I've some doubts concerning some BS EN standards and I couldn't think of a better way to get help, other than seeking the expertise of my fellow clovers here. Just a brief info: our company produces medical devices to be sold in US and EU.

1. BS EN ISO 14971:2012 (Deviation #3: ALARP vs AFAP)
- We're using a template for our RMF. The definition for "ALARP" as per ISO 14971:2007 is still considered acceptable. But as for EN ISO 14971:2012 is concerned, can we associate ALARP with AFAP? (meaning whichever risks that fall under AFAP are considered as ACCEPTABLE - provided we?ve given all means of risk reduction that we could think of)

2. BS EN 62366:2008
- We've been told by our Technical File's reviewer that not having this will cause us to have a major NCR.
- Since this is something new for us. We currently don?t have an SOP for this. Do we need one solely for this or can we incorporate it in our RMF procedure?

3. BS EN 15223-1:2012
- A feedback from our recent Technical File review, we were told that the "Single Use Only" symbol is insufficient to state that product is Not For Reprocess. And since I can't find any "Non-reprocessable" symbol in the standard, will the use of both "Single Use Only" and "Do Not Resterilize" constitute to the meaning?

Thanks in advance, guys! Have a blessed week ahead! :)
 
Elsmar Forum Sponsor

renenatasha

Involved In Discussions
#2
BS EN standards

Hi fellow 'covers'!
First and foremost, Happy New Year to all.

I'd like to seek your help concerning some doubts I have over BS EN standards and I couldn't think of a better way to get help, other than seeking the expertise of my fellow 'covers' here. Just a brief intro: our company produces medical devices to be sold in US and EU.

1. BS EN ISO 14971:2012 (Deviation #3: ALARP vs AFAP)
- We're using a template for our RMF. The definition for "ALARP" as per ISO 14971:2007 is still considered acceptable. But as far as EN ISO 14971:2012 is concerned, can we still associate ALARP with AFAP? (meaning whichever risks that fall under AFAP are considered as ACCEPTABLE - provided we’ve given all means of risk reduction)

2. BS EN 62366:2008
- We've been told by our Technical File's reviewer that not having this (in the ER) will result in a major NCR.
- Since this is something new for us, we currently don’t have an SOP for this. Do we need one solely for this or can we incorporate it in our RMF procedure?

3. BS EN 15223-1:2012
- A feedback from our recent Technical File review, we were told that the "Single Use Only" symbol is insufficient to state that product is Not For Reprocess. And since I can't find any "Non-reprocessable" symbol in the standard, will the use of both "Single Use Only" and "Do Not Resterilize" constitute to the meaning?

Thanks in advance, guys! Have a blessed week ahead! :)
 
Last edited by a moderator:
S

SteveK

#3
Re: BS EN standards

2. BS EN 62366:2008
- We've been told by our Technical File's reviewer that not having this (in the ER) will result in a major NCR.
- Since this is something new for us, we currently don?t have an SOP for this. Do we need one solely for this or can we incorporate it in our RMF procedure?
As per my RMF I just use a template for Useability (not an SOP) for my TF.
Have a look at this post for an example:

http://elsmar.com/Forums/showthread.php?p=400541#post400541


And yes I would expect a NC if it is not included - my NB has been OK with the form of the above example with recent TFs submitted for approval.

Steve
 
#4
You mention that your Notified Body has stated that missing out 62366 would be a Major NC. But remember that ALL use of standards is voluntary. What is compulsory (for EU) is to show that the Essential Requirements (Med Device Directive Annex 1) have been met. However, the use of standards is often the most straightforward way to show compliance.

Can the usability requirements of the MDD be shown just by risk management? Probably. Depends on the complexity of your device and how significant usability is to safe use.

But..
If you want to be able to claim compliance with EN 60601-1 then the collateral parts of that have to be included. So that would include EN 60601-1-6, the Usability standard. That it turn demands that a .."usability process in accordance with IEC 62366 has been undertaken." So it is difficult to sate unconditional compliance with 60601-1 without already having 62366 in place.

This situation does make things difficult, especially for products that have been on the market for some time, so they do not have the usability design process documents that 62366 demands. Over on the 62366 sub-forum there is news (from Marcello) of the development of another IEC standard to specifically address legacy devices, but this is still a couple of years (at least) away.
 

