.3 Impact test, broken enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Impact of an MDSAP certificate with all countries coverage Other Medical Device Regulations World-Wide 8
Edward Reesor Canadian News article on impact of MDSAP ISO 13485:2016 - Medical Device Quality Management Systems 10
qualprod Propose Aspects to evaluate risk impact? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
qualprod Risk impact can be modified? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
R Brexit - European Commission - The impact of the Brexit per March 29, 2019 EU Medical Device Regulations 3
B The impact of IoT in ISO documentation IT (Information Technology) Service Management 1
J Business Impact Analysis SOP Business Continuity & Resiliency Planning (BCRP) 2
T GDPR impact on ISO 9001 and Quality ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Raffy What is the first step in doing PIA (Privacy Impact Assessment)? IEC 27001 - Information Security Management Systems (ISMS) 3
Sidney Vianna IATF 16949 Areas of Impact - Rules 5th Edition Document (February 2017) IATF 16949 - Automotive Quality Systems Standard 0
I Biocompatibility Impact on Autoclaved Silicone Other Medical Device Related Standards 3
R Tensile Test Strain Rate - Impact on the Test Results Manufacturing and Related Processes 3
K Thoughts on the impact of the General Data Protection Regulation? Medical Information Technology, Medical Software and Health Informatics 5
Q Risk Impact - Risk Assessment Sample/Method per ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
A Difference between Environmental Impact and Risk ISO 14001:2015 Specific Discussions 7
M 15.3.3 Ball Impact Test requirements in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Calculating impact force, expressed in lbs, of a 5 lb weight dropped 48 inches Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
J Charpy impact test - Mill cert for API 13CrL80 coupling stock material with is PLS 2 Oil and Gas Industry Standards and Regulations 1
K TS 16949 Planning Phase: Raw Materials purchased from Distributer and the PPAP impact IATF 16949 - Automotive Quality Systems Standard 3
A Product Impact When Thread Rings Require Adjustment General Measurement Device and Calibration Topics 9
Marc What companies think climate change will have a material impact on their business? Sustainability, Green Initiatives and Ecology 11
K IEC 60601-1:2012 - Impact Assessment for changes from IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Business Impact Analysis Questionnaire - Financial Implications Business Continuity & Resiliency Planning (BCRP) 1
Howard Atkins Rules 4th Edition Areas of Impact for Client Consideration IATF 16949 - Automotive Quality Systems Standard 1
S Exploring the Impact of Changes in ISO 14001:2015 ISO 14001:2015 Specific Discussions 3
Q Complete Company Restructuring & Impact on ISO 9001 Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
M Do Volume Spikes impact Quality Scores and Utilization? Which Tool To Use Statistical Analysis Tools, Techniques and SPC 2
A Organization Name Change & Its Impact in QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Medical Device Software Version Upgrade Regulation Impact 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
F Moving our Distribution Company - ISO 9001 Certification Impact ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
AnaMariaVR2 Legal Flash: 8 Regulations set to impact the Pharma & Biotech Industry in 2013 US Food and Drug Administration (FDA) 0
G TS-16949 7.6.2 ? Calibration/Verification Records - Assessment of the Impact IATF 16949 - Automotive Quality Systems Standard 5

Similar threads

Top Bottom