When is Compliance required for 4th edition IEC 60601-1-2

B

bassa

#1
Hello all,

Since our device did not pass the tests for EMC (60601-1-2) we have to improve it.
The device was tested as a Class A device, (not for domestic use)
The company that is perfoming the tests, advised us to test the device as Class B, conform the new 4th edition requirements.

We are unsure what to do, is it still possible to stick to the 3rd edition?
When/If this 4th edition comes out in November this year, when is compliance required?

Thanks for your advice.
 
Elsmar Forum Sponsor

Stijloor

Staff member
Super Moderator
#2
Re: When is Compliance required for 4th edition 60601-1-2

A Quick Bump!

Can someone help?

Thank you very much!!
 
#3
Re: When is Compliance required for 4th edition 60601-1-2

From the IEC website for SC62A (the committee responsible) 60601-1-2 Ed 4 is at the ADIS stage (approved for voting to create the FDIS {Final Draft for Circulation}). A standard at this stage should be stable with only minor editorial changes to occur before the finished standard is published.

It also shows that the vote was scheduled for June. It may be that the website needs to be updated.

Publication date is shown as November of this year. The programme to create this Ed 4 started in 2006.

To answer the question on when it will replace the 2007 version - that depends upon the market you are dealing with. If you are considering the EU situation, then once IEC publish Ed 4 it will be scrutinised by CEN/CENELEC who then publish it as an 'EN' standard. It then can be listed in the 'Official Journal of the EU' which makes it an official standard for European compliance. This process would likely take from 6 months to a year.

There is, usually, a 3 year overlap period before the old version ceases to be usable to show compliance.

I don't know the details of the revised standard, but am aware that 60601-1-11 (MEE for use at home) demands Class B, so that may have been incorporated within -1-2.

From an (re)-engineering basis it would, probably, make sense to use the forthcoming standard to set your requirements even if you then use the 2007 version to claim compliance. That way you are more likely to "future proof" your device. With Ed 4 so late in development it is highly unlikely to significantly change now.
 

Peter Selvey

Staff member
Moderator
#4
Re: When is Compliance required for 4th edition 60601-1-2

I am surprised to hear that Class B will be required, I would like to see this in writing.

As I understand, Class A is important for domestic because a house may be located far away from the transmission tower, so they get for example a weak TV signal. That means it is easier for an electrical device to interfere with the weak signal. But in a clinical (or any business) situation receiving TV/Radio signals are not considered a primary concern, and the business is anyhow more likely to be located near a strong transmission source.

It is sometimes thought that Class A is safer as the emission is lower. But that is a mistake; the signal levels are so low that they are negligible from an immunity point of view. The immunity tests are designed around transmitting devices, which are in the order of V or V/m, while Class A limits are in the order of uV or uV/m. If a medical device was affected by Class B emissions, it would really be an incredibly weak device.
 
#5
The Class A / B thing is often a 'bit of a fudge'. We commonly specify that our equipment is Class A, thus stating that it is not suitable for "domestic establishments and those directly connected to the public low-voltage power supply network that supplies buildings used for domestic purposes".

This, for us, makes sense as our devices are for professional clinic based use. The "grey" area comes from something like a primary care physicians office, which may very well be in a domestic area.

To be clear, the Class A / B classification only applies to the Emissions part of the testing. It derives from the base CISPR 11 standard. When you receive the spectrum analyser plot test results you will probably see 2 coloured lines marked; these mark the 'pass levels' with the Class A being about 10dB higher, depending upon the type of receiver used for the tests.

It is assumed (by CISPR) that domestic settings have equipment that is less well protected from emissions (such as the TV that Peter notes) as well as having a less-robust power supply network than a location where ISM (Industrial Scientific Medical) equipment would typically be used.

The justification for Class A vs. B is dependent upon your device type and how sure you can be as to where it will be used. More 'Risk Analysis' of course!

As far as I am aware, only 60601-1-11 demands Class B, not surprising for the standard for home-use MEE.
 

Peter Selvey

Staff member
Moderator
#6
I just noticed that I got the Class A and B mixed in my post.

I would think that choosing Class A or Class B is not a subject for risk management. The only significant risk is interference with a weak TV or radio signal, not with other medical devices.

You might try to argue that this is still a significant risk, e.g. maybe the TV is important for emergency info. But chances are that if the signal is so weak that +10dB emission will make a difference, then reception is going to be a bit dodgy anyway e.g. cloud cover, someone walking around the room etc might just as easily interfere with reception.

Rather, it is just an annoyance issue. Which is of course important, but just not an issue in medical device regulation or risk management.
 
