Can anyone clear up my confusion on the UK responsible person.

There are grace periods for registering products, but do the grace periods also apply to assigning the UK RP? Or is the UK RP required to be designated for 1 Jan 2021?

The guidance isn’t really explicit on this point (as far as I have seen). But the UK exit regulations do suggest UK RP would need to be in place for Jan 1 2021.
 
Elsmar Forum Sponsor

twanmul

Involved In Discussions
#2
It says in the draft legislation:
" (4) The obligations in paragraph (2) begin to apply—
(a)where a device is being placed on the market by a manufacturer with a registered place of business in Northern Ireland or by a person who has designated an authorised representative with a registered place of business in Northern Ireland, on 1 January 2021;
(b)in circumstances other than those described in sub-paragraph (a)—
(i)in the case of a relevant device that is a List A device, on 1st May 2021;
(ii)in the case of a relevant device that is a device for self-testing, on 1st September 2021; and
(iii)in the case of a relevant device that is a List B device, on 1st September 2021."

All requirements for UK RP in the draft legislation link back to paragraph 2 and therefore; unless you're "placing devices" on the market in Northern Ireland (where you will need some representation from 1st January), the grace periods apply for designating a UK RP (from what I can see anyway).
 

Awadine

Starting to get Involved
#3
It says in the draft legislation:
" (4) The obligations in paragraph (2) begin to apply—
(a)where a device is being placed on the market by a manufacturer with a registered place of business in Northern Ireland or by a person who has designated an authorised representative with a registered place of business in Northern Ireland, on 1 January 2021;
(b)in circumstances other than those described in sub-paragraph (a)—
(i)in the case of a relevant device that is a List A device, on 1st May 2021;
(ii)in the case of a relevant device that is a device for self-testing, on 1st September 2021; and
(iii)in the case of a relevant device that is a List B device, on 1st September 2021."

All requirements for UK RP in the draft legislation link back to paragraph 2 and therefore; unless you're "placing devices" on the market in Northern Ireland (where you will need some representation from 1st January), the grace periods apply for designating a UK RP (from what I can see anyway).
Thanks!

For general medical devices (not IVDs) my reading of the amendments is that the contents ‘substitution of regulation 19’ (registration of persons placing medical devices on the market) would apply to registration of devices and there’s a similar entry on timing

(4) The obligations in paragraph (2) begin to apply—
(a)in the case of a device that is a Class I device, on 1st January 2021;
(b)in the case of a device that is a Class III or IIb implantable device, on 1st May 2021;
(c)in the case of a device that is a Class IIa or Class IIb non-implantable device, on 1st September 2021

which makes sense for the actual registration of the device.

But I thought a separate registration was needed for the UK RP. One UK RP might be responsible for multiple device classes for example.

Questions:
1.Even if there’s no requirement to register a device/responsible person until the grace periods end, does that mean that a UK RP is not needed to be designated prior to deadlines for registration?

2. Who are MHRA going to hold accountable for products placed on the UK market post Jan 1 if there’s no UK RP in place, whether or not they are registered with MHRA?
 

Rincewind

Involved In Discussions
#4
Hello,
can we already designate an UK Rep right now and if yes, where would you do this? I couldn't find a form or anything on the MHRA website.
My understanding is that we need a contract with the UK Rep. and the UK Rep himself will then register with MHRA for our products.

I was also sligthly confused by an Emergo article where they said that the UK Rep also has to be the importer of the devices into the UK. Is this correct?
 

Awadine

Starting to get Involved
#5
It’s not correct that the importer and UK RP have to be the same entity. I am sure I have seen this stated somewhere. My understanding is that the I importer would be required to inform the UK RP they intended to import product.
 

Rincewind

Involved In Discussions
#6
I am wondering how exactly you register the UK Rep. I mean actually where? If I look at the MHRA webpage and database it only allows registration as manufacturer but not as a UK Rep from 01.01.2021.

Does anyone know where to exactly do that?
 

