When does design transfer take place?

U

Uberbert

#1
In searching this forum, I found threads on the “what” and “how” of design transfer but not so much on when. Conversations with quality engineers have suggested my perspective is off base.


As I read the standard and guidance, design transfer should occur before design verification and certainly before design validation. As a design engineer, 14 years ago I helped set up a quality system at a design engineering firm. We had a design transfer process and form that we used almost every time we released drawings to make an injection mold. I am helping set up another system again and wanted to confirm that strategy.


I think my argument can be summed up this way.
1. the definition of design validation indicates that there are device specifications to validate, therefore device specifications come before validation
2. the output of design transfer is production specifications, therefore design transfer comes before production specifications
3. Given 1 and 2 and assuming “device specifications" and "production specifications" are equivalent, design transfer comes before validation.


Other perspectives I have seen are (1) that design transfer occurs after design validation is complete and (2) is intended to ensure that the DMR is complete and adequate. My argument above refutes the (1) and regarding (2), neither the standard nor the guidance say anything about DMR in the design transfer section.


Finally, assuming design transfer occurs before design validation, I think the FDA chose poorly in listing it in the standard after design verification and design validation.


If I am off base or missing something, please explain.
 
Elsmar Forum Sponsor

mihzago

Trusted Information Resource
#2
Design Transfer is not a single point in time, but rather a set of tasks and activities that may start as early as the definition of requirements (design inputs), and end when the device is released for actual production.

Design transfer does not have to happen before Design Validation, because nothing says it has to. The regulations states "initial production units ... or their equivalents".
Frequently the design validation will have an input into the production process (e.g. process validation) and vice versa, a transfer activity such as pilot run (or another activity) may require changes that impact Design Validation.
 
U

Uberbert

#4
Mizhago,

We agree on many, but not all, points. I think my 3 point argument indicates that design transfer must to occur, or at least start, before design validation. In other words, how can manufacturing provide production units or their equivalents to quality for design validation if design has not "transferred" a device design to manufacturing?

Bert
 
J

Julie O

#5
I think design transfer is the process of providing manufacturing with the information it needs to manufacture a new design. If there is information available at the inputs stage that they will find useful, and you pass it along at that stage, you have begun design transfer. The transfer ends whenever they have enough information to start production. This does not have to be production of the final product. It could be production of units needed for validation. Depending on how the validation works out, this might not be the design that will be manufactured for marketing. In that case, additional transfer activities (usually of design modifications) will occur between validation and market production.
 

Ronen E

Problem Solver
Moderator
#6
Regulatorily, "Design Transfer" is not the transfer act itself but the (culminating) act of methodically making sure that the design is correctly translated into production specifications. For instance, a plastic part design may be represented by a drawing with some informative comments, while that drawing is not yet sufficient for mass production through injection molding (IM). The production specifications may (should) include, among other things, the IM process parameter settings. In Design Transfer, the manufacturer ascertains that the IM process has been properly fine-tuned for mass production and that the results have been documented in a formal and binding way, to guarantee good parts production. Of course, this is just one aspect of production specifications - I highlighted it as an example of the difference between "pure" design and production specifications.

The language in 21 CFR 820.30(g) (Design Validation) suggests that for these regulations "Production" refers to serial ("marketing") production, rather than to prototype / R&D style production; hence IMO Design Transfer refers to the formalization / setting of specifications related to mass-, commercial production. Informing those responsible for making units for Design Verification, Design Validation, Process Validation etc. (e.g. manufacturing staff) of the design / product requirements which are necessary for delivering test units would fall under the scope of these steps in the Design Control process (Design Verification, Design Validation etc.); not under Design Transfer in the regulatory sense.
 
S

Scott A. Bednar

#7
The transfer to manufacturing (not design transfer) is a long drawn out process that builds over the life of the project. As the design evolves, manufacturing documentation is developed and processes are optimized until the design is frozen. The processes are validated and formal design verification is completed. The next run of product may be considered production and these units are used for validation testing and in some cases summative usability studies. Once all of the boxes are checked and the manufacturing engineer accepts the package, I consider that point in time design transfer. This scenario follows the layout of the standard and is typical of a sterile disposable development cycle.
 
#8
Design transfer is critical to product quality and it takes place when production specifications/working drawings are transferred from the design specifications.820.30 (h)
 

mihzago

Trusted Information Resource
#9
I think I should clarify my response in earlier post on May-11; in the sentence "Design transfer does not have to happen before Design Validation ..." , the word "happen" should be replaced with "complete". So to reiterate, the design transfer has started, but is not necessarily complete prior to design validation.


There seems to be a common theme in all of the responses, which is also in line with my opinion, that the "Design Transfer" is not a distinct gate or a phase but rather a set of tasks and activities that progress throughout the project and may start as early as project initiation. It ends when Production team has all the necessary information to begin manufacturing devices that can be sold to the public.
 

Ronen E

Problem Solver
Moderator
#10
"Design Transfer" is not a distinct gate or a phase but rather a set of tasks and activities that progress throughout the project and may start as early as project initiation. It ends when Production team has all the necessary information to begin manufacturing devices that can be sold to the public.
I respectfully disagree.