Marcelo

Inactive Registered Visitor
#5
Over on the 62366 sub-forum there is news (from Marcello) of the development of another IEC standard to specifically address legacy devices, but this is still a couple of years (at least) away.
Just a heads-up, althought the revision of IEC 62336 (creating IEC 62366-1 and IEC 62366-2) will take some years, JWG will publish an amentment to the original IEC 62366 which is the annex detailing how to deal with legacy devices in the following months (probably May/June 2013)
 
Last edited:

rob73

looking for answers
#6
3. BS EN 15223-1:2012

- A feedback from our recent Technical File review, we were told that the "Single Use Only" symbol is insufficient to state that product is Not For Reprocess. And since I can't find any "Non-reprocessable" symbol in the standard, will the use of both "Single Use Only" and "Do Not Resterilize" constitute to the meaning?"
MDD states (13.3 (f)) that "where appropriate, an indication that the device is for single use. A maufacturer's indication must be consistent across the community", so using the symbol (which is globally recognised) should not be a problem, BUT the MDD also states (13.6 (h)) that IFU must include "information on known characteristics and technical factors known to the manufacturer that could pose a risk if the device were to be reused"
Maybe this was what they where querying?
Some authorities (FDA is one I think) also that require you to show the symbols used on packaging in the IFU with a description, so for the single use symbol you could have the descriptions "single use/do not reuse/do not reprocess"


.
 
Last edited by a moderator:

renenatasha

Involved In Discussions
#7
Thanks, Pads38.
Actually, the feedback came from another consultant. Our very own NB didn't spot the missing EN 62366 standard in our ER (Essential Requirements) checklist. However, we're using EN ISO 14971 to show conformance. In another word, we're addressing the questions in Annex E of EN 62336:2008 in our Risk Analysis. Wouldn't this suffice?

Gee... you mean there's this EN 60601-1-6 that needs to be in place too? Oh dear - spare me the gallows please! :D I guess I'm gonna give EN 60601-1 a skip if it's an option.

Anyway, thanks for the info. - Rene :)
 

renenatasha

Involved In Discussions
#8
Thanks, Rob!

I guess that's what we're missing. We assume that "Do not reuse" symbol is pretty obvious that device is for single use and no reprocessing is allowed. Well, I guess it won't hurt to elaborate that in our IFU too. BUT what about product that doesn't have an IFU (e.g Class I and IIa)? Does that mean we have to include that statement on the label?

Again, thanks for your help. - Rene
 
E

ET_Marley

#9
Marcelo

Thank you for the information regarding the amendment of 62366 covering legacy systems. I had heard about this amendment and it is much needed.

How is the revision of 62366 progressing? are you still looking for comments about what the revision should bring?

Thanks

Marley
 

Marcelo

Inactive Registered Visitor
#10
How is the revision of 62366 progressing? are you still looking for comments about what the revision should bring?
The first CD of IEC 62366-1 was circulated and comments were sent. The group will revise the comment and make changes for the circulation of a CD in the next meeting of TC 62 (in April in Shanghai).

And not, I'm not looking for comments anymore, that was before the development of the working draft, were the comments could be incorporated into the document if they were reasonable. We are past this stage now.
 
