Hello all,
It is a power adaptor under test. The plastic enclosure of the adaptor is 81[FONT=宋]℃ at maximum ambient of 40[FONT=宋]℃[/FONT]. Table 23 specifies the maximum limit of 86[FONT=宋]℃[/FONT] for t <1s and 71[FONT=宋]℃[/FONT] for 1s£ t<10s. [/FONT]
The standard also says the probability of occurrence of contact and of the duration of contact is determined by the RMF. We then documented t < 1s rather than 1s < t< 10s in the RMF just because of we think the power adaptor is not a handle or control panel which in normal use would have a brief or long contact with a user. So, the measured temperature 81[FONT=宋]℃ is below the limit 86[FONT=宋]℃[/FONT]. Is it reasonable? [/FONT]
In addition, could someone give some examples (equipment parts) for each contact time specified in the table 23 of IEC 60601-1:2005?
Thanks and regards,
Roland
It is a power adaptor under test. The plastic enclosure of the adaptor is 81[FONT=宋]℃ at maximum ambient of 40[FONT=宋]℃[/FONT]. Table 23 specifies the maximum limit of 86[FONT=宋]℃[/FONT] for t <1s and 71[FONT=宋]℃[/FONT] for 1s£ t<10s. [/FONT]
The standard also says the probability of occurrence of contact and of the duration of contact is determined by the RMF. We then documented t < 1s rather than 1s < t< 10s in the RMF just because of we think the power adaptor is not a handle or control panel which in normal use would have a brief or long contact with a user. So, the measured temperature 81[FONT=宋]℃ is below the limit 86[FONT=宋]℃[/FONT]. Is it reasonable? [/FONT]