While I agree that the transfer of information from the design team to the manufacturing team, covering specifications that enable commercial production, is a long process that may start very early in the project, "Design Transfer" - in the regulatory (FDA) sense - IS sort of a gate event, in the sense that it should be relatively short end decisive; it requires a certain input to begin (design freeze, or the successful completion of design & process validation) and when successful/complete it allows the NPI process to proceed to commercial production. As I noted above, Design Transfer in the regulatory sense is the act of ensuring (checking, verifying) that the actual transfer work product is complete and adequate.
 
Thread starter Similar threads Forum Replies Date
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 25
K Does company who manufacture but does not design or carry out clinical trials is responsible for CE marking for single use medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 4
M Does a CDR (Critical Design Review) have to have a moderator that is unconnected to the team? Design and Development of Products and Processes 4
C Optic Patchcord Cables - Does 8.3 (Design) Apply Here? ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M Does the Scope of 21 CFR Part 820.72 (Equipment) apply to Design? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Where does process validation fit within design & development? Design and Development of Products and Processes 7
N What does an effective design process look like? (shoots flare in the air) Design and Development of Products and Processes 2
O Design and Development Process - Does one size fit all? ISO 13485:2016 - Medical Device Quality Management Systems 13
L Medical Device New Product Design & Development- How long does this process take you? Other Medical Device and Orthopedic Related Topics 1
L When does Medical Device New Product R&D stop and Design start? Design and Development of Products and Processes 6
S Definition Design Capacity - What does 'Design Capacity' (or 'Rated Capacity') mean? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 10
T Does the Layout Design of a PCB fall under clause 7.3 in ISO 9001 Design and Development of Products and Processes 7
C At what stage does Design Control play a role in Medical Device Development ISO 13485:2016 - Medical Device Quality Management Systems 2
Q Definition Design Margin - What Does Design Margin (DM) mean? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 16
L Does 820.30 Design Controls apply to Unclassified Device with 510(k)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
Z Does a Tooling Build or Design Benefit From TS16949 IATF 16949 - Automotive Quality Systems Standard 8
S Does ISO 9001 Clause 7.3 Design and Development apply? Design and Development of Products and Processes 18
D Does TS Organization have to use the Design Information Checklist? Design and Development of Products and Processes 1
A DOE - Design of Experiments? What does that mean? Quality Tools, Improvement and Analysis 5
I When does development & design verification take place? Design and Development of Products and Processes 1
S A-2 Design Information Checklist - Does anyone have the A-2 checklist in Word format? Design and Development of Products and Processes 3
J When does an engineering change become a design change? Design and Development of Products and Processes 7
Marc Wine By Design - Does Development of a New Wine Qualify as Design? Design and Development of Products and Processes 16
G When does containment event not require 100% sort? Nonconformance and Corrective Action 5
H Does this clause mean i must be in compliance with AS9100 or is 9001 enough? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Does ISO state in anyway that we MUST keep physical first piece parts? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 20
S Brexit ISO 13485:2016 + Corrigendum - What does a UKCA DoC require? EU Medical Device Regulations 2
G Does anyone know about tobacco-free nicotine pouches? US Food and Drug Administration (FDA) 0
cgaro62 Does FDA apply to a non-medical 13485 certified custom manufacturing company? ISO 13485:2016 - Medical Device Quality Management Systems 11
R Does anyone use iQMS for their ISO based document control? Manufacturing and Related Processes 1
V How many hrs does it take to create a Technical File (TD)? EU Medical Device Regulations 4
M How does FDA determine OAI and VAI after inspection? ISO 13485:2016 - Medical Device Quality Management Systems 0
G Does TPI agencies comes under outsourced process as per Q1 Oil and Gas Industry Standards and Regulations 11
N FPY requested for an industry that does not have rework IATF 16949 - Automotive Quality Systems Standard 4
William55401 21 CFR 11 Password Aging - What does "periodically" revised mean in your org? Qualification and Validation (including 21 CFR Part 11) 3
I CQI-15 Welding System Assessment 2nd edition - Does not require Ultrasonic Welding anymore? Customer and Company Specific Requirements 0
Ajit Basrur Does "Refurbishment" fall under "Service"? ISO 13485:2016 - Medical Device Quality Management Systems 13
T Does marketing company require CE mark if manufacturer has CE mark on product? EU Medical Device Regulations 5
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 6 – How Does This Even Happen? Other US Medical Device Regulations 0
T What does VW mean by annual requalification? Customer and Company Specific Requirements 4
I Does BSI require suppliers to be ISO 9001 Certified? EU Medical Device Regulations 12
Watchcat Does Regulatory Affairs put the "no" in innovation? Other US Medical Device Regulations 14
Q Does anyone know what this call out means? Manufacturing and Related Processes 3
K COPLANARITY: Composite profile tolerance on multiple surfaces- what does" lower dimensional reference frame tolerance" control? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
P Does FDA require certification for quality system internal audit for auditor? Qualification and Validation (including 21 CFR Part 11) 1
H Does anyone has feedback on a notified body for small companies? EU Medical Device Regulations 3
M Does 4.5 - Alternative RISK CONTROL apply to the Particular Standards? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Does IVD need to integrate with hospital IT infrastructure? Medical Information Technology, Medical Software and Health Informatics 2

Similar threads

Top Bottom