Actually you can sell till one year after your MDD EC certificate get expired.
As I can see your exp date is 15 Feb 2022; You can sell device till 14 Feb 2023; on 15 Feb 2023 - ① Cleared customs (in EU region) by 15 February 2023
Not true.
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#23
As far as I know, you "can" sell any product anywhere, any time, as long as a customer will pay you for it. There is no law or regulation, including the MDR, that determines what anyone "can" and "cannot" do.
Not accurate.
The EU has an enforcement policy and regulations, captured under the term "Market Surveillance" (annoyingly similar term to "Postmarket Surveillance", but not the same). How well it's budgeted and implemented is a separate matter, but in legality terms, it's there.

Of course, you are "allowed" to break the law anywhere and anytime, you'd "only" have to bear the consequences if you're caught.
 

Watchcat

Trusted Information Resource
#24
I was speaking from the latter perspective, although I would argue further than no one can "allow" you to do something unless they are in a position to stop you. Virtually all enforcement is reactive, action that occurs (and legally, can only occur) once you have taken an action (or sometimes neglected to take an action). You "can" put someone in prison for a crime they didn't commit, but it's pretty hard to put anyone in prison for a crime that has not been committed. Customs is a good example of an exception, as they *can* stop your product from getting to your customer and, in that case, the customer is probably not going to be willing to pay for it.

My problem with the use of "can" and "can't" in these situations is that it is often not only inaccurate and therefore not useful (occasionally harmful), but also often neglects to communicate the information that is accurate and therefore useful.
 

Ronen E

Problem Solver
Staff member
Moderator
#25
My problem with the use of "can" and "can't" in these situations is that it is often not only inaccurate and therefore not useful (occasionally harmful), but also often neglects to communicate the information that is accurate and therefore useful.
I was being brief in my response to @SKM.Sunil because I'm tired of explaining the same thing for the 100th time. I simply wanted there to be a warning for anyone coming across that post. And now there is.

I wasn't trying to be useful and I'm sure I wasn't being harmful.

Ronen out.
 

SKM.Sunil

Involved In Discussions
#27
I was just interpreting these 2 lines from article 120 (MDR), hence I said after expired certificate, the device can be put on service/sell till one year.

Article 120:
Certificates issued by notified bodies in accordance with Directives 90/385/EEC and 93/42/EEC from 25 May 2017 shall remain valid until the end of the period indicated on the certificate, which shall not exceed five years from its issuance. They shall however become void at the latest on 27 May 2024.

Devices lawfully placed on the market pursuant to Directives 90/385/EEC and 93/42/EEC prior to 26 May 2020, and devices placed on the market from 26 May 2020 by virtue of a certificate as referred to in paragraph 2 of this Article, may continue to be made available on the market or put into service until 27 May 2025.
 

Ronen E

Problem Solver
Staff member
Moderator
#28
Article 120:
Certificates issued by notified bodies in accordance with Directives 90/385/EEC and 93/42/EEC from 25 May 2017 shall remain valid until the end of the period indicated on the certificate, which shall not exceed five years from its issuance. They shall however become void at the latest on 27 May 2024.
The highlighted sentence just means that even if a certificate states an expiry date after 27 May 2024, it will nevertheless expire on 27 May 2024.
Devices lawfully placed on the market pursuant to Directives 90/385/EEC and 93/42/EEC prior to 26 May 2020, and devices placed on the market from 26 May 2020 by virtue of a certificate as referred to in paragraph 2 of this Article, may continue to be made available on the market or put into service until 27 May 2025.
That's why it's important to distinguish "Placing on the Market" from "Making Available on the Market" (both are officially defined EU terms), and from "sell" which can mean many different things unless a clear definition is attached. The original discussion was related to Placing on the Market, and I was commenting from that perspective. Placing on the Market does not occur at first entry to the EU (e.g. customs). I will stop here because I'm already repeating myself, which is exactly what I wanted to avoid.
 
Last edited:

myusoffice

First Time Right...
#29
Hi, although the question is related to MDD, i thought it wouldn't harm pulling a reference from MDR.
According to MDR Article 2 definition 28, IVDR Article 2 definition 21,
‘Placing on the market’ means the first making available of a device, other than an investigational device, on the EU market.
Then...
According to MDR Article 2(27).
‘Making available on the market’ means any supply of a device, other than an investigational device, for distribution, consumption or use on the Union market in the course of a commercial activity, whether in return for payment or free of charge.

So, to answer your question, whether it is cleared from custom or stored in distributor warehouse, as long as it enters in European Market in course of a commercial activity (Free of charge or in return of payment), it is considered placing on the market. Only exception is device for investigation purpose.

Hope this helps.
 

Ronen E

Problem Solver
Staff member
Moderator
#30
as long as it enters in European Market in course of a commercial activity (Free of charge or in return of payment), it is considered placing on the market.
The question remains, what does "enters in European Market" mean exactly. I answered this question more than once (hint: it's not intuitive), won't go there again.
 
