Dear All
In Thailand, we (mean many MR & me too, except Non-Thai) have a big problem concerning who should be the person autholizing-release of product for delivery to customer as mention in (ISO 9001:2008 & ISO 13485;2003) clause 8.2.4.
and what he should or shall do to release the product ( I beleve not only sign the delivery document...)
:frust:
DRDDO
 
Elsmar Forum Sponsor
P

Polly Pure Bread

#2
Dear All
In Thailand, we (mean many MR & me too, except Non-Thai) have a big problem concerning who should be the person autholizing-release of product for delivery to customer as mention in (ISO 9001:2008 & ISO 13485;2003) clause 8.2.4.
and what he should or shall do to release the product ( I beleve not only sign the delivery document...)
:frust:
DRDDO
authorities for the disposition of products after verification should be identified
 
J

JaneB

#3
who should be the person autholizing-release of product for delivery to customer as mention in (ISO 9001:2008 & ISO 13485;2003) clause 8.2.4.
and what he should or shall do to release the product ( I beleve not only sign the delivery document...)
I can't speak for 13485, but in 9001, it requires you to:
have records that indicate the person(s) authorizing release of product for delivery to the customer' 4.2.4

That means at least one record that shows at least one specific role/person (and who that was) decided that product X (or batch X of product) was OK to go.

Bottom line: before it leaves you who produced it and are selling it, you should have a way of demonstrating (here, to an external party) that it met all the requirements.

Who decides that? And what kind of record/s are needed? You get to determine that. But whoever it is should have the authority to decide it and be competent to make that decision, and have some kind of criteria on which that decision is made (as opposed to 'gut feel').

I mean, you could have the truck driver authorising release, but you'd have trouble convincing me (let alone anyone else) that a truck driver had the necessary competency (per 6.2 of ISO 9001) to distinguish between product that was OK to go and product that wasn't.

Some ways of achieving this (for example) might be to have inspection points along the way, with a signoff /acceptance of satisfactory pass on the final inspection being the record of release.

Signing a 'delivery docket' might be OK, but only if there were sound inspection processes that ensured that before it got to the delivery stage it had been inspected by someone competent and = OK.

But if it's simply a matter of something being moved across to despatch (without anyone checking final) and then the warehousing person or whoever sending it out... then that sounds a bit loose, to put it mildly, and not likely to meet the requirements.

Hope this helps.
 

Ajit Basrur

Staff member
Admin
#4
Dear All
In Thailand, we (mean many MR & me too, except Non-Thai) have a big problem concerning who should be the person autholizing-release of product for delivery to customer as mention in (ISO 9001:2008 & ISO 13485;2003) clause 8.2.4.
and what he should or shall do to release the product ( I beleve not only sign the delivery document...)
:frust:
DRDDO
Since this question is raised in the ISO 13485 forum, the answer is "Quality Assurance".

They will gather all the pertinent documents vix. Manufacturing records, Quality Control testing reports, all Warehousing particulars on raw materials and packaging materials and then RELEASE the batch that meets the customer specifications.
 
D

DRDDO

#5
I can't speak for 13485, but in 9001, it requires you to:
have records that indicate the person(s) authorizing release of product for delivery to the customer' 4.2.4

That means at least one record that shows at least one specific role/person (and who that was) decided that product X (or batch X of product) was OK to go.

Bottom line: before it leaves you who produced it and are selling it, you should have a way of demonstrating (here, to an external party) that it met all the requirements.

Who decides that? And what kind of record/s are needed? You get to determine that. But whoever it is should have the authority to decide it and be competent to make that decision, and have some kind of criteria on which that decision is made (as opposed to 'gut feel').

I mean, you could have the truck driver authorising release, but you'd have trouble convincing me (let alone anyone else) that a truck driver had the necessary competency (per 6.2 of ISO 9001) to distinguish between product that was OK to go and product that wasn't.

Some ways of achieving this (for example) might be to have inspection points along the way, with a signoff /acceptance of satisfactory pass on the final inspection being the record of release.

Signing a 'delivery docket' might be OK, but only if there were sound inspection processes that ensured that before it got to the delivery stage it had been inspected by someone competent and = OK.

But if it's simply a matter of something being moved across to despatch (without anyone checking final) and then the warehousing person or whoever sending it out... then that sounds a bit loose, to put it mildly, and not likely to meet the requirements.

