In the IEC 60601-1-8, it can be understand in a way that the requirements are mandatory.
However, in the Subclause 6.3.3.1 of the same standart it is clearly stated that "The MANUFACTURER can provide more than one set of auditory ALARM SIGNALS. VALIDATION by USABILITY testing is not required if each set complies with Table 3 and Table 4 (or Annex F). If additional non-standard auditory ALARM SIGNAL sets (i.e., those that do not comply with Table 3 and Table 4 or Annex F) are provided, they require clinical VALIDATION to ensure that they provide at least an equivalent degree of safety as the standard sounds. Permission to provide non-standard sounds is intended to allow a RESPONSIBLE ORGANIZATION to continue to use non-standard but “historically validated” sound sets that have been successfully used for significant periods of time in their PATIENT care areas, and to ensure that this collateral standard is not excessively design restrictive."
I understand that if following the standart is mandatory unless the auditory alarm is clinically validated and tested for usability.
Could anyone please enlighten me?
However, in the Subclause 6.3.3.1 of the same standart it is clearly stated that "The MANUFACTURER can provide more than one set of auditory ALARM SIGNALS. VALIDATION by USABILITY testing is not required if each set complies with Table 3 and Table 4 (or Annex F). If additional non-standard auditory ALARM SIGNAL sets (i.e., those that do not comply with Table 3 and Table 4 or Annex F) are provided, they require clinical VALIDATION to ensure that they provide at least an equivalent degree of safety as the standard sounds. Permission to provide non-standard sounds is intended to allow a RESPONSIBLE ORGANIZATION to continue to use non-standard but “historically validated” sound sets that have been successfully used for significant periods of time in their PATIENT care areas, and to ensure that this collateral standard is not excessively design restrictive."
I understand that if following the standart is mandatory unless the auditory alarm is clinically validated and tested for usability.
Could anyone please enlighten me?