Where does Contact begin for Biocompatability?

F

flyingfrisbee

#1
Dear cove experts,

I am in the process of trying to ensure that our new CPAP prototype is certifiably biocompatible for CE marking, but I am having trouble figuring out what needs to be biocompatible and what testing needs to be done.

The biggest question I have is based on the contact of the device. We have air streaming from the walled oxygen and the compressors through the internal tubing, filters, and etc inside the device to the external humidifier then to the tubing, then to the patient interface. From ISO 10992-1, I classified the patient interface (which is not currently involved in our CE marking process, we may recommend buying a CE marked one already) as having prolonged contact with the mucosal membranes.

The air, oxygen, or a combination of the two pass a long way before they enter the patient's lungs. Would everything need to be biocompatability tested from the air compressor to the filters to the external tubing or would just the external parts of the device (tubing and humidifier) that are disinfected regularly need to be certified biocompatable. Also, would the same cytotoxicity, sensitization, and irritation or intracutaneous reactivity testing be required for each component (filter, regulator, tubing, humidifier, etc.)?

Also, why is ISO 10993-10 "Tests for irritation and delayed-type hypersensitivity" not on the on the harmonized standards list?

Thanks in advance,

Friz
 
Elsmar Forum Sponsor
M

MIREGMGR

#3
I don't think the particular question is answered adequately by existing standards and guidances. The ISO 10993 standard-series of course does say that it's non-applicable to non-patient-contact devices, but that's not the same question.

I used to think that the requirement was that only that part of a medical device that would, or plausibly could, have patient (or in some cases, clinician) contact needed to be biocompatible. For me, anyway, that approach is logical.

I recently was informed by a US FDA examiner, though, that their view (or at least, that particular examiner's) is that if any part of a device is required to be biocompatible, the entire device--including non patient contact elements, I guess including fully enclosed, internal components if such are present--must be biocompatible.

That of course isn't definitive for the EU. I don't know of a definitive answer for EU...maybe someone else does.
 
Last edited by a moderator:

somashekar

Staff member
Super Moderator
#4
Contact to the healthy skin is the first level of the contact for bio.
The air, O2 or the combination of other gases are not a part of your device. Your device assists in its delivery. The purity of these may not be compromised by your device in terms of contaminating or adding particulate matter, and your design and material selection must be addressing this.
 
F

flyingfrisbee

#5
Thanks for you help.

I will just conduct the cytotoxicity, sensitization, and irritation or intracutaneous reactivity biocompatibility tests on the humidifier, connectors, and tubing then.

As for the rest, I will have to make sure everything is ROHS compliant (along with the rest of the device), and free of DEHP, BPA, and latex.

Take care,

Friz
 

Ronen E

Problem Solver
Staff member
Moderator
#6
Hi,

In my opinion, there's a significant grey area WRT this issue.

The approach "only direct contact must be investigated" is detached from 10993 itself. For instance, one of the contact categories is "externally communicating, blood path, indirect". it's intended for all sorts of administration tubing (among aother things), and accounts for the possible conveyance of substances from components not-directly contacting the patient, by means of fluid media.

The opposite approach of "everything must be compatible" is ridiculous / outrageous. If an eletrical device has a fully enclosed circuitry that is detached from any applied parts in terms of substance transfer, does that circutry have to be biocompatible to the same level as the applied part?... I believe the majority of MEE on the market would fail such a requirement.

As with all things grey, a reasonable and fairly-efficient approach can be based on risk management. For instance, if you have a set of components behind a filter in a flow path (with known filter performance and reliability characteristics), there's not much point in testing for stuff that won't go past the filter. You probably need to start off with a detailed, structured analysis along the flow path, identifying the possible concerns and their sources, then evaluate the risks in terms of amounts and likelyhood of anything ending up in direct contact with a patient tissue. once you've identified the hazards, focus on those with significant risk and implement appropriate mitigation means. Back to the filter example, such means may make actual biocompatibility testing redundant in some cases. In case of doubt, you could consult an expert toxicologist (such as those working for any respectable biocompatibility test house), to frame those risks that need actual testing. For some of the lower risks, a documented opinion from such toxicologist may suffice.

This issue is not unique to you, about every CPAP device developer had to deal with it at some point. If you go to one of the big test houses, I guess that they will already have some relevant experience.

Cheers,
Ronen.
 
