When Does Resolution of Measurement Standard Apply in Uncertainty Calculations
Jerry Eldred
I've had my own views on this for a while, but haven't had to deal with it. But it is coming up again, and needs to be properly addressed....
For example, a Fluke 5720A is used to verify 10 Volts DC on an Agilent 34401A DMM. There are various contributors to measurement uncertainty, which is not my question/focus for this question. But regarding resolution....
It makes perfect sense that the resolution of the DMM is a contributor. But where I have a problem is if I simply enter an output of 10 Volts DC from a Fluke 5720A calibrator, there are opinions out there that the resolution of the calibrator is also a contributor. I could even see the least significant non-zero digit of the calibrator being considered the resolution perhaps. But at a setting for a cardinal value, or even non-cardinal value, the resolution is accounted for in its specifications already.
Any knowledgable inputs as to exactly why resolution must be considered a contributor? Or, are there only certain AB's that think that, or is this subject to adjudication and engineering judgment.
I'd appreciate any knowledgable feedback on this anyone could give. Have a great evening all.
dv8shane
I personally account for the resolution of the calibrator too. The contribution is small but does exist.
dv8shane
Jerry
Attached is a copy of a budget I did 7 years ago for a Fluke 8506A calibrated with a Datron 4000A. Both units had 7.5 digits of resolution and had accredited calibration with adjustment to nominal just before I took the readings. The contribution is really small but it is there.
dwperron
Here is a link to an A2LA guidance document which covers the rationale for including resolution of the standard in uncertainty calculations. It's written by Klaus Jaeger, and as usual is an excellent presentation.
Go to the A2LA website
Click on Publications
Click on Search A2LA Site
Search for document G110
Atul Khandekar
Hershal
Metrologist-Auditor
The real question is not, is there a difference? There is a difference.
The real question is, does that difference truly make a difference? That I would argue is not always true, because the difference may be so small that the reporting of the uncertainty is not affected. Given Jerry's example, if the uncertainty absent resolution is 10mV, and adding the resolution results in a change of say 0.05mV, then the difference is insignificant, and can simply be accounted for the uncertainty budget as a comment.
Typically, resolution is a more significant factor in mechanical and dimensional calibration, rather than electronic calibration.
Spleen
In theory, your standard (Fluke 5720A) resolution contribution is supposed to be included in your imported calibration uncertainty (The Fluke 5720A calibration certificate). In your case, the Short-term repeatability will be a bigger contributor. It should be already taken into account during the calibration.
brian33a
Yes, you are absolutely correct. Unfortunately, you have to prove it. The easiest way to do that, for any given model, is to include the resolution of the calibrator in the list of contributors.
Also, it should be noted that A2LA states (R205 4.2.1):
Every measurement uncertainty shall take into consideration the following standard
contributors, even in cases where they are determined to be insignificant, and documentation of the consideration shall be made:
a) Repeatability (Type A)
b) Resolution
c) Reproducibility
d) Reference Standard Uncertainty
e) Reference Standard Stability
f) Environmental Factors
So, even though the resolution of the 5720A is insignificant, they still require it. I can only imagine they require that because people kept leaving out the resolution of the standard when it was significant (even if barely).
dwperron
b) Resolution
c) Reproducibility
You are correct, A2LA does require resolution in the uncertainty budget. Period.
But then, they also get inconsistent, such as this from the same R205 document:
4.0 Requirements for Uncertainty Calculations that support the Calibration and
Measurement Capability (CMC) on the Scope of Accreditation
4.1 Terms and Definitions:
b) Significant: “significant” further means a contributor whose contribution increases the CMC by five percent (5%) or greater
But even though you know that a 5720's resolution will not change the CMC by more than 5% you still have to include it - the reason we were given is because "you can't tell if it is significant unless you actually calculate it". So A2LA says you just have to do it, even if it isn't a significant contributor. But if you have different assessors...
Hershal
Metrologist-Auditor
I would say where the resolution makes a difference in the uncertainty that can be observed, then yes it must be considered. This is typical in calipers and micrometers for example.
A 5720 calibrating a 34401 likely no noticable impact.
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 0
0 To which part of 13485 does this refer? ISO 13485:2016 - Medical Device Quality Management Systems 3
lanley liao Does the customer`s trademark belong to customer-supplied property? Oil and Gas Industry Standards and Regulations 2
H How does a gas turbine work on diesel fuel? Oil and Gas Industry Standards and Regulations 12
G What does performance specification include? US Food and Drug Administration (FDA) 1
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S How long does it take to register a product with MHRA? UK Medical Device Regulations 3
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
A Brexit When does the UK responsible person need to be in place? UK Medical Device Regulations 10
N Does anyone have experience of GB/T 34986-2017? China Medical Device Regulations 1
Z Does anyone have experience with EN ISO 17664 ? IEC 62366 - Medical Device Usability Engineering 9
F Does anyone have an ESD quality/cooler talk to share? Training - Internal, External, Online and Distance Learning 4
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
B General Motors and Honda Alliance - What does this mean to suppliers? IATF 16949 - Automotive Quality Systems Standard 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
A Does ISO 9001:2015 cover all the requirements of ISO 10012:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
A Does anyone have a checklist of API Spec 650 13th Edition? Oil and Gas Industry Standards and Regulations 0
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
L Turkish Requirements - Does the Software need to be translated? CE Marking (Conformité Européene) / CB Scheme 2
J Does Pakistan Medical Device Import License allows parallel import? Other Medical Device Regulations World-Wide 0
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
P Does anyone have a API Q1 Documentation Package? Quality Management System (QMS) Manuals 1
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
N Small Company - Internal audit process - Who does the audit? Internal Auditing 16
J Does anyone have an excel IATF 16949 Internal Audit checklist I could use? IATF 16949 - Automotive Quality Systems Standard 7
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
Watchcat Does "Similar Device" = "Predicate"? EU Medical Device Regulations 7
Y Does Solidworks (2D/3D drafting modules) need validation? Other Medical Device and Orthopedic Related Topics 5
M Definition Open Audit - What does an Open Audit mean? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
I Does training have to be written? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14