Hope this helps.
Perfect anwer
:thanks:
DRDDO
 
S

SteveK

#6
Since this question is raised in the ISO 13485 forum, the answer is "Quality Assurance".

They will gather all the pertinent documents vix. Manufacturing records, Quality Control testing reports, all Warehousing particulars on raw materials and packaging materials and then RELEASE the batch that meets the customer specifications.
I would beg to differ on this. My thought was indeed that "Quality" was the final sign off for release of product (medical devices). However my auditor decided this was not the case, it was the person in charge of the production department because he was responsible for the shipping operation. Nearly got a NC for this.:2cents:
 

Ajit Basrur

Staff member
Admin
#7
I would beg to differ on this. My thought was indeed that "Quality" was the final sign off for release of product (medical devices). However my auditor decided this was not the case, it was the person in charge of the production department because he was responsible for the shipping operation. Nearly got a NC for this.:2cents:
The FDA (Food Drug Administration) requires the release of bacthes from the Quality unit. :)
 
A

Allan66

#8
I would beg to differ on this. My thought was indeed that "Quality" was the final sign off for release of product (medical devices). However my auditor decided this was not the case, it was the person in charge of the production department because he was responsible for the shipping operation. Nearly got a NC for this.:2cents:
It is always a difficult area when an auditor has a strong opinion which differs from that of the client - or in the case of products being sold to multiple companies - other auditors. In these cases it is often useful to go back to the specification defining the requirements. With the 'new' answer, I would quite strongly question how the person in charge of a production department head could be considered to have the necessary degree of independence to authorise a final release, so I am with Ajit on this one. Do you know how experienced your auditor was in this field?
 
C

Chris Ford

#9
The FDA (Food Drug Administration) requires the release of bacthes from the Quality unit. :)
Actually, the Quality System Regulation does not require a quality function to be responsible for final release of product. The requirements are generally the same as 13485.

21CFR820.80(d)Final acceptance activities. Each manufacturer shall establish and maintain procedures for finished device acceptance to ensure that each production run, lot, or batch of finished devices meets acceptance criteria. Finished devices shall be held in quarantine or otherwise adequately controlled until released. Finished devices shall not be released for distribution until:
(1) The activities required in the DMR are completed;
(2) the associated data and documentation is reviewed;
(3) the release is authorized by the signature of a designated individual(s); and
(4) the authorization is dated.
(e)Acceptance records. Each manufacturer shall document acceptance activities required by this part. These records shall include:
(1) The acceptance activities performed;
(2) the dates acceptance activities are performed;
(3) the results;
(4) the signature of the individual(s) conducting the acceptance activities; and
(5) where appropriate the equipment used. These records shall be part of the DHR.
 
Q

Qualqueen

#10
Dear All
In Thailand, we (mean many MR & me too, except Non-Thai) have a big problem concerning who should be the person autholizing-release of product for delivery to customer as mention in (ISO 9001:2008 & ISO 13485;2003) clause 8.2.4.
and what he should or shall do to release the product ( I beleve not only sign the delivery document...)
:frust:
DRDDO
We're also an ISO 9001 / 13485 company and what JaneB states is indeed the answer.

Some ways of achieving this (for example) might be to have inspection points along the way, with a signoff /acceptance of satisfactory pass on the final inspection being the record of release.

We didn't name any specific person or department. We simply stated records provide the evidence of product conformance, indicating the person authorizing the release of the product.

In our cross reference table for 8.2.4 are our WI's for inspection activities, sampling plan and so on with a needed sign-off for each phase of the product. By the time it's in shipping, all of the production has been inspected at incoming, inprocess and final along with an inspection card of acceptance. A traveler and copy of the process sheet accompany finished product to shipping and they sign off. This is all kept on file and that's our record of release....in one neat bundle without naming one specific person.
:magic:
 
