In which cases is it more appropriate to use k-factors and expanded uncertainty over the Test Uncertainty Ratio (TUR)?

My understanding of it is that k-factors are to be used when there is sufficient data to be able to calculate the combined uncertainty, and that the TUR is simply approximating from the accuracy of the reference standard being used.

Does ISO 17025 give preference to using one over the other?

Also if the reference standard is the dominant contribution to the combined uncertainty is giving a k-factor = 2 equivalent to stating a TUR of 2:1?
 
Elsmar Forum Sponsor

dgriffith

Quite Involved in Discussions
#2
Re: k-Factors vs TUR

Also if the reference standard is the dominant contribution to the combined uncertainty is giving a k-factor = 2 equivalent to stating a TUR of 2:1?
The TUR and K factor are not related that way. The reference standard is not necessarily the dominant contribution--that may very well be resolution of the UUT.
In general, the TUR is the ratio of the calibration system uncertainties to the UUT specification.
The k factor is a coverage multiplier of the standard uncertainty to give a confidence level. One standard uncertainty x 2 = approximately 95% confidence (1.96 specifically). A k=3 would yield a 99.73% confidence, and so on.

By the way, many think the TUR is easier to do, but you cannot get a TUR without performing an uncertainty budget and analysis at least once. A TAR (test accuracy ratio) only concerns itself with the specifications of the standard and UUT, and is not a good indicator of the quality (risk) of the calibration, in my opinion.
 
Last edited:

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#3
dgriffith is correct.

As for when to use and regarding 17025, in the US, the TUR is considered an "identified metrological specification" and as such is allowed under Clauses 5.6.2.1.1 and 5.10.4.1.b. ILAC P14 was also revised after challenge to continue to allow the use. If you use the TUR, remember it must be backed up by an uncertainty budget.

However, expanded uncertainty is better, even though it is more work initially. Internationally, outside the US, it is typically the requirement, and most countries will not allow the TUR.

Also, the customer can request uncertainty instead of the TUR.

Hope this helps.
 
J

jerryb1426

#4
The use of TUR is for Z540-3 not ISO 17025 which this is not excepted and not allowed. A lot of calibration lab use this in error. If you go to section 5.6 of the standard and ILAC P14 you will have further Guidance.
Lead Assessor
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#5
Actually, in the US, the TUR is allowed, as ILAC revised P14 last year, to conform to ISO/IEC 17025 which allows an "identified metrological specification" in two separate Clauses. In the US, that is codified in the American National Standard ANSI/NCSL Z540.3 as the TUR.

Note, the TUR is based on uncertainty calculation, which inherently means uncertainty is a requirement anyway. To me it may simply be easier to have the uncertainty, rather than the extra work to determine the TUR.

Hope this helps.
 
J

jerryb1426

#6
Note ILAC P14 has nothing to do with TUR ratio and the P14 doc. applies to all ISO17025 Labs world wide TUR in the US are only required for compliance to Z540
If I see a Cal cert for a Standard which is used to calibrate another Instrument with only TUR values and No UNC Values it in violation of P14 and section 5.10 of the ISO17025 standard
Lead Assessor
Jerry Bambach
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#7
It is true that P14 has nothing directly to do with the TUR, except that it is an identified metrological specification, allowed under the revised version of P14. You are correct that one referencing only the TUR cannot be used as a standard to calibrate other items, but is acceptable going to the end user (customer). However, since it is codified in the US under the American National Standard, it is therefore allowed under ISO/IEC 17025 Clauses 5.6.2.1.1 and 5.10.4.1.b as an identified metrological specification. It is of course also up to the AB whether to accept the Z540 series documents within the US.
 