Last edited:
Thread starter Similar threads Forum Replies Date
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
R Where does IATF 16949 address Process mapping? IATF 16949 - Automotive Quality Systems Standard 3
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
D Where does "as far as possible" stop? FMEA - EN 14971 ISO 14971 - Medical Device Risk Management 29
A What if Contract Manufacturers does not have an ISO 13485 certificate? Where will the NB audit take place, at legal mfg. site or contract mfg. site? Other Medical Device Regulations World-Wide 3
D Where does FMEA fit in your ISO 14971 Risk Management process? ISO 14971 - Medical Device Risk Management 13
P Warning Letter 1999 - Where does the FDA make older warning letters available? US Food and Drug Administration (FDA) 4
M Where does our regulatory responsibility end? ISO 13485:2016 - Medical Device Quality Management Systems 11
K Where does the FDA post its various compliance dates? US Food and Drug Administration (FDA) 2
C Where does Quality Authority come from? Career and Occupation Discussions 4
J Where does process validation fit within design & development? Design and Development of Products and Processes 7
Q When does the FDA deem something "where appropriate"? 21CFR820.30(g) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
M Where does Document Revision Control Start? Rev 0 or Rev 1 Document Control Systems, Procedures, Forms and Templates 19
D Where does the QMS (Quality Management System) fit in your organization? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
Q Where does 5S technique fit into ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Where does OEE (Overall Equipment Effectiveness) belong? Quality Tools, Improvement and Analysis 11
somashekar Where does RoHS stuff get audited by CB's? Quality Manager and Management Related Issues 3
S IEC 60601-1 3rd Gap analysis template TO BUY - Does anyone know where? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Where does 20% Lot Qualification Rule Come from? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
G Customer Property Cl. 7.5.4 - Where does ISO 9001 stop and ISO 27001 start? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
R Where does the definition of ME (Medical Electrical) Equipment originate? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J Quality Organization - Does it matter where it reports to? Other US Medical Device Regulations 13
I Where does it state Supplier Corrective Action is required? Supplier Quality Assurance and other Supplier Issues 5
J Where does Lean Six Sigma Originate? Six Sigma 5
G 12 or 13 Men - Where does the extra man come from? Brain Teasers and Puzzles 5
S Where does least significant digit come into play? Measurement Uncertainty (MU) 8
T Organizational Charts - Where does the Quality Assurance Manager fit in Quality Manager and Management Related Issues 30
T Does anyone know where I can view Bac5300 Various Other Specifications, Standards, and related Requirements 2
A Where does the Lean Manufacturing philosophy / tool set come from? Lean in Manufacturing and Service Industries 32
H Where does the 16949 number in TS16949 come from? ISO Numbering System IATF 16949 - Automotive Quality Systems Standard 1
Wes Bucey Where does your hospital rank on quality? Hospitals, Clinics & other Health Care Providers 20
J GMP CFR 21 Part 210 or 211 - Where does GMP address Document Control Document Control Systems, Procedures, Forms and Templates 6
A Where does one find the complete list of ISO standards? Book, Video, Blog and Web Site Reviews and Recommendations 1
D Does anyone know where to buy an instant read dosimeter? ISO 13485:2016 - Medical Device Quality Management Systems 10
C What is Money for? Where does Barter come into play? Coffee Break and Water Cooler Discussions 19
E Where Does One Get Customer Specific Requirements? Customer and Company Specific Requirements 26
CarolX Where does Quality begin? Philosophy, Gurus, Innovation and Evolution 36
M Research and Development Process - Where does research end and development begin? Misc. Quality Assurance and Business Systems Related Topics 15
M Attribute Charting - Where does Cp and Cpk come into the equation? Capability, Accuracy and Stability - Processes, Machines, etc. 5
K Where and How Does Marketing Fit In? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
L Does anyone know where I can get documentation for the PUMA method. General Measurement Device and Calibration Topics 1
E SOP or system Work Instruction? Controlled Document - Where does one draw the line? Document Control Systems, Procedures, Forms and Templates 5
D Number of Distinct Data Categories:nDC: Where does the constant 1.41 come from? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S Does anyone know where Guide 25/Palcan Training is available? Training - Internal, External, Online and Distance Learning 1
R Internal audit schedules - Where does it say that I need an annual schedule? Internal Auditing 13
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
M Testing criteria - where to place Design and Development of Products and Processes 2
J UDI - Where to document it? EU Medical Device Regulations 8
J Where to get a Black belt certification ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 13

Similar threads

Top Bottom