As Ronen notes early in the thread, there is a confusion in this thread about the meaning of "verification".
Medical device regulators (generally) require that essentially every unit be assuredly compliant. A typical meaning for "essentially every" in a common medical process, i.e. sterilization, is assurance of compliance to at least a 10E-6 standard, i.e. an uncertainty of no worse than 1 in a million.
A process that "can be verified", or for which the first unit, the last unit and a few intermediate units are checked, hasn't been verified. Verification applies to each unit. You either do it for each unit, or that unit hasn't been verified.
Don't want to verify every unit? Validation is what you do to prove that your process is reliable and you don't need to verify every unit.
You can't say that you're verifying if you're only sample-checking. Not per the applicable medical device process standards, anyway.