Where to submit Adverse Events Reports in Europe

image182

Involved In Discussions
Hello all

I'm getting a bit confused with regards to where we need to submit adverse event reports in Europe....
  • We are a UK based company
  • BSI is our notified body
  • MHRA is our national competent authority
  • We sell our devices into multiple countries in Europe

As an example. if we had an event occur in France and we decided to submit a report, how many places would we be required to submit the report to?

Our notified body (BSI)
BSI have a submission portal and I believe we are required to submit the report to them. I'm not sure if this is something unique to BSI or all notified bodies require this?

Our national competent authority (MHRA)
MEDDEV 2.12 guides us to submit the report to the "national competent authority so I would have presumed this would be the only place we would be required to submit to

The competent authority in the country where the incident took place (ANSM)
We've been contacted by our distributor and they have advised us that the ANSM require the event to be reported to them

Other competent authorities who receive the device?
If the device is sold to various countries, Germany, Italy etc.... then would we also need to submit the report to these countries too?

Ideally I would only like to submit one report to one location... but i'm getting worried that each report would need to be submitted to a large number of locations and cause a considerable amount of work...

Thanks in advance for the help!
 

dgrainger

Trusted Information Resource
From the MEDDEV - P22:
5.1.8 TO WHOM TO REPORT
In general, the report should be made to the National Competent Authority in the country of occurrence of the INCIDENT unless specified differently in this guideline.
 

chris1price

Trusted Information Resource
Hi

If the incident occured in EU, it should be notified to the Competent Authority of the country in which the event occurred. This will be the full report of the incident and invenstigations. They will take the lead in the investigation. (see meddev 2-12-1 section 5.2)

If it occured outside of EU, then report to your national CA.

You also need to notify all other EU countries in which the same product was sold, and your national CA if you havnt already done so. A copy of MEDDEV FSCA form and FSN in appropriate language. (see meddev 2-12-1 section 5.4.4.1)

If you dont sell to a specific country and don't notify them, you will probably get requests for information anyway.

Don't forget any non eu countriles where the device may have been sold, USA, Australia, Canada, etc. Also dont be surprised when you get requests for information from other countries such as Hong Kong and Singapore.

Notified bodies vary in their requirments, BSI do want notification, others may not. Check your contract with them.

Unfortunately, there isnt a single sentral location to send notifications. Under the new regulations, one may be created, but done hold your breath.

Chris
 

image182

Involved In Discussions
If the incident occured in EU, it should be notified to the Competent Authority of the country in which the event occurred. This will be the full report of the incident and invenstigations. They will take the lead in the investigation. (see meddev 2-12-1 section 5.2)

If it occured outside of EU, then report to your national CA.

You also need to notify all other EU countries in which the same product was sold, and your national CA if you havnt already done so. A copy of MEDDEV FSCA form and FSN in appropriate language. (see meddev 2-12-1 section 5.4.4.1)

So following the MEDDEV and submitting to the competent authority where the incident occurred is not sufficient?

If we sold to every country in Europe, we would need to submit the report to every competent authority?
 

chris1price

Trusted Information Resource
I'm believe so, 5.4.4.1 of the meddev says......

"The MANUFACTURER should issue a notification (see below) to the Competent Authorities
of all countries affected at the same time and also to the National Competent Authority
responsible for the MANUFACTURER or AUTHORISED REPRESENTATIVE. Use the
format recommended in annex 4."

Annex 4 is the FSCA form. The good news is that I have always filled it out in English and have not been asked for a translation. The FSN (which is also sent to customers) should be translated.
 

image182

Involved In Discussions
I'm believe so, 5.4.4.1 of the meddev says......

"The MANUFACTURER should issue a notification (see below) to the Competent Authorities
of all countries affected at the same time and also to the National Competent Authority
responsible for the MANUFACTURER or AUTHORISED REPRESENTATIVE. Use the
format recommended in annex 4."

Annex 4 is the FSCA form. The good news is that I have always filled it out in English and have not been asked for a translation. The FSN (which is also sent to customers) should be translated.

Isn't section 5.4.4.1 related to FSCA and not applicable for general reporting where a FSCA is not required?

5.1.8 is related to reporting incidents and states
"In general, the report should be made to the National Competent Authority in the country of occurrence of the INCIDENT unless specified differently in this guideline."