Thread starter Similar threads Forum Replies Date
D TUR's to alpha / beta risk factors Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
G What to do with correction factors overlooked by an accredited standard (6.4.11)? ISO 17025 related Discussions 5
P Human Factors / Usability validation in the time of COVID Human Factors and Ergonomics in Engineering 17
C Kaizen Events - Factors Affecting Failure Lean in Manufacturing and Service Industries 13
D Is Human Factors testing mandatory for a 510(k) submission? Human Factors and Ergonomics in Engineering 16
J Selecting factors for screening DoE design Statistical Analysis Tools, Techniques and SPC 8
T Determination of TENSION SAFETY FACTORS - Table 21 IEC 60601-1 Other Medical Device Regulations World-Wide 5
J DoE (Design of Experiments) - Multiple responses with different factors Using Minitab Software 2
P Usability/human factors engineering requirement (standard) for IVD medical device Other Medical Device Related Standards 2
A Interesting Discussion Human Factors as Root Cause AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 13
P Global medical device human factors/usability requirement IEC 62366 - Medical Device Usability Engineering 3
M Medical Device News FDA News - 14-09-18 - Benefit-Risk Factors to Consider for Substantial Equivalence Other US Medical Device Regulations 0
H Minitab 15 - Factorial Design - 3 factors: 4x3x2 - How to? Using Minitab Software 4
P Interesting Discussion Addressing Human Factors in Corrective Action AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 23
M I have 3 different factors - DOE help Statistical Analysis Tools, Techniques and SPC 3
A How do I create a 3^k factorial design with factors be treated as continuous where I Using Minitab Software 0
V Screening DOE with 7 Input factors and 4 responses - Significant factors Using Minitab Software 1
H Taguchi mixed model DoE [L16 (4^3 2^6): factors interaction and ANOVA calculation Using Minitab Software 3
S Taguchi Design with ANOVA - 3 factors each with 3 levels Quality Assurance and Compliance Software Tools and Solutions 13
R Is there any Excel .xls Spreadsheet available for DOE with 3 factors? Six Sigma 1
T New EU Medical Device Regulations - Q1/2 2017 and Human Factors Other Medical Device Related Standards 10
M New MHRA guidance on Human Factors - Usability Engineering IEC 62366 - Medical Device Usability Engineering 1
L BSI White Paper on Human Factors/Usability Engineering IEC 62366 - Medical Device Usability Engineering 12
M Requiring action on factors alone of RPN in FMEA APQP and PPAP 7
S IEC 62366 vs. FDA Human Factors Requirements Human Factors and Ergonomics in Engineering 2
S DOE Analysis - Experiments with 5 Factors Statistical Analysis Tools, Techniques and SPC 11
Ajit Basrur FDA issues Guidance Document - Benefits-Risks Factors to consider for 510(K) US Food and Drug Administration (FDA) 1
C MU (Uncertaincy of Measurements) definition and MU Factors Measurement Uncertainty (MU) 4
E ANOVA for Taguchi method - Four factors and three levels and nine runs (Minitab) Using Minitab Software 6
M How to compare 2 independent samples having 3 factors 3 levels Statistical Analysis Tools, Techniques and SPC 5
Y Taguchi Design with 6 levels and 3 factors in Minitab Using Minitab Software 11
R Taguchi design for 3 level 4 factors Using Minitab Software 12
V Selection of Design - 8 Factors, almost sure about Target Levels Using Minitab Software 8
R 4 Factors 3 Level DOE to determine which factors are significant Using Minitab Software 2
C Cost Factors of Iso 8 vs. ISO 8/9 Cleanroom Other Medical Device and Orthopedic Related Topics 2
M Human Factors Testing for CE Marking of a Cosmetic Device CE Marking (Conformité Européene) / CB Scheme 9
Z A DOE in Minitab - 3 factors and 3 levels for each factor Using Minitab Software 12
J What are the Design Factors in APQP? APQP and PPAP 3
Q Risk Factors Checklist identifying the Risks for meeting the Customer Indent AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
B DOE Question: 3 factors with 2 levels and 1 factor with 3 levels Six Sigma 2
C Analyzing 3 Factors 3 Levels using Minitab Using Minitab Software 9
J Metal Stampers Feasibility form that includes Risk Factors IATF 16949 - Automotive Quality Systems Standard 5
J DOE Analysis Experiment, 5 factors, 4 factors having 3 levels, and 1 factor having 5 Using Minitab Software 38
D What Factors to consider to determine the Number of Auditors Internal Auditing 3
E 5^2 DOE (Design of Experiments) with 3 Replications, 5 Levels 2 Factors Statistical Analysis Tools, Techniques and SPC 6
N Applying Correction Factors to Furnaces Manufacturing and Related Processes 5
D Resolution V Factorial Design with 5 factors (3 numerical and two text) Using Minitab Software 1
N CE Mark and Third Party Companies Evaluation Factors CE Marking (Conformité Européene) / CB Scheme 5
L DOE with 2 factors (3,4 levels) and optimized factor Using Minitab Software 12
C Packaging Shock Testing - Examples of Fragility and g-factors Other ISO and International Standards and European Regulations 3

Similar threads

Top Bottom