Hello All:

When do ISO 17025:1999 accreditations expire? When are ISO 17025:2005 accreditations required in place of an ISO 17025:1999 accreditations?

Thank you,

Dirk
 
Elsmar Forum Sponsor

Sidney Vianna

Post Responsibly
Staff member
Admin
#2
vanputten said:
When do ISO 17025:1999 accreditations expire? When are ISO 17025:2005 accreditations required in place of an ISO 17025:1999 accreditations?
The International Laboratory Accreditation Cooperation (ILAC) has set a transition period of two years from date of publication of the new edition - 12 May 2005 - for accredited laboratories to comply with the standard's requirements.

Excerpt
 

CalRich

Involved In Discussions
#3
17025 accreditation

Each accreditation has its own expiration date, based on when the organization was last assessed. The 1999 accreditations should all be expired in the next year or two, as they normally would - ours expires in March 2007. The next time you are assessed to 17025, it will be to the 2005 version of the standard.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#4
Accreditation assessments are currently conducted to the 2005 revision. ILAC/APLAC has set a drop-dead date of June 2007 for accreditations to be under the 2005 revision.

Hershal
 
K

Ken K

#5
We just had our audit a few weeks ago and were given the choice of either being audited to the 1999 or 2005 version. We chose the 2005 version.

L-A-B will stop auditing to the 1999 version in October 2006.
 
Thread starter Similar threads Forum Replies Date
K Differences - ISO17025:1999 vs. ISO17025:2005 - The addition of "Improvement" in 4.10 ISO 17025 related Discussions 16
M What are the changes in supervising according to ISO17025? ISO 17025 related Discussions 7
W My company is looking to start a materials testing lab that conforms to ISO17025:2017 ISO 17025 related Discussions 3
B Grading of non-conformities for ISO17025:2017 ISO 17025 related Discussions 5
C ISO17025 for In-House Calibration Labs: Is it required? ISO 17025 related Discussions 30
Jerry Eldred ISO17025 Accreditation Not Required For Compliance ISO 17025 related Discussions 1
M ISO17025 Testing Laboratory Equipment Calibration Requirements ISO 17025 related Discussions 4
S Advice for ISO17025 First Round of Internal Audits ISO 17025 related Discussions 10
G ISO17025 requirement in TS16949 for External Calibration Laboratories ISO 17025 related Discussions 4
Jerry Eldred ISO17025 vs. Z540 M&TE vs. Standards Laboratory ISO 17025 related Discussions 7
P Dimensional Mechanical Inspection only as a Service - ISO17025 or ISO17020 ISO 17025 related Discussions 5
D Cross-Reference between ISO9001:2008 and ISO17025:2005 ISO 17025 related Discussions 3
Z Don't quite understand interpretation of ISO17025: 2005 clause 1.4 and 1.2 ISO 17025 related Discussions 13
M Drafting Quality Manual for ISO17025 ISO 17025 related Discussions 10
B If not ISO17025, what are the other types of accreditations that I can use? ISO 17025 related Discussions 6
B Effect of Calibration Laboratory Changing Ownership & Moving on ISO17025 registration ISO 17025 related Discussions 4
D Defining ISO17025 Calibration Laboratory Technical Competency ISO 17025 related Discussions 1
E Conflict? 2 Technical Managers identified (one for EMC, one for Reliability) ISO17025 ISO 17025 related Discussions 8
B EU REACH Regulation - is ISO17025 accreditation required for internal labs? ISO 17025 related Discussions 6
B Measurement Uncertainty (MU) vs. Level of Certainty - ISO17025 para. 5.1.1 ISO 17025 related Discussions 8
A Are ISO 9001 certified companies required to use ISO17025 facilities for calibration? ISO 17025 related Discussions 13
D External Laboratory Inquiries (ISO17025) - Product validation process Per AEC-Q100 ISO 17025 related Discussions 3
T Help with MU and ISO17025 basics - Hired as a CMM programmer ISO 17025 related Discussions 17
R Looking for a company in Colorado - ISO17025 certification consulting & training ISO 17025 related Discussions 1
B Z540 vs. ISO17025 Comparison Matrix - Calibration Customers who still insist on Z540 ISO 17025 related Discussions 3
K Combining ISO17025 with ISO9001 Documentation - Product is standard solutions ISO 17025 related Discussions 3
G ISO17025 Certification with out ISO9001 ISO 17025 related Discussions 2
A ISO17025 Quality Manual - Calibration - Radiotherapy and radiation protection ISO 17025 related Discussions 0
S Is a low calibration TUR acceptable in the eyes of ISO17025 and TS16949? 4:1 TUR Rule ISO 17025 related Discussions 2
S Comparison: ISO9001:2000 v ISO17025 ISO 17025 related Discussions 4
S ISO17025 vs. ISO9001:2000 Comparison ISO 17025 related Discussions 10
K Should QS/TS certified internal test/calibration labs be required to be ISO17025 ISO 17025 related Discussions 16
R Guide 25 vs. ISO17025 - Substantially more requirements in ISO 17025 ISO 17025 related Discussions 12
Sidney Vianna Informational ISO 14001 Certificates Top 15 Countries - 1999 to 2017 ISO 14001:2015 Specific Discussions 3
M Informational US FDA – MDR Data Files – Alternative Summary Report Data Since 1999 Available Medical Device and FDA Regulations and Standards News 0
P Warning Letter 1999 - Where does the FDA make older warning letters available? US Food and Drug Administration (FDA) 4
O EN ISO 12322:1999 vs. EN ISO 13612:2002 - Which to use Other Medical Device Related Standards 3
L What changed in Amendment 1:2011 in the ISO 2859-1:1999? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Application of IEC 60601.2.2:1999 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S ISO/IEC Guide 51: 1999, Safety Aspects - Guidelines for their inclusion in standards Food Safety - ISO 22000, HACCP (21 CFR 120) 4
E Can someone help me with ISO 11539:1999 for contact lens classification? Other US Medical Device Regulations 2
Q EN 55011:1998 +A1:1999 +A2:2002 vs. EN55011:2009 Test Levels - Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Validity of Old Biocompatibility Data (1999) - Catheter Other Medical Device and Orthopedic Related Topics 3
B OHSAS 18001:1999 vs. OHSAS 18001:2007 - Differences Occupational Health & Safety Management Standards 1
R Copy of TS16949:1999 for reading purpose IATF 16949 - Automotive Quality Systems Standard 7
G Sampling plans for large quantities along lines of ISO 2859-1:1999 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
R Sampling Plans - ISO 2859-1: 1999 - How to calculate the switching score? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
G Differences between ISO 17025:1999 and ISO 17025:2005 ISO 17025 related Discussions 4
Antonio Vieira ISO 14031:1999 "Environmental Performance Evaluation - Guide Lines" - Replaced? Miscellaneous Environmental Standards and EMS Related Discussions 2
A Radio & Telecommunications Terminal Equipment Directive 1999/5/EC CE Marking (Conformité Européene) / CB Scheme 5

Similar threads

Top Bottom