This suggests only sending the report to one CA

However, section 5.4.3 which deals with the final report states
"The report is made by the MANUFACTURER to the National Competent Authority(ies) to whom the MANUFACTURER sent the initial report. "

This suggests that we could have submitted to multiple CA?
 

dgrainger

Trusted Information Resource
Yes, section 5.4.4 relates to Field Safety Corrective Action.

Section 5.1 relates to incident reporting.

You would only send a final report to multiple NCAs if you had reported incidents in those NCAs.
 
Thread starter Similar threads Forum Replies Date
I Where do we submit the FAA Preapplication forms? Federal Aviation Administration (FAA) Standards and Requirements 1
E Class II symbol - where to apply it? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T Good Documentation Practices (GDocP) - Where are these described in the regulations? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
A Where 60601-1 represented in the 510k submission? US Medical Device Regulations 6
Q Where did my Alignment go wrong? MCOSMOS Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
Q Where to place our Safety Program in our QMS? Occupational Health & Safety Management Standards 9
MaHoDie Where to show the Software version IEC 62304 - Medical Device Software Life Cycle Processes 2
E Where can i get latest toyota and honda Supplier Quality Assurance Manual (SQAM) Customer and Company Specific Requirements 3
I Where to find MDR compliance device? EU Medical Device Regulations 6
N Writing Risk Management procedure for small manufacturing and we don't know where to start. Manufacturing and Related Processes 9
I ISO 9000 Where to Start ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 93
Ajit Basrur Where to find a pagination requirement as "Page x of y"? ISO 13485:2016 - Medical Device Quality Management Systems 7
T Assessing risk where harm is indirect - Generic devices / accessories / intermediates ISO 14971 - Medical Device Risk Management 8
J Incoming Stock QC, where to start? Manufacturing and Related Processes 6
J Stock Traceability Methods, where to start? Manufacturing and Related Processes 0
shimonv Where do you place Human Factors report in a 510(k) application? Other US Medical Device Regulations 3
B Where do the definitions of the types of processes come from? IATF 16949 - Automotive Quality Systems Standard 9
Awais Where to get started on DFM / DFMA Manufacturing and Related Processes 2
S Where to start for MDSAP for Canada? Canada Medical Device Regulations 5
D Deviations - Where in ISO 13485 deviations are covered? ISO 13485:2016 - Medical Device Quality Management Systems 7
D Question regarding where "validations" fit according to ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
M Testing criteria - where to place Design and Development of Products and Processes 2
J UDI - Where to document it? EU Medical Device Regulations 8
J Where to get a Black belt certification ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 13
D Any recommendations on where to find experienced quality hires in Chicago area? Job Openings, Consulting and Employment Opportunities 2
J Where is the definition of a critical supplier? ISO 13485:2016 - Medical Device Quality Management Systems 5
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
B When Travel is allowed - Where are you going? Travel - Hotels, Motels, Planes and Trains 24
G Where to buy Template / sample SCSMS Manual, checklist Supply Chain Security Management Systems 1
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
C Where to draw the line for "sufficient evidence" to verify safety/performance of a device? CE Marking (Conformité Européene) / CB Scheme 2
L Where to place the what/when/how/who/where procedures in a multi-regulation AS9110C organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
B Where to acquire EN 868-5 required dye (Amaranth red)? Other Medical Device Related Standards 2
N Where to find Regulations for Reprocessing and Reuse of Single-Use Devices Other Medical Device Related Standards 2
H Where to get EK MED 1927/16 Other Medical Device Regulations World-Wide 3
D Where to find Guidance Class II Amendment Canada Medical Device Regulations 1
O EN 301 489-1 and EN 301 489-17 - Where do we get the information of the Published versions? CE Marking (Conformité Européene) / CB Scheme 1
A AS9100 Training - Where do I start? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
qualprod Records where apply and stop recording everything? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
G Gage R&R - Where am I going wrong? Part of a FAIR submission (Aerospace) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Where does IATF 16949 address Process mapping? IATF 16949 - Automotive Quality Systems Standard 3
C Where to place CE/label in Sterile IVD EU Medical Device Regulations 1
G New to ISO 9001 - Where to begin? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
K No IFU, where the requirements would go? EU Medical Device Regulations 5
T Assessing Hazard-Related Use Scenarios where control measures exist through standards IEC 62366 - Medical Device Usability Engineering 32
J Where to Place Process Maps in our Documentation? Process Maps, Process Mapping and Turtle Diagrams 4
GreatNate Where to get a KPI dashboard for QMS processes Quality Tools, Improvement and Analysis 5

Similar threads

Top Bottom