When are Medical Device Accessories MD's?

Highground

Involved In Discussions
#1
Hi Everyone,
I received this question from a person asking about the TGA...or telling me. Can anyone let me know how I need to register accessories if they apart of a whole system? We listed them as accessories and have been registered for 5 years with TGA, they also have their own UDI# Help?

This is the email.

The “system” is registered as a whole, including any components.

However, should a probe or other components, which are termed as Medical Devices (e.g. probe / sensor) are to be supplied separately at a later stage, the TGA requires us to register them separately as Medical Device on its own class (if they are classified as Medical Devices).

Therefore, I would like to have a clarification, whether on their own, the sensors such as venous xxx or arterial xxx or the flow probes for the system are considered Medical Devices or not?

If you are classifying them as “Non-Medical Devices” in EU, I would like you to confirm that.
======================
Any help would be appreciated.:confused:
 
Elsmar Forum Sponsor

sreenu927

Quite Involved in Discussions
#2
TGA registration (in Australia) is for each individual medical device/accessory and if it is a system, then each individual component of that system/pack.
The registration of your accessories has to be by GMDN Code. Identify the GMDN Code for those specific accessories, classify accordingly and provide necessary documentation to your Australian Sponsor.

Regards
Sreenu
 

Highground

Involved In Discussions
#3
TGA registration (in Australia) is for each individual medical device/accessory and if it is a system, then each individual component of that system/pack.
The registration of your accessories has to be by GMDN Code. Identify the GMDN Code for those specific accessories, classify accordingly and provide necessary documentation to your Australian Sponsor.

Regards
Sreenu
Thank you! This helps greatly!
 
G

GCF21

#4
If your products have been registered with the TGA for 5 years my suggestion would be to speak to your Australian sponsor who should be able to advise on this.
 

Ronen E

Problem Solver
Staff member
Moderator
#5
Hi Everyone,
I received this question from a person asking about the TGA...or telling me. Can anyone let me know how I need to register accessories if they apart of a whole system? We listed them as accessories and have been registered for 5 years with TGA, they also have their own UDI# Help?

This is the email.

The “system” is registered as a whole, including any components.

However, should a probe or other components, which are termed as Medical Devices (e.g. probe / sensor) are to be supplied separately at a later stage, the TGA requires us to register them separately as Medical Device on its own class (if they are classified as Medical Devices).

Therefore, I would like to have a clarification, whether on their own, the sensors such as venous xxx or arterial xxx or the flow probes for the system are considered Medical Devices or not?

If you are classifying them as “Non-Medical Devices” in EU, I would like you to confirm that.
======================
Any help would be appreciated.:confused:
Hi,

My response below is based on my knowledge of the Australian medical devices regulatory system, and on my experience, including as an Australian sponsor. It's somewhat a shot in the dark though because I haven't seen what the TGA has written to your company, exactly.

In my understanding if you qualify and register a system (a medical device) with the TGA, and the scope covers any components / spare parts / accessories that are initially supplied with / as part of the system, and some of those are later re-supplied to the system's end-user, they wouldn't require a separate registration (ARTG inclusion). However, if some of those items are supplied to other users independent of the system under which they were initially cleared, they might require their own inclusion, depending on what they actually are in the regulatory sense.

An accessory to a medical device is defined as an item that's intended by its legal manufacturer to enable another* medical device to fulfill its own intended use (as intended by the medical device's legal manufacturer). If a given medical device is able to fulfill its own intended use without the item in question, or the item's legal manufacturer didn't specifically intend their item to serve that specific medical device, the item is not an accessory to that medical device. If there is no specific medical device which an item in question is an accessory to, it is not an accessory to a medical device at all.

The Australian system is very much the same as the EU one (under the MDD) and therefore there shouldn't be a major difference in the way the same situation is handled in both.

Cheers,
Ronen.

*) If the item is enabling a medical device that the item is officially a part of (eg a system), it's a component of that device and not an accessory to it.
 
Last edited:

Highground

Involved In Discussions
#6
I have another question...several EU and Asian countries are asking for a DofC with all the accessories listed, even accessories to accessories. I'm not sure how to provide this information or if it should even be on the DofC since they are systems. The customer has a choice of different accessories so the system can be built to their needs. Should I just make this a statement of some sort with all possible accessories?
 
L

locutus

#7
Just a comment that I have seen multiple regulatory agencies wanting all of accessories of devices listed, as mentioned above especially if they are sold separately or for different purposes. This is a way that the regulatory agencies are keeping track of all products coming into their country. It also might benefit in order to not have your shipments held at border control.
 

Ronen E

Problem Solver
Staff member
Moderator
#8
I have another question...several EU and Asian countries are asking for a DofC with all the accessories listed, even accessories to accessories. I'm not sure how to provide this information or if it should even be on the DofC since they are systems. The customer has a choice of different accessories so the system can be built to their needs. Should I just make this a statement of some sort with all possible accessories?
This is too much of a general question to answer. To provide some meaningful and responsible answer I whould have needed access to the relevant specific documentation and details. Sorry.
 

Highground

Involved In Discussions
#9
Okay. Let me put it this way.

