Why was Drip-Rate Infusion Pump removed from IEC 60601-2-24:2012?

Roland chung

Trusted Information Resource
If anyone knows why drip-rate infusion pump was removed from the IEC 60601-2-24:2012? Does it mean that the accuracy test for drip-rate mode is not necessary any more?

Thank you!
 

Peter Selvey

Leader
Super Moderator
It's not clear and there seem to be some issues with the upgrade from the 1998 edition to 2012 edition. For example, Type 5 has been deleted from Note 1 definition but still referred to in the accuracy tests

(and by the way, it never made sense to refer to Type 1 - 5 since this was a "note" and yet used as a formal definition in the text).

I believe there are some other errors and confusing points in the standard which pop up in practical application.

Beyond that there is the whole issue of syringe pump testing which is extremely weakly dealt with (huge variations in performance with different syringes, gets hidden as the manufacturer can choose which data to display in the operation manual).

Expect a bit of an overhaul in the future.

This might also be the reason why it is yet to be adopted in Europe as a harmonized standard
 

Roland chung

Trusted Information Resource
If a nurse used a non-compatible syringe and did not follow the IFU to perform calibration prior to infusion and then got a bad accuracy, this should have nothing to do with manufacturer.
 

Peter Selvey

Leader
Super Moderator
Things might have improved (I haven't been involved in testing recently), but my experience with syringes is they have a huge amount of variable stiction, the stop start motion, where the syringe sticks for a few minutes and suddenly jumps forward with bolus. It was also worse with some brands compared to others.

The standard says you need to put the performance data in the operation manual, but there is no criteria or requirement to select data that is representative.

As such, it was common practice among manufacturers to wait for a good result and use that one in the operation manual.

Actually this is not just for syringe pumps, I found other examples of infusion pump manufacturers selecting "optimistic" data, and it is common in other medical devices to fudge the data.

The difference is that the variability with syringe pumps can be huge and clinically significant.

For my side, the standard should require enough samples to show for example a trumpet curve that covers 95% of cases. It's an issue raised in the rationale, just not mentioned in the normative text.
 
Thread starter Similar threads Forum Replies Date
E MDCG 2023-3 - is it me or was the rug just pulled? EU Medical Device Regulations 7
M How to 'delist' a Class II device that was incorrectly registered as such? Medical Device and FDA Regulations and Standards News 2
K Someone asked me this morning what a pulse oximeter was? Other Medical Device and Orthopedic Related Topics 2
Sidney Vianna Informational Changes to Terms and Definitions in ISO/TC 176 Standards since ISO 9000:2015 was issued. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
D PCBA was swollen after SMD Manufacturing and Related Processes 12
M Are you allowed to reuse the same asset number if the device was replaced? General Measurement Device and Calibration Topics 10
L Significant Production Run - How 300 was determined to be the minimum quantity APQP and PPAP 2
K OEM specific requirements - PPAP was rejected by STA Customer and Company Specific Requirements 5
K %GRR was between 10-30% so we have to have a "backup plan" per auditor IATF 16949 - Automotive Quality Systems Standard 15
I When exactly was "competence" added to the ISO-9001 standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
I "On the job training" as a response to how someone was trained? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 92
T Was Just told we could ONLY buy material for AS9100 customers if the DISTRIBUTION house is AS9100 / AS9120 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
U Own Procedure was not effectively implemented Problem Solving, Root Cause Fault and Failure Analysis 3
B How to reply NCR on ineffectiveness of corrective action during IATF external audit? This is repeated issue whereby some mistake was done. IATF 16949 - Automotive Quality Systems Standard 7
R Major nonformance finding was given during a closing meeting of a ISO9001 certification audit General Auditing Discussions 76
M Case study - If the restaurant (ISO 9001:2015 certified) was run by 2 persons covering cooking and purchasing processes (Mother and Father) supported ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Ed Panek ISOTECH TTI-10 found out of calibration - Calibration house was wrong? General Measurement Device and Calibration Topics 7
E Audited to ISO 13485 standard although customer was aware we're not and have no plans to be ISO 13485:2016 - Medical Device Quality Management Systems 23
H Claiming equivalence with the product that was sold earlier in the market but not now EU Medical Device Regulations 1
E Interesting Discussion Was this the fastest ever stage 2 ISO9001:2015 audit? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 76
M Products Manufactured while CE was valid CE Marking (Conformité Européene) / CB Scheme 1
S Corrections and Removals - Fiber Content in our textile products was incorrect Other US Medical Device Regulations 1
qualprod Risk closeout , mitigation was not effective, next? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
J AS9100D Clause Brain Fade - Tool was past due for Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
P Retired equipment that was never validated Qualification and Validation (including 21 CFR Part 11) 2
Jane's What steps do you take to terminate a product which was licensed for sale in Canada Canada Medical Device Regulations 14
Q Change revision number in document when only codification was changed Document Control Systems, Procedures, Forms and Templates 8
Marc Before the 760mph Hyperloop dream, there was the atmospheric railway Coffee Break and Water Cooler Discussions 1
Ninja Getting eaten by Big Business - My company was bought by a huge multinational Coffee Break and Water Cooler Discussions 54
M When was ISO started and Where? Other ISO and International Standards and European Regulations 3
J Dinged on Internal Audits for supervising an auditor I was training Internal Auditing 10
Q 21CFR820 as it was in 1993 or 1997? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R 5.4.1 Quality Objectives - Nonconformity: The process was not fully effective ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 36
C CB Accreditaton - CB that was accredited by IAS Registrars and Notified Bodies 5
GStough Rx-360 - Has Anyone Used This and What Was Your Experience? Supplier Quality Assurance and other Supplier Issues 3
R IEC 62304 was brought up during an FDA Inspection/Audit IEC 62304 - Medical Device Software Life Cycle Processes 6
M I was wondering what happened to them (Joke) Funny Stuff - Jokes and Humour 5
U Major Nonconformance - We had a control in place that was not on the control plan IATF 16949 - Automotive Quality Systems Standard 14
J What are the changes in CMDR that was revised in Dec 2013? Canada Medical Device Regulations 2
W The 2nd printing of the AWS D17.1 2010 which was 2011 - Page missing? Various Other Specifications, Standards, and related Requirements 3
T Suppose Aristotle was holding a conference Brain Teasers and Puzzles 10
P ISO 15197:2013 (Blood glucose monitoring system) was released in May Other US Medical Device Regulations 10
B Was EN 980:2008 superseded by EN 15223-1:2012 on January 31, 2013 CE Marking (Conformité Européene) / CB Scheme 2
Q Reduced Device License Fee was rejected for being sent in too early Canada Medical Device Regulations 2
M AS9100A (old revision) - Required C=0 sampling plans if sampling was used? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
S The days before International Standards - Has ISO done what it was supposed to? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
R Internal Auditing Checklist - Major NCR because the Checklist was not Completed Internal Auditing 17
Richard Regalado ISO 22301 was published 2012 May 15 - New Standard for BCM - Published! Business Continuity & Resiliency Planning (BCRP) 4
Q I was assigned as Kaizen Leader for our MRB Area - Your advice appreciated Quality Tools, Improvement and Analysis 5
R New to MSA - I was told to use ANOVA for the GRR's Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5

Similar threads

Top Bottom