In 3rd edition, single momentary flashover is not regarded as insulation breakdown during carrying out Hi-Pot test. How about multi-time flashovers?


During the test, breakdown constitutes a failure. Insulation breakdown is considered to have occurred when the current which flows as a result of the application of the test voltage rapidly increases in an uncontrolled manner, that is, the insulation does not restrict the flow of the current. Corona discharge or a single momentary flashover is not regarded as insulation breakdown.​
 

Peter Selvey

Leader
Super Moderator
The "Hi-Pot" test is intended to verify solid insulation. If the corona discharge has nothing to do with the insulation being tested, you can ignore it.

At voltages above 3000Vpeak, corona is common and may occur away from the actual solid insulation under test, especially if care is not taken with the set up.

However, there is a possibility that corona / flashover is related to the insulation partially breaking down and then recovering. For this it is important to make sure to use the 100mA supply as required by CB scheme/CTL, to avoid that the recovery is limited current capacity in the hi-pot supply.

In general though, it is normally possible to design solid insulation that is >> requirement, so corona/flashover is a test set up issue.
 

Roland chung

Trusted Information Resource
Thanks, peter.

The continuous flashover occurred in transformer's core when hi-pot test was carried out between input and output of switching power supply. The applied voltage was 4 kV, the test instrument didn't fail the test since the current was only 2 mA (tripping current was set to 10 mA). Did you mean that the test result should be judged based on the current only?
 

Peter Selvey

Leader
Super Moderator
Almost certainly a set up issue, you rarely see 2mA at 4kV if the set up is correct (but often see if the set up is wrong).

The high leakage indicates a side path is getting picked up, probably via EMC capacitors, from secondary to earth which is the typical case.

This usually means one or both of the following:
1) areas not designed or required for 4kV get overstressed (hence the flashovers)
2) the area designed for 4kV is not being stressed/tested
 

Peter Selvey

Leader
Super Moderator
I should distinguish between corona and flashover, I assumed at the start you were talking about corona.

Corona can occur around a surface near an electrode without actually bridging to the other electrode. It's just the point where the local field strength is higher than the molecular O2 bonds. Near an electrode the field strength (V/m) can be high, but it usually drops off quickly as you move away. As such steady corona (light buzzing sound, smell of ozone) is normal for the 4kV test, and can be ignored.

A big flashover is different and needs investigation. Flashover is a full path from two sides. Because of the high cr/cl limits in IEC 60601-1 (e.g 8.0/5.0mm for 2MOPP) ), flashover is not really possible in normal testing.

So, if it occurs it is likely to be a set up issue.

If one were to insist that the set up correct and flashover still occurs I would fail it as the standard says only "single" flashover is allowed.
 

Marcelo

Inactive Registered Visitor
AS defined in the standards, "breakdown is considered to have occurred when the current which flows as a result of the application of the test voltage rapidly increases in an uncontrolled manner, that is, the insulation does not restrict the flow of the current".

So, you have to verify the current and, if it flows with our restriction, it's a breakdown.

Just take care with incremental flow due to capacitance, if your hipot does not have enough power and have a limit to current (most have to avoid hipot damage), it may trip due to the capacitance (this is easily checked, if you increase the voltage and there's a related increase in current, and when your step a step of the voltage and the current also stops, this is a capacitance problem, not a breakdown - in this case you probably forgot to remove the capacitance from the tests).
 

somashekar

Leader
Admin
With a valid setup, can the flashover be the extension of the corona effect perhaps due to faster rate of voltage ramp up or high humidity in the test place ....
 

Peter Selvey

Leader
Super Moderator
You would need to be pretty quick with the dial to have any concern about the ramp up effects. The only risk would be if the voltage was already set up and then applied by a switch, that could cause an issue.

The breakdown voltage in air is around 3kV/mm, this means 5mm clearance requires about 15kVpeak. The 4kVrms test is about 5.6kVpeak, well below this level. Influence of humidity is around 2% from 40-80% and actually higher humidity means higher breakdown voltage (it's counter-intuitive, but fairly well documented). Different geometry such as sharp points can influence, but sharp points are rare and probably not enough.

If the dielectric strength test is not done in the RH chamber, condensation can cause breakdown along any surfaces. But this is not a failure as it is unrelated to the solid insulation under test. It is just lazy testing :).

As already stated, far more likely is a set up issue where the volts are being transferred to a location not required or designed for the 4kV.

If you read the standard carefully, solid insulation is best removed from the equipment to avoid stray path issue and to ensure the targeted insulation is actually being tested. For example, I would normally remove a switching power supply PCB from the equipment and any metal frame and then solder in test wires to the primary and secondary, inspect to make sure all parallel paths are removed, and then put that into the chamber. If you do that, you will never get any flashover.

I very (very) rarely do 4kV test on a fully assembled device. It's almost impossible to eliminate the stray paths and make sure the target insulation is is tested. Rare cases are Class II devices where the lack of earthing eliminates many stray path, but even then not always.
 

Marcelo

Inactive Registered Visitor
The only risk would be if the voltage was already set up and then applied by a switch, that could cause an issue.

Yeah, that's the case I was thinking, for automated testers. But in this case there would not a corona in fact, only a tripping.

As already stated, far more likely is a set up issue where the volts are being transferred to a location not required or designed for the 4kV.