It's a monitor that is useless without the sensors and vice versa The sensors are 6 different sizes according to the physiology of the patient. So a hospital might buy only 2 sizes, which does happen. They can buy other sizes at a later date.
So it's a system when used together.
Component of the MD or an Accessory? I think I'm going to list all the sensors with their GTIN numbers.

I had to come back and edit this. I understand finally!! I just had to reread the responses from everyone.
 
Last edited:
Thread starter Similar threads Forum Replies Date
C CE Marking - Medical Device Accessories EU Medical Device Regulations 0
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
R An indication that the device is a medical device (MDR, Annex 23.2q) - applicable for accessories? EU Medical Device Regulations 5
S How to register class IIA medical device accessories EU Medical Device Regulations 1
J Are DAM tools, Mobile application, clinical dashboard, etc. considered accessories to medical device or supportive functions? Manufacturing and Related Processes 4
M OTS (Off-the-shelf) Medical Device Accessories - Camera EU Medical Device Regulations 1
K Medical Device Symbol(s) on Accessories EU Medical Device Regulations 28
M UDI labeling MDR requirement for reusable medical device accessories EU Medical Device Regulations 12
M Medical Device Accessories - Health Canada - CMDR Section 21 Canada Medical Device Regulations 1
shimonv Summary of FDA Guidance on “Medical Device Accessories” (Dec 2016) Other US Medical Device Regulations 1
M CE Scope on Medical Device & Device Accessories to the Foley Catheter CE Marking (Conformité Européene) / CB Scheme 2
R Medical Device Accessories vs. Detachable Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J 510k file requirements for non-sterile medical device accessories 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
D Separate Classifications for Medical Device Accessories EU Medical Device Regulations 2
M Who is the Manufacturer - Medical Device Accessories? EU Medical Device Regulations 6
M Providing accompanying Medical Device Accessories CE Marking (Conformité Européene) / CB Scheme 2
A Classification of Medical Device Accessories 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
Q Technical File for Medical Device Accessories CE Marking (Conformité Européene) / CB Scheme 4
G Medical Device Accessories (e.g.: ECG Patient Cables) - Traceability Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
A Customized Medical Device Accessories for Research Purposes Only EU Medical Device Regulations 14
T Device Master Record For Medical Device Accessories ISO 13485:2016 - Medical Device Quality Management Systems 2
K Regulatory Aspect of Embedding 4G/5G communication technologies into medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M FDA Medical device reporting (Manufacturer in US; contract manufacturer OuS) US Food and Drug Administration (FDA) 0
G Medical Device Auditor (CMDA) certification exam by ASQ - looking for input Career and Occupation Discussions 3
B Documenting Medical Device Complaints after End of Life? Medical Device and FDA Regulations and Standards News 0
M Medical Device License as Distributor Canada Medical Device Regulations 8
M Preparing a document for Raw Material of Medical Device EU Medical Device Regulations 1
R Medical device regulations in African countries Other Medical Device Regulations World-Wide 0
Z Swiss Authorized representative & non-medical device regulations Other Medical Device Regulations World-Wide 0
T FDA UDI Question - Class II Medical Device Other US Medical Device Regulations 1
blackholequasar ISO 13485 certification prior to Medical Device Manufacturing... worth it? ISO 13485:2016 - Medical Device Quality Management Systems 4
P Brexit Germany - Import Medical Device For Clinical Study EU Medical Device Regulations 0
M Medical Device Marketing Material - Control of Social Media 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Ed Panek Make sure you hire competent QA RA Folks before making a "Medical Device." Coffee Break and Water Cooler Discussions 1
B Acquired Medical Device Product Line - Documentation Requirements for Device Master Record ISO 13485:2016 - Medical Device Quality Management Systems 7
MedicalDevicesCanada How to find a medical device contract manufacturer, MDSAP certified? Canada Medical Device Regulations 5
dgrainger Informational MHRA's Software and AI as a Medical Device Change Programme UK Medical Device Regulations 0
S Medical Device - Technical Documentation structure EU Medical Device Regulations 1
LostLouie Archived Specifications and Drawings for Medical Device Technical Files EU Medical Device Regulations 4
B Reprocessing of an accessory for a medical device EU Medical Device Regulations 4
A Class 1 medical device - Thailand Other Medical Device Regulations World-Wide 0
P Interchangeable/alternative parts in BOM (medical device) Manufacturing and Related Processes 4
U Is Initial Importer Status Required if a Medical Device is Manufactured and Sterilized by an OEM in the US Other US Medical Device Regulations 1
J Shoe Covers - medical device class I EU Medical Device Regulations 3
J Software as a Medical Device - SaMD IEC 62304 - Medical Device Software Life Cycle Processes 3
H Existing cloud based medical device - questions regarding improving the processes IEC 62304 - Medical Device Software Life Cycle Processes 6
K Unused Service Parts in Newly Manufactured Medical Device? Other US Medical Device Regulations 1
B A.I. diagnostic software is considered as medical device in FDA? US Food and Drug Administration (FDA) 6
J Medical device repairs (to upholstery) Manufacturing and Related Processes 4
Ed Panek Apple Provides New Medical Device Cautions 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2

Similar threads

Top Bottom