M

MIREGMGR

#21
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

What type of questions should be asked to the R&D, Mfr Depts that will be initiating the change?
You should be starting from the decision-question-list in the FDA deciding-whether-to-submit guidance.
 
Elsmar Forum Sponsor
D

drewsky1

#23
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

:thanks:yes, that would be best. Just wanted to get input on other SOP's that other's may be using.

Thanks
 
C

cff920

#24
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

Lately, it appears that FDA is making efforts (or struggling with) to clarify two grey areas:
(1) Malfunction and the criteria of its reportability;
(2) Threshold that a design change to trigger 510(k) submission

Both areas are sensitive, and a firm could easily be given 483 and Warning Letter, if any violation is detected in FDA inspection.

But, despite the suffering that the industry had, FDA has the most comprehensive and informative database of device adverse events among all regulators. I personally respect FDA's achievements in this, although I dislike FDA.
 
S

ss2013

#25
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

I AM SEEKING A GENERAL TEMPLATE/SAMPEL OF A 510(K) LETTER TO FILE (ADD TO FILE) FOR INTERNAL FILING.

DOES ANYONE HAVE A REALLY GOOD TEMPLATE/TOC/SAMPLE?

THANK YOU!!

:thanx:
 
M

MIREGMGR

#26
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

US FDA today issued a 114-page document, "Report on FDA's Policy to be Proposed Regarding Premarket Notification Requirements for Modifications to Legally Marketed Devices", available at http://www.fda.gov/downloads/Medica...ivery&utm_medium=email&utm_source=govdelivery

Summarizing, they say they're going to keep the K97-1 process, but tune it up in several ways.

Next step is a draft guidance.
 
S

ss2013

#27
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

Does anyone have a current template for Letter to File with cover or statement?

Thanks
 
T

tallu

#28
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

Could some one provide "table of content" for Letter to File?

Or is it something like this:
- design change input
- risk management review
- in-house v&v activities
- applicable 60601 series test reports

What needs to be included?
 
S

ss2013

#29
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

This is great. I truly appreciate this new information!!

ss2013
 

Ronen E

Problem Solver
Staff member
Moderator
#30
Re: Change of Existing Medical Devices and 510(k) - "Letter to File" and "Add to File

Could some one provide "table of content" for Letter to File?

Or is it something like this:
- design change input
- risk management review
- in-house v&v activities
- applicable 60601 series test reports

What needs to be included?
Hi,

In my opinion this is way too much. All these belong in the DHF.

A letter to file should only contain a brief (yet unambiguous) description of the change, or a reference to other documents that describe it; and a thorough documented analysis showing how - by FDA standards (at the moment, it's mostly K97-1) - the change doesn't require a 510k.

It's "a letter", not "a book" :lol:

Cheers,
Ronen.
 
Thread starter Similar threads Forum Replies Date
K Change of Existing Medical Devices and 510(k) US Food and Drug Administration (FDA) 9
N Is a new 510K required for change of manufacturing location for an existing device? US Food and Drug Administration (FDA) 1
V Procedure to change the Company name in existing TS16949-2009 certificate ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
shimonv Deciding When to Submit a 510(k) for a Change to an Existing Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
N Company Name Change - Existing Inventory with Old Company Name Other US Medical Device Regulations 9
sagai New Draft Guidance - When to Submit a 510(k) for a Change to an Existing Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
C Design Change needed for Changing an Angle in existing Product from 15-25 degrees? ISO 13485:2016 - Medical Device Quality Management Systems 11
S Time for a Special 510k? Making a formulation change to an existing 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
A Submit a 510(k) to an existing device vs. Just Documenting a Change ISO 13485:2016 - Medical Device Quality Management Systems 2
M Change of Organization Chart - The effect on existing procedures Document Control Systems, Procedures, Forms and Templates 4
Ed Panek Notified Body Substantial Change MDD EU Medical Device Regulations 1
A Purchasing Controls on Change Management of Literature/Labeling ISO 13485:2016 - Medical Device Quality Management Systems 3
A For software change - New Channel of interoperability CE Marking (Conformité Européene) / CB Scheme 4
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
R AS9102 FAI Change in Material / Process Supplier AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Time change via change of Time Zone Qualification and Validation (including 21 CFR Part 11) 3
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
U Power supply - Change From 160 to 300 Watts - Significant change? EU Medical Device Regulations 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
M Label change - 510(k) Medical Device and FDA Regulations and Standards News 5
M EMS change management ISO 14001:2015 Specific Discussions 2
E Document Change Request forms (or no forms)? ISO 13485:2016 - Medical Device Quality Management Systems 9
J Significant change related to design and intended use EU Medical Device Regulations 3
J Need Change Control Yes/No Decision Tree Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
S Cleaning process center change ISO 13485:2016 - Medical Device Quality Management Systems 4
U NOC - What is considered a "design change" EU Medical Device Regulations 5
qualprod What do CB´s change when doing a partial moving ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Y Algo change significancy Medical Information Technology, Medical Software and Health Informatics 3
Revision36 Engineering Change Order Industry Standards Other Medical Device and Orthopedic Related Topics 4
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 12
D ECO (Engineering Change Order) process questions ISO 13485:2016 - Medical Device Quality Management Systems 7
R Medical Device - Change manufacturing plant Design and Development of Products and Processes 6
Tagin Template or Checklist for Process Change Document Control Systems, Procedures, Forms and Templates 5
R Do we need issue ECN (Engineering Change Notice) towards updated Material Specification? Design and Development of Products and Processes 2
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
M Change management procedure when 7.3 is not applicable ISO 13485:2016 - Medical Device Quality Management Systems 5
D Change Approval Requirements - Does every change need formal customer approval? Design and Development of Products and Processes 17
Q Change the shape of the raw material FAI AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Engineering Change Order (ECO) metrics to track ISO 13485:2016 - Medical Device Quality Management Systems 6
P Notified Body Approval of Change EU Medical Device Regulations 2
F Change to Formulation due to Reach (China Medical Device Regulations) China Medical Device Regulations 0
E Change in control plan - Do I have to do sampling? IATF 16949 - Automotive Quality Systems Standard 1
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
M Address change for a company with CE/ISO13485 EU Medical Device Regulations 2
G Need to change KPI we called NC parts (maximum 3%.) to FTQ (first time quality) IATF 16949 - Automotive Quality Systems Standard 4
J FDA notification of address change US Food and Drug Administration (FDA) 2

Similar threads

Top Bottom