Thread starter Similar threads Forum Replies Date
B Timeframe for updating QMS / transitioning from ISO 14971:2012 to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 10
Q Information for safety EN ISO 14971:2012 - Customer Risk Reduction ISO 14971 - Medical Device Risk Management 6
A Our auditor requires us to attend a training on EN ISO 14971:2012 Other ISO and International Standards and European Regulations 3
W Risk Benefit Analysis - ISO 14971:2012 Requirements ISO 14971 - Medical Device Risk Management 27
B Interpreting Deviations 5 & 6 in Annex ZA in ISO 14971:2012 ISO 14971 - Medical Device Risk Management 1
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
F ISO 14971:2012 and the FDA ISO 14971 - Medical Device Risk Management 5
M Labelling and the harmonized EN ISO 14971:2012 ISO 14971 - Medical Device Risk Management 9
F ISO 14971:2012 - 4.3 Initiating Events and Circumstances ISO 14971 - Medical Device Risk Management 4
M ISO 14971:2012 - Verification of Implementation of Risk Control Measures ISO 14971 - Medical Device Risk Management 12
M MDD Europe: Use EN ISO 14971:2012 or ISO 14971:2007 Quality Manager and Management Related Issues 9
J EN ISO 14971:2012 Annex ZA Deviation #4 Clarification ISO 14971 - Medical Device Risk Management 3
J Annex ZA of ISO 14971:2012 - Deviation 7 - Incorrect Interpretation ISO 14971 - Medical Device Risk Management 10
Sam Lazzara Experiences with Notified Bodies addressing EN ISO 14971:2012 Annex Zs ISO 14971 - Medical Device Risk Management 5
K RISK ANALYSIS SAMPLE according to Annex ZA of EN ISO-14971-2012 Other Medical Device and Orthopedic Related Topics 1
R EN ISO 14971:2012 - Implementing Annex Z content deviations ISO 14971 - Medical Device Risk Management 1
rob73 New version BS EN ISO 14971:2012 released - July 2012 ISO 14971 - Medical Device Risk Management 37
M ISO 14971 Determination of Competent Persons ISO 14971 - Medical Device Risk Management 4
M ISO 14971:2019: Criteria for overall residual risk ISO 14971 - Medical Device Risk Management 6
R Risk control measures as per ISO 14971 ISO 14971 - Medical Device Risk Management 6
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
S Practical Implementation of ISO 14971 ISO 14971 - Medical Device Risk Management 6
R Identify Medical Device characterstics as Annex C of ISO 14971 Risk Management ISO 14971 - Medical Device Risk Management 5
A ISO 14971 PFMEA Manufacturing Risk ISO 14971 - Medical Device Risk Management 2
K Overall residual risk according to ISO 14971:2019 ISO 14971 - Medical Device Risk Management 5
M Gap analysis on ISO 14971:2019 with previous revision ISO 14971 - Medical Device Risk Management 8
Bill Hansen New ISO 14971:2019 Harm: unreasonable psychological stress, and cybersecurity ISO 14971 - Medical Device Risk Management 13
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
B ISO 14971 Applied to Software ISO 14971 - Medical Device Risk Management 2
D Recent changes to ISO 14971 - SOP required for managing standard revisions ISO 13485:2016 - Medical Device Quality Management Systems 1
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 5
Ronen E Informational What's new in ISO 14971:2019 ISO 14971 - Medical Device Risk Management 2
T ISO 14971-2019 doubt - Evaluate if estimated risks are acceptable ISO 14971 - Medical Device Risk Management 9
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
Y When will Notified Bodies require MedDev manufacturers to fully implement ISO 14971:2019? ISO 14971 - Medical Device Risk Management 1
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
S ISO 14971 Risk Management - Questions for Hazard identification ISO 14971 - Medical Device Risk Management 2
M Informational ISO 14971 / ISO TR 24971 revision update – atualizações sobre a revisão Medical Device and FDA Regulations and Standards News 1
R The difference b/w FMEA & Risk analysis as per iso 14971 ISO 14971 - Medical Device Risk Management 8
D Risk management according to ISO 14971 - When to document risk controls? ISO 14971 - Medical Device Risk Management 10
D Where does FMEA fit in your ISO 14971 Risk Management process? ISO 14971 - Medical Device Risk Management 13
M Informational ISO TC 210 JWG 1 meeting in São Paulo – Revision of ISO 14971 and ISO TR 24971 – Medical Device Risk Management Medical Device and FDA Regulations and Standards News 0
S In a risk analysis, how can we tie mobile app security breach to ISO 14971? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
F IMDRF opened a Consultation on Annex E & F and the link to ISO 14971 ISO 14971 - Medical Device Risk Management 4
M Updates to EN 62366 & ISO 14971? Other Medical Device Related Standards 3
D IEC 60601-1 and ISO 14971 Assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
M Example ISO 14971 policy and risk criteria ISO 14971 - Medical Device Risk Management 0
P ISO 13485 and ISO 14971 - one mandates the other? ISO 13485:2016 - Medical Device Quality Management Systems 8
D Rationale for Risk Acceptability Matrix - ISO 14971 ISO 14971 - Medical Device Risk Management 9