If you read the standard carefully, solid insulation is best removed from the equipment to avoid stray path issue and to ensure the targeted insulation is actually being tested. For example, I would normally remove a switching power supply PCB from the equipment and any metal frame and then solder in test wires to the primary and secondary, inspect to make sure all parallel paths are removed, and then put that into the chamber. If you do that, you will never get any flashover.

I very (very) rarely do 4kV test on a fully assembled device. It's almost impossible to eliminate the stray paths and make sure the target insulation is is tested. Rare cases are Class II devices where the lack of earthing eliminates many stray path, but even then not always.

This is a common problem in some places (it is in Brazil, at least), when manufactures do not clearly state insulations and then declare to put the "4k from main to AP", but in fact there is a double insulation, and the tester simply put 4k from one to the other. This goes back to the comment in insulation diagrams not bring clear enough.
 
Thread starter Similar threads Forum Replies Date
P Increase in audit days for multi-site companies based on the new AS9104-1 2022 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
D Control Limit Calculation with multi-cavities molding Statistical Analysis Tools, Techniques and SPC 5
K "Supplier's multi-disciplinary approach to decision making" IATF 16949 - Automotive Quality Systems Standard 4
R MSA Multi System Correlation Capability, Accuracy and Stability - Processes, Machines, etc. 21
F Multi-Sensor CMM Measurement General Measurement Device and Calibration Topics 5
A ISO 9001- any advise please with transitioning from a paper based system to eQMS and multi site scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
xfngrs Multi-lingual documentation IATF 16949 - Automotive Quality Systems Standard 6
eldercare Multi-Site ISO/AS Certification Requirement for some sites ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 22
B Site removal from MDSAP multi-site Certificate ISO 13485:2016 - Medical Device Quality Management Systems 4
L Where to place the what/when/how/who/where procedures in a multi-regulation AS9110C organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
M Sites to List on Multi-Site Certifications Registrars and Notified Bodies 1
D How to list multi-product sample pack in GUDID 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
F 5520A High Performance Multi-Product Calibrators General Measurement Device and Calibration Topics 0
S REF and CE mark symbols on multi-component device EU Medical Device Regulations 3
M Inputs on definition of very similar processes for multi site audit sample - IAF MD1 2018 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
B Multi-site certification (IAF MD1) - for 9001, 14001, OHSAS/45001 ISO 14001:2015 Specific Discussions 19
F Business expanding - Campus vs Multi Site Certification Structure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Q Controlling Multi-Tab Microsoft Excel Forms Excel .xls Spreadsheet Templates and Tools 5
G Integrated ISO EHSQ Multi-Site Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
M Multi-Site Umbrella Certification Transition Plan ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
J Multi-level QMR Process Management Review Meetings and related Processes 6
L Corporate Manual - Global Multi-Site ISO9001:2015 Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
A One site of multi-site company closing - Quality transfer issues Quality Manager and Management Related Issues 7
B External Label Content of multi-device (sterile and nonsterile) EU Medical Device Regulations 3
L IMDS for a Part of a Multi-Part Assembly? RoHS, REACH, ELV, IMDS and Restricted Substances 10
S Understanding Subgroup Size - Multi Cavity (Minitab) Statistical Analysis Tools, Techniques and SPC 4
L Automate GR&R using Minitab on complex, multi-measurement assemblies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
L Responsibility for Labeling in Multi-Location Group Document Control Systems, Procedures, Forms and Templates 5
S MSA Help or guidelines - Plastics - Setting up an MSA on a multi cavity fixture Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
O Transitioning Locations into a Multi Site ISO 9001 Registration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
K Validation of an Analytical Method that relies on Multi-Variate Calibration General Measurement Device and Calibration Topics 7
G Multi Site EMS and QMS Management System Implementation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
T Single or Multi site Internal Audits - Any difference in how to carry out? Internal Auditing 2
G Internal Auditing in a Multi-Site Environment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
V Do we require impurities to be monitored for multi-vitamin products? US Food and Drug Administration (FDA) 5
A QMS and Certification in a Multi-Company/Group Environment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
N How to qualify (IQ, OQ, PQ) a multi-purpose equipment ? Imported Legacy Blogs 1
S How to implement protection for multi-winding transformer? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P How to streamline different documents in a multi plant ISO13485 QMS ISO 13485:2016 - Medical Device Quality Management Systems 8
A Multi-Site Certification to ISO 9001, 14001 and OHSAS 18001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
J How to qualify (IQ, OQ, PQ) a multi-purpose equipment ? Qualification and Validation (including 21 CFR Part 11) 5
0 Gage R&R Standard for Multi Location Fixtures Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
S Integrating a Multi-Management System - What documents & records must we have? Various Other Specifications, Standards, and related Requirements 6
R Mix Multi-Site/Individual Certification - Has this ever happened to anyone? Miscellaneous Environmental Standards and EMS Related Discussions 14
W Digital Multi-Meter Calibration Procedure for various Meter Brands and Functions General Measurement Device and Calibration Topics 7
G Multi Language Instructions - Standard for Type Size when Printing Instructions? Other ISO and International Standards and European Regulations 2
M Multi-factor Anova with Minitab Using Minitab Software 1
T DOE Multi level Taguchi - I am doing a Capstone Project Using Minitab Software 10
B Multi Cavity Plastic Injection Mould SPC Statistical Analysis Tools, Techniques and SPC 4
S Multi-Vari Help on Experiment Results Six Sigma 1

Similar threads

Top Bottom