Where to place label for an IVD test?

HarKor

Registered
Hi,
I work for a company where we test patient samples using IVD test. Patient report containing the results is sent to the clinicians. My question is regarding label requirements. I have generated a label that has all the necessary information from IVDR Annex 1, Chapter II (20.2). I need help clarifying some requirements;
  • What does date of manufacture mean?
At our company, people think that each time a sample is run, it is the date of manufacturing. I am of different opinion. I think that the device is already manufactured once design and development was completed and we got the approval. Therefore, date of manufacturing should not change every time a new sample is run.

  • Where should we include the label?
Since we don't manufacturing any reagents or kits, we don't have a physical label. Is it ok to create a label and then store electronically and may be, even attach to each patient report?

Thank you so much!
 
When we talk about the “date of manufacture” under IVDR, this normally refers to when the device itself is made available for use. For example, in the case of a software as a medical device (SaMD), this would typically correspond to the software release date or to its last significant update. It would not change each time a patient sample is analysed or a report is generated.

As for the second question, it really depends on the nature of the IVD. If there is somehow a physical device or kit, the labelling information must be associated with that physical item and cannot exist only electronically or much less in the patient report. For the same SaMD, a controlled electronic label linked to the device documentation is generally acceptable, while the patient report should only reference the test or software version used.
 