Basic UDI and UDI-DI for hip implant systems

ab1992

Registered
Hello all,

I'd like to ask for some help/advice trying to figure out basic udi and udi-di's for a hip implant acetabular system. The various components (shells, liners, femoral heads, associated instruments) all have their own UDI-DI. Presumably each type of component would need to have it's own basic udi-di due to differences in manufacturing characteristics, intended use etc? Or would the acetabular system just need to have a single basic udi, covering all of those separate components?

The acetabular system has a single technical documentation file under MDD with a separate technical documentation file covering the reusable instruments and this plan would be the same for EU MDR.
 

sibelyum

Registered
Hello,
All components with different production processes and different usage purposes must have the basic UDI-DI code separately.
Each technical file does not have to have a single basic UDI-DI code, it is necessary to evaluate each component of the product separately and derive the code.

I wish you success in your work.
 