Awadine

Starting to get Involved
#7
I would guess that there will be some guidance coming from MHRA on this point. I was imagining that they might use the existing manufacturer or Authorised Rep registration functionality. But I am not sure how that would work.
 

Rincewind

Involved In Discussions
#9
I would guess that there will be some guidance coming from MHRA on this point. I was imagining that they might use the existing manufacturer or Authorised Rep registration functionality. But I am not sure how that would work.
Hello, I thought the same thing but different consulting companies (in Webinars I attended) said they are already registered as a UK-Rep, this what got me wondering.
 
Thread starter Similar threads Forum Replies Date
T Does person responsible for standard have to be the Managment Representative? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Does company who manufacture but does not design or carry out clinical trials is responsible for CE marking for single use medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 3
I Does BSI require suppliers to be ISO 9001 Certified? EU Medical Device Regulations 12
Watchcat Does Regulatory Affairs put the "no" in innovation? Other US Medical Device Regulations 14
Q Does anyone know what this call out means? Manufacturing and Related Processes 3
K COPLANARITY: Composite profile tolerance on multiple surfaces- what does" lower dimensional reference frame tolerance" control? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
P Does FDA require certification for quality system internal audit for auditor? Qualification and Validation (including 21 CFR Part 11) 1
H Does anyone has feedback on a notified body for small companies? EU Medical Device Regulations 3
M Does 4.5 - Alternative RISK CONTROL apply to the Particular Standards? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Does IVD need to integrate with hospital IT infrastructure? Medical Information Technology, Medical Software and Health Informatics 2
D What does a level 1 (PSW) PPAP actually promise? APQP and PPAP 16
R GS1 DataMatrix code does not meet GS1 Specification EU Medical Device Regulations 0
M Does the scope of ISO 9001:2015 applies to tenders, pricing and sales department of a medical devices distributor? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
K Why does load cell supplier requires force verification General Measurement Device and Calibration Topics 3
M Does C=0 strictly mean 1 bad, all bad, all the time? ISO 13485:2016 - Medical Device Quality Management Systems 6
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 5
D Does Risk Management apply to re-labeler (MDR) EU Medical Device Regulations 1
Ed Panek Does this FDA Requirement Apply to international (not USA) distributors for USA based manufacturing companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 3
S Does a refurbished product required a new UDI? US Food and Drug Administration (FDA) 3
D Change to labelling - does it require a new 510(k)? US Food and Drug Administration (FDA) 5
M What does "constantly" mean ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
G Does FDA allows remote approvals of quality documentation. Is there any specific guidance on signing any quality records remotely? Document Control Systems, Procedures, Forms and Templates 1
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 1
lanley liao Does all of the suppliers need to integrated into the supplier list qualified of the company? Oil and Gas Industry Standards and Regulations 2
0 To which part of 13485 does this refer? ISO 13485:2016 - Medical Device Quality Management Systems 3
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
lanley liao Does the customer`s trademark belong to customer-supplied property? Oil and Gas Industry Standards and Regulations 2
H How does a gas turbine work on diesel fuel? Oil and Gas Industry Standards and Regulations 12
G What does performance specification include? US Food and Drug Administration (FDA) 1
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
B Does anybody know how to get older versions of Minitab to work in Windows 10? Quality Tools, Improvement and Analysis 9
M Does the ISO 9001:2015 standard require a disaster recovery plan or emergency response plan ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S How long does it take to register a product with MHRA? UK Medical Device Regulations 14
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N Does anyone have experience of GB/T 34986-2017? China Medical Device Regulations 1
Z Does anyone have experience with EN ISO 17664 ? IEC 62366 - Medical Device Usability Engineering 9
F Does anyone have an ESD quality/cooler talk to share? Training - Internal, External, Online and Distance Learning 4
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
D Change Approval Requirements - Does every change need formal customer approval? Design and Development of Products and Processes 17
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
B General Motors and Honda Alliance - What does this mean to suppliers? IATF 16949 - Automotive Quality Systems Standard 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9

Similar threads

Top Bottom