#11
In theory, your standard (Fluke 5720A) resolution contribution is supposed to be included in your imported calibration uncertainty (The Fluke 5720A calibration certificate). In your case, the Short-term repeatability will be a bigger contributor. It should be already taken into account during the calibration.
This is a great point. During an audit I fount that I was listing the resolution of the calibrator and including the mfg unc for the calibration of the same calibrator. The mfg calibration uncertainty should already include the resolution as a contributor which would add additional uncertainty that was already accounted for. You will need to watch out for situations where the calibration certificate for your calibrator doesn't account for the resolution of the UUT(your calibrator). I think most of the major calibration mfgs have there act together on including the resolution in their calibration certificates but it is always a good idea to double check.
#12
Jerry
Attached is a copy of a budget I did 7 years ago for a Fluke 8506A calibrated with a Datron 4000A. Both units had 7.5 digits of resolution and had accredited calibration with adjustment to nominal just before I took the readings. The contribution is really small but it is there.
What program using in this pdf? Can u give it?
P
Pezikon
#13
I would argue that a calibrator doesn't have resolution uncertainty. That would be like saying a standard resistor, with nothing more than "10 ohms" written on it, has resolution. The whole point of including resolution as an uncertainty contributor is because the UUT has an additional digit of resolution that is invisible and is always being rounded up or down and impacts the smallest visible digit. This is not applicable on a calibrator. It's not rounding anything.
:2cents:
Jerry Eldred
#14
Pezikon, I actually agree with you completely. This is one of those areas where I have been advised that I should just enter it into budgets.
However, just to stir things up a bit (not to argue the point - only to solicit some others thoughts) to talk out of both sides of my mouth a little, If you think about it, everything (in this context) has a resolution. I have a Fluke 732A 10 Volt DC source with no display values, no digits, just a pair of terminal really nice Fluke banana binding post terminals.
If I measure this with my Fluke 8508A DMM, there it the 8.5 digits of resolution, repeatability, etc. But, as I think about it, I believe the 732A also has a resolution. It was certified by Fluke and they provided me with a value with a given resolution (can't remember it at the moment). So that is its resolution.
I have some nice standard resistors that I send to a particular standards lab for calibration. They are just analogue resistors. But the certifications provided have the resolution of the measured value, etc. Even the resistance outputs of a Fluke 5700A/5720A calibrator have determined resolutions. The overall uncertainty may be enough larger that it isn't significant. but it is still there.
Even on the DC volts output of a Fluke 5720A calibrator, there is a resolution - the minimum discernment level of its settability, or perhaps the resolution of the certified values from its certificate of calibration.
Just a few thoughts to see others thoughts.
P
Pezikon
#15
I acknowledge that we are in full agree. Now for the sake of argument...
I agree, the values on the certificate have resolution uncertainty. But I would argue that resolution uncertainty is included in the uncertainty found on the calibration certificate for that measurement. Ultimately, one would import that uncertainty for their calibrator uncertainty budget. To include resolution uncertainty as a separate line item on the uncertainty budget would be excessive, it's already there as part of the imported uncertainty.
User Spleen said as much:
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 0
0 To which part of 13485 does this refer? ISO 13485:2016 - Medical Device Quality Management Systems 3
lanley liao Does the customer`s trademark belong to customer-supplied property? Oil and Gas Industry Standards and Regulations 2
H How does a gas turbine work on diesel fuel? Oil and Gas Industry Standards and Regulations 12
G What does performance specification include? US Food and Drug Administration (FDA) 1
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S How long does it take to register a product with MHRA? UK Medical Device Regulations 3
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
A Brexit When does the UK responsible person need to be in place? UK Medical Device Regulations 10
N Does anyone have experience of GB/T 34986-2017? China Medical Device Regulations 1
Z Does anyone have experience with EN ISO 17664 ? IEC 62366 - Medical Device Usability Engineering 9
F Does anyone have an ESD quality/cooler talk to share? Training - Internal, External, Online and Distance Learning 4
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
B General Motors and Honda Alliance - What does this mean to suppliers? IATF 16949 - Automotive Quality Systems Standard 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
A Does ISO 9001:2015 cover all the requirements of ISO 10012:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
A Does anyone have a checklist of API Spec 650 13th Edition? Oil and Gas Industry Standards and Regulations 0
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
L Turkish Requirements - Does the Software need to be translated? CE Marking (Conformité Européene) / CB Scheme 2
J Does Pakistan Medical Device Import License allows parallel import? Other Medical Device Regulations World-Wide 0
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
P Does anyone have a API Q1 Documentation Package? Quality Management System (QMS) Manuals 1
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
N Small Company - Internal audit process - Who does the audit? Internal Auditing 16
J Does anyone have an excel IATF 16949 Internal Audit checklist I could use? IATF 16949 - Automotive Quality Systems Standard 7
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
Watchcat Does "Similar Device" = "Predicate"? EU Medical Device Regulations 7
Y Does Solidworks (2D/3D drafting modules) need validation? Other Medical Device and Orthopedic Related Topics 5
M Definition Open Audit - What does an Open Audit mean? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
I Does training have to be written? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14