Thread starter Similar threads Forum Replies Date
Anerol C Where should I start as the person responsible for the calibration laboratory? General Measurement Device and Calibration Topics 7
Ajit Basrur Should the Protocol and Report be signed by the same person(s)? Qualification and Validation (including 21 CFR Part 11) 7
S FDA approval (Med Device) - Should we use Accredited Person or proceed on Our Own? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
C Hopefully supply to GM - What should a small 4 person company do IATF 16949 - Automotive Quality Systems Standard 8
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance standard to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A Should I take an online course for a career in Occupational Health and Safety? Career and Occupation Discussions 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
K Should APQP/PPAP has its own section in a QM? Quality Management System (QMS) Manuals 1
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
P Should eIFU link per ISO 15223-1:2016 be added to labels out of scope of Reg 207/2012? EU Medical Device Regulations 1
S Which Sampling Plan(s) Should I Use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
A Document release vs its related training. Which should come first? ISO 13485:2016 - Medical Device Quality Management Systems 18
S Which department should prepare the control plan? could you show me a standard regarding to this matter. FMEA and Control Plans 17
J Help settle a disagreement: Should external providers of preventive maintenance be on your ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Master Samples - What should we be keeping? IATF 16949 - Automotive Quality Systems Standard 9
G Supplier delivered recent PPAP, should he deliver yearly layout inspection? IATF 16949 - Automotive Quality Systems Standard 4
John Broomfield Vote - Should ISO9004 Become a Requirements Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
A Capability Study - in the beginning of your career what should you have known about the tool Quality Tools, Improvement and Analysis 11
J Should Loading and Unloading be Included in Cycle Times? Lean in Manufacturing and Service Industries 14
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0
T 510(k) submission - Which name should I use in the submission? Other US Medical Device Regulations 3
N ISO 19011:2018 - 5.4.2 "...audit program should engage in appropriate continual development..." Training - Internal, External, Online and Distance Learning 4
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
DuncanGibbons Should the requirements FAA/EASA Part 21 be addressed within the QMS and AS9100D quality manual? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Should 510(k) Predicates be Actively Listed Devices? Other US Medical Device Regulations 12
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
J On PFMEA for danger labels - Label always should be assigned severity 10 ? FMEA and Control Plans 3
H Who should be listed as the manufacturer/distributor on the box? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 3
BeaBea How Many Processes should be created for each Department? Process Maps, Process Mapping and Turtle Diagrams 5
M Should volume of sales be factored into risk probability assessments? ISO 14971 - Medical Device Risk Management 33
MrTetris Should potential bugs be considered in software risk analysis? ISO 14971 - Medical Device Risk Management 5
S Should safety checks be included in the Control Plan? IATF 16949 - Automotive Quality Systems Standard 5
M Which incubation condition should be selected to recover both bacteria and fungus effectively Miscellaneous Environmental Standards and EMS Related Discussions 3
D Is there a specific location for PPE such as safety glass holders and glove dispensers should be mounted Occupational Health & Safety Management Standards 10
Robert Stanley Which Registrar Should I Choose for ISO 9001:2015 registration? Registrars and Notified Bodies 10
M Who should receive the bills from suppliers and vendors, account payable or procurement? Consultants and Consulting 4
V IATF 16949 8.4.1 Control of externally provided processes, products and services - Should the CB be on our Approved Supplier List? IATF 16949 - Automotive Quality Systems Standard 10
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Received a Major finding during IATF Surveillance audit for loss of BIQS Level 3 (more than 6 SPPS in 6 months)...how should we address SYSTEMIC CA? IATF 16949 - Automotive Quality Systems Standard 11
J Organization merger. Should we keep two separate ISO 13485 certificates? ISO 13485:2016 - Medical Device Quality Management Systems 6
S Companies that maintain your machine should be in ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Use of "Shall" versus "Should" in Procedures ISO 13485:2016 - Medical Device Quality Management Systems 26
D Class II medical device - When should a complaint be closed? Customer Complaints 6
Sidney Vianna IATF 16949 News Presentations from the latest IATF Stakeholder Event - Expectation that IATF 16949 certification should equate with product quality. Misguided? IATF 16949 - Automotive Quality Systems Standard 7
L Clause 0.4 of ISO 9001 and EHS - Where should I stop the inclusion of EHS in my QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Ed Panek Part 11 Self Certify Memo - What else should it cover? Qualification and Validation (including 21 CFR Part 11) 5
H Should I mention machine/Equipment password In SOP? Qualification and Validation (including 21 CFR Part 11) 4
D How long should we keep the spare parts available for our medical device, after we have stopped the production? ISO 13485:2016 - Medical Device Quality Management Systems 0
H Statistical Techniques Procedure - What should be included Document Control Systems, Procedures, Forms and Templates 4

Similar threads

Top Bottom