Thread starter Similar threads Forum Replies Date
M Sample record for verification performed by importers before placing a device on the market EU Medical Device Regulations 0
R SaMD - Placing on the Market before DoA EU Medical Device Regulations 2
N "Placing on the market" & "putting into service"- one time regulatory event or Continual supply event? EU Medical Device Regulations 2
D ROHS 2 - Medical Device: Placing vs Making Available on the market EU Medical Device Regulations 3
R RoHS and placing on the market in Non EU countries RoHS, REACH, ELV, IMDS and Restricted Substances 3
C Medical Device Marketing vs. Placing Product on the Market EU Medical Device Regulations 8
C What is "Placing on the Market" in EU? Brochures or Install? CE Marking (Conformité Européene) / CB Scheme 6
A Ownbranding or placing on market without CE mark ISO 13485:2016 - Medical Device Quality Management Systems 7
I Additional National requirements for placing MD in the EU countries EU Medical Device Regulations 7
Y CE Mark - Placing the CE mark on medical device labels EU Medical Device Regulations 2
J Facilities - Placing our assembly group into a separate building ISO 13485:2016 - Medical Device Quality Management Systems 8
G Placing a dollar value on our QMS (we are looking at IP protection) Quality Manager and Management Related Issues 3
C Reducing paperwork and placing in Systems electronically - Electronic documents ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
J Automated Processing and Placing Parts Manufacturing and Related Processes 4
J How can I evaluate a new supplier prior to placing an order? Supplier Quality Assurance and other Supplier Issues 7
thisby_ Post Market/Production Risk Assessment ISO 14971 - Medical Device Risk Management 0
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
D Economic Operators, EU Market Access EU Medical Device Regulations 2
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 2
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5
R Information on obtaining a market authorization for China China Medical Device Regulations 2
M EU MDR - Labilities when a device is discontinued from the Market EU Medical Device Regulations 1
A Packaging device with accessories already on the market Other Medical Device Regulations World-Wide 0
S Can a Contract Manufacturer "release" product on behalf of client to the market? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
dgrainger Clue to UK's new regulatory requirements - approval of all devices placed on the market? Other Medical Device Regulations World-Wide 1
F When to market and distribute device in EU? EU Medical Device Regulations 3
N Post Market Surveillance Package Example Wanted ISO 13485:2016 - Medical Device Quality Management Systems 1
T Modification of a product on the market EU Medical Device Regulations 2
M Differences in post market safety reporting for Combination Product Applicants Medical Device and FDA Regulations and Standards News 1
K Product (Device) Life cycle - Product Life AFTER the Post Market Phase Medical Device and FDA Regulations and Standards News 3
M Post Market Surveillance for Class II Medical Devices Medical Device and FDA Regulations and Standards News 2
H Post-market surveillance strategy EU Medical Device Regulations 2
B Procedures for Complaint Handling and Post Market Surveillance EU Medical Device Regulations 7
A Post-market clinical followup studies (PMCF) EU Medical Device Regulations 16
M Informational Creating a post market surveillance (PMS) system for medical devices – Part 1 Medical Device and FDA Regulations and Standards News 7
E Post Market Surveillance Standard Various Other Specifications, Standards, and related Requirements 4
K EU MDD Class I Device market access / certification after 26th May 2020 EU Medical Device Regulations 9
T Post-Market Surveillance Plan - Requirements outlined in Annex III (1.1b) EU Medical Device Regulations 2
M Informational Health Canada guidance document – Pre-market Requirements for Medical Device Cybersecurity Medical Device and FDA Regulations and Standards News 0
Y Post Market Surveillance Indicators and Analysis method EU Medical Device Regulations 1
M Post Market Surveillance Plan Template EU Medical Device Regulations 10
F Vietnamese Post-Market Surveillance Requirements for a medical device in Japan Other Medical Device Regulations World-Wide 0
M PMCF (Post Market Clinical Followup)/ PSUR template EU Medical Device Regulations 8
C Shipping Medical Devices to US before pre market clearance US Food and Drug Administration (FDA) 1
M New medical device on the market before MDR EU Medical Device Regulations 12
M Informational The ANSM decides, as a precautionary measure, to withdraw from the market macrotextured breast implants and breast implants with polyurethane-coated s Medical Device and FDA Regulations and Standards News 0
S PMCF (Post Market Clinical Followup): Rationale for Exemption - Software Medical Devices EU Medical Device Regulations 6
M Informational MDCG 2019-3 Interpretation of Article 54(2)b – Pre- market clinical evaluation consultation procedure with the involvement of expert panels Medical Device and FDA Regulations and Standards News 0
S Post Market Surveillance and Annual Product Review in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 2
M Informational TGA – Q&A: Use of market authorisation evidence from comparable overseas regulatory bodies for medical devices Medical Device and FDA Regulations and Standards News 0

Similar threads

Top Bottom