Hi,
Pardon my naivet?. I am pretty much all self taught at this stuff.
My small biomedical engineering company in Hanoi, Vietnam is testing our CPAP for CE marking as our first device.
I was wondering if we are required to test the device for xray radiation in accordance with clause 10.1.1 of IEC 60601-1, since our device is not intended to produce xray radiation. I am not sure whether it "might" produce xray radiation or not.
Also, I was wondering if we need to do an infrared radiation test in accordance with 10.6. The device has heaters for tubing, an applied part and for humidification, not applied. If we need one, how is this different from temperature test in accordance with risk management.
Take care
Elsmar Forum Sponsor
#2
If you have a look at the informative annex A, with regard to 10.1.1, it identifies VDUs as being a possible source of unintentional x-ray radiation. If you do not have such a component (or similar) then I think you would be justified as claiming this clause as 'Not Applicable'. That's what I do!
For 10.6, the requirement is for infrared to be considered in your risk process. So, if you can say that the infrared is separated / shielded / too low power for any harm to come to the patient then you probably have it covered.
Thread starter Similar threads Forum Replies Date
alonFAI Determining PCBA Xray Test Sampling Size Statistical Analysis Tools, Techniques and SPC 1
S Nadcap audit May 18 - May 21 - NDT: PT/XRay and Chemical Processing AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
S CSA Requirements - CMDCAS - ISO 13485 - Medical Xray camera for Canada ISO 13485:2016 - Medical Device Quality Management Systems 7
D Gage R&R for an Xray instrument for measuring plating thickness Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
N Infrared Tachometer vs. Laser Non-Contact Tachometer vs. Stroboscope - Differences Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
D Control of monitoring and measuring equipment - infrared thermometer for cross-check ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
L Creating Meaurement Uncertainty (MU) Budget for Infrared Thermometers Measurement Uncertainty (MU) 1
Adriane Need Help to Verify an Infrared Cure Oven General Measurement Device and Calibration Topics 9
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
M Informational USFDA – Radiological Health Regulations; Amendments to Records and Reports for Radiation Emitting Electronic Products; Amendments to Performance Stand Medical Device and FDA Regulations and Standards News 0
M FDA News USFDA Draft Guidance – Clarification of Radiation Control Regulations for Manufacturers of Diagnostic X-Ray Equipment Medical Device and FDA Regulations and Standards News 0
B Radiation Sterilization Dose Audit: Bioburden frequency Other Medical Device Related Standards 1
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
J Impact using two different materials for product enclosure in radiation and immunity IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Y ISO or BS standard for Measurement of Occupational Exposure to Ionization Radiation IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B CE Marking - Lead Aprons - Protection against diagnostic medical X-radiation CE Marking (Conformité Européene) / CB Scheme 5
Y User Manuals for Radiation-Emitting Electronic Products US Food and Drug Administration (FDA) 0
J FDA Launches Medical Device and Radiation-Emitting Product Transparency Web site US Food and Drug Administration (FDA) 2
Wes Bucey Cell phone radiation? After Work and Weekend Discussion Topics 6
S Radiation Survey per AMS2750D - Looking for a supplier Ontario/Quebec Manufacturing and Related Processes 1
R Customer Radiation Requirement headache RoHS, REACH, ELV, IMDS and Restricted Substances 7
M Radiation (Gamma or E-beam) Dose Audit SOP - Quarterly dose audits based on ISO 11137 Other Medical Device Related Standards 8
S Do any of the certification bodies specialize in radiation measurement testing? General Measurement Device and Calibration Topics 4
M IEC 60601-1-3 2nd Edition (X-ray Radiation Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
C New Danger Sign For Radiation Coffee Break and Water Cooler Discussions 2
A ISO17025 Quality Manual - Calibration - Radiotherapy and radiation protection ISO 17025 related Discussions 0
Similar threads
Top Bottom