Thread starter Similar threads Forum Replies Date
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
S What records are required to show compliance to ISO 13485:2016? ISO 13485:2016 - Medical Device Quality Management Systems 1
R GHS compliance - Are we are required to provide an SDS? Manufacturing and Related Processes 3
Jerry Eldred ISO17025 Accreditation Not Required For Compliance ISO 17025 related Discussions 1
M Compliance Marks on Labeling - Required, recommended, or useless? EU Medical Device Regulations 3
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Is compliance to IEC 60601-1 Required for my Product? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Confused: IEC 60601-1-2:YYYY and Required Compliance Dates IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M FULFILMENT of compliance obligation versus COMPLY with compliance obligations ISO 14001:2015 Specific Discussions 2
K ISO 13485 and compliance of electronic signature ISO 13485:2016 - Medical Device Quality Management Systems 5
L Medical device HIPAA compliance in encryption Medical Information Technology, Medical Software and Health Informatics 1
J Strategy for MDR Regulatory Compliance Procedure ISO 13485:2016 - Medical Device Quality Management Systems 4
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
M Advice needed for SEH Compliance Software and ISNETWord Compatabiliy Occupational Health & Safety Management Standards 2
D HIPAA, HITECH and Interoperability compliance route Medical Device and FDA Regulations and Standards News 2
N Quality Compliance Officer - ISO 13485, London Job Openings, Consulting and Employment Opportunities 1
A Environmental Compliance obligations and risks (ISO 14001:2015 6.1.3) ISO 14001:2015 Specific Discussions 3
M Tracking Expiration dates on compliance certifications REACH and RoHS Conversations 2
T Training recommendations? Bringing our RoHS and REACH compliance efforts in-house REACH and RoHS Conversations 2
optomist1 Informational Training IMDS - Management of Product Chemical Regulatory Compliance RoHS, REACH, ELV, IMDS and Restricted Substances 2
G ISO 14001 - 6.1.3 Compliance Obligations ISO 14001:2015 Specific Discussions 1
Ed Panek Compliance with Standards? When a standard is updated/revised CE Marking (Conformité Européene) / CB Scheme 3
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
N Which EN ISO 17664 version compliance to EU MDR? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 3
N Audit non-compliance API Q1 - Use of External Documents 4.4.4 in Product Realization Oil and Gas Industry Standards and Regulations 8
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
J Management Representative and PRRC (Person Responsible for Regulatory Compliance) ISO 13485:2016 - Medical Device Quality Management Systems 5
N Audit non-compliance - API Spec Q1 9th Ed 5.6.1.2 b Oil and Gas Industry Standards and Regulations 10
J Interesting Discussion Compliance with regulations in exceptional circumstances EU Medical Device Regulations 5
L Wearables 21 CFR Part 11 compliance Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
V Preparing the IFU in compliance with MDR 745, Chapter III EU Medical Device Regulations 2
L AS9146 Implementation and Compliance AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
P MDR PRRC (person responsible for regulatory compliance) and personal liability EU Medical Device Regulations 2
R Foam mattresses used in hospitals - compliance with MDR requirements? EU Medical Device Regulations 6
E Machines in Europe not in compliance with the EC directive CE Marking (Conformité Européene) / CB Scheme 0
N EUDAMED postponement and compliance with Article 120 (3) MDR for Legacy Devices EU Medical Device Regulations 8
R Role of quality compliance in SAP Software Quality Assurance 2
E Part 11 Compliance, Excel living documents (i.e. document master list, equipment list, approved supplier list) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
M MDR Legacy Medical Device Labeling compliance timeline EU Medical Device Regulations 3
D What is the best software used for the pharma compliance management? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
T Rumours that class 1 NS, NM, NR devices will have a new deadline for MDR compliance EU Medical Device Regulations 3
M Informational From RAPS: Danish Regulators Seek to Help Smaller Companies With EU MDR Compliance Medical Device and FDA Regulations and Standards News 0
G ISO 9001 Legal Compliance and Legal Register Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Ed Panek Do Cloud services require 21 CFR Part 11 compliance? Qualification and Validation (including 21 CFR Part 11) 7
R Certificate of compliance to RoHS/REACH/WEEE - any such animal? REACH and RoHS Conversations 18
Sidney Vianna Interesting Discussion Legal compliance as part of ISO 45001 accredited certification. Major OSHA penalties in the USA. Occupational Health & Safety Management Standards 15
D Control chart applicable? Percentage of compliance with a standard Statistical Analysis Tools, Techniques and SPC 2
J Business Intelligence and 21 CFR Part 11 Compliance Qualification and Validation (including 21 CFR Part 11) 1
M Informational US FDA issued the first warning letter for UDI violations to help ensure compliance Medical Device and FDA Regulations and Standards News 0
D USFDA vs NRTL/IEC 17025 Differences - Compliance testing lab ISO 17025 related Discussions 0

Similar